Airsense 10 auto

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
curiousbetsy69
Posts: 6
Joined: Fri Jun 21, 2019 7:18 am

Airsense 10 auto

Post by curiousbetsy69 » Wed Jun 26, 2019 5:02 pm

Undiagnosed- buying tommorow

when i buy the machine i suppose ill reset it then set the date and time. Apparently it comes with mask hose cord and a card in it.

i know nothing about how this machine or sleep apnea works, and so i was wondering where i can read up on pressures, what pressure i should start with, and how to get and interpret my data with this machine.

Am i ast risk of causing more health problems by using the wrong pressure, or is it more risky to not use the machine entirely. Are there helpful videos i can watch about how to get myself set up right and how to manage/monitor the pressure setting i need?

User avatar
Grace~~~
Posts: 662
Joined: Tue May 10, 2016 3:27 pm
Location: Florida

Re: Airsense 10 auto

Post by Grace~~~ » Wed Jun 26, 2019 5:10 pm

Hi curiousbetsy :D

I am always impressed with proactive people.

Others will be along to welcome and advise you soon, but I was wondering if you would be willing to share what caused you to seek out a cpap?

What health conditions are you hoping to fix? Do you have friends or relatives that use a cpap.

The airsense AUTOSET is one of the best machines you can get. So you're already on your way to a good start.
Began XPAP May 2016. Autoset Pressure min. 8 / max 15. Ramp off. ERP set at 2. No humidity. Sleepyhead software installed and being looked at daily, though only beginning to understand the data.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Airsense 10 auto

Post by palerider » Wed Jun 26, 2019 5:15 pm

curiousbetsy69 wrote: ↑
Wed Jun 26, 2019 5:02 pm
Undiagnosed- buying tommorow

when i buy the machine i suppose ill reset it then set the date and time. Apparently it comes with mask hose cord and a card in it.

i know nothing about how this machine or sleep apnea works, and so i was wondering where i can read up on pressures, what pressure i should start with, and how to get and interpret my data with this machine.

Am i ast risk of causing more health problems by using the wrong pressure, or is it more risky to not use the machine entirely. Are there helpful videos i can watch about how to get myself set up right and how to manage/monitor the pressure setting i need?
Start with a min pressure of 6,and a Max of 20, then post some charts,as directed in the "NEWBIES,READ THIS FIRST!" post.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11501
Joined: Fri Jun 19, 2015 4:36 pm

Re: Airsense 10 auto

Post by zonker » Wed Jun 26, 2019 5:21 pm

curiousbetsy69 wrote: ↑
Wed Jun 26, 2019 5:02 pm
Undiagnosed- buying tommorow
welcome to the zoo! when you have your machine, please come back to this post. set up as palerider suggests. also as he suggests, read the newbies post. it's the link at the bottom of my post.

you aren't going to harm yourself. it may take awhile, but your health and sleep WILL improve.

please read all the permanent posts that are the first thing you see on the page. some of it will be outdated, maybe even with some broken links, but it will help familiarize you with the lingo.

others will be along with more advice.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

curiousbetsy69
Posts: 6
Joined: Fri Jun 21, 2019 7:18 am

Re: Airsense 10 auto

Post by curiousbetsy69 » Wed Jun 26, 2019 5:35 pm

Grace~~~ wrote: ↑
Wed Jun 26, 2019 5:10 pm
Hi curiousbetsy :D

I am always impressed with proactive people.

Others will be along to welcome and advise you soon, but I was wondering if you would be willing to share what caused you to seek out a cpap?

What health conditions are you hoping to fix? Do you have friends or relatives that use a cpap.

The airsense AUTOSET is one of the best machines you can get. So you're already on your way to a good start.
i havent slept right in years, and i feel sick as hell. i stopped breathing when i was a child. i cant get deep sleep anymore so i figure its worth a shot.

curiousbetsy69
Posts: 6
Joined: Fri Jun 21, 2019 7:18 am

Re: Airsense 10 auto

Post by curiousbetsy69 » Wed Jun 26, 2019 5:40 pm

palerider wrote: ↑
Wed Jun 26, 2019 5:15 pm
curiousbetsy69 wrote: ↑
Wed Jun 26, 2019 5:02 pm
Undiagnosed- buying tommorow

when i buy the machine i suppose ill reset it then set the date and time. Apparently it comes with mask hose cord and a card in it.

i know nothing about how this machine or sleep apnea works, and so i was wondering where i can read up on pressures, what pressure i should start with, and how to get and interpret my data with this machine.

Am i ast risk of causing more health problems by using the wrong pressure, or is it more risky to not use the machine entirely. Are there helpful videos i can watch about how to get myself set up right and how to manage/monitor the pressure setting i need?
Start with a min pressure of 6,and a Max of 20, then post some charts,as directed in the "NEWBIES,READ THIS FIRST!" post.


Should i use sleepy head or oscar

curiousbetsy69
Posts: 6
Joined: Fri Jun 21, 2019 7:18 am

Re: Airsense 10 auto

Post by curiousbetsy69 » Wed Jun 26, 2019 5:44 pm

palerider wrote: ↑
Wed Jun 26, 2019 5:15 pm
curiousbetsy69 wrote: ↑
Wed Jun 26, 2019 5:02 pm
Undiagnosed- buying tommorow

when i buy the machine i suppose ill reset it then set the date and time. Apparently it comes with mask hose cord and a card in it.

i know nothing about how this machine or sleep apnea works, and so i was wondering where i can read up on pressures, what pressure i should start with, and how to get and interpret my data with this machine.

Am i ast risk of causing more health problems by using the wrong pressure, or is it more risky to not use the machine entirely. Are there helpful videos i can watch about how to get myself set up right and how to manage/monitor the pressure setting i need?
Start with a min pressure of 6,and a Max of 20, then post some charts,as directed in the "NEWBIES,READ THIS FIRST!" post.
i see oscar had actually replaced it. does the software still look the same? will reading this guide to sleepyhead be beneficial for using oscar?

User avatar
Thumper68
Posts: 129
Joined: Sun Sep 16, 2018 9:48 am

Re: Airsense 10 auto

Post by Thumper68 » Wed Jun 26, 2019 5:51 pm

curiousbetsy69 wrote: ↑
Wed Jun 26, 2019 5:44 pm
palerider wrote: ↑
Wed Jun 26, 2019 5:15 pm
curiousbetsy69 wrote: ↑
Wed Jun 26, 2019 5:02 pm

i see oscar had actually replaced it. does the software still look the same? will reading this guide to sleepyhead be beneficial for using oscar?
To me Oscar looks the same. the sleepyhead guide should be beneficial as well. Oscar was forked from sleepyhead to fix bugs and provide better support.

_________________
Mask: DreamWear Gel Nasal Pillow CPAP Mask with Headgear - Fit Pack (All Nasal Pillows with Medium Frame)
Additional Comments: Settings 11-20, EPR 2. Dreamstation Full Auto for back up

User avatar
zonker
Posts: 11501
Joined: Fri Jun 19, 2015 4:36 pm

Re: Airsense 10 auto

Post by zonker » Wed Jun 26, 2019 6:07 pm

curiousbetsy69 wrote: ↑
Wed Jun 26, 2019 5:44 pm

i see oscar had actually replaced it. does the software still look the same? will reading this guide to sleepyhead be beneficial for using oscar?
yup, same-same.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Airsense 10 auto

Post by palerider » Wed Jun 26, 2019 7:41 pm

curiousbetsy69 wrote: ↑
Wed Jun 26, 2019 5:44 pm
palerider wrote: ↑
Wed Jun 26, 2019 5:15 pm
curiousbetsy69 wrote: ↑
Wed Jun 26, 2019 5:02 pm
Undiagnosed- buying tommorow

when i buy the machine i suppose ill reset it then set the date and time. Apparently it comes with mask hose cord and a card in it.

i know nothing about how this machine or sleep apnea works, and so i was wondering where i can read up on pressures, what pressure i should start with, and how to get and interpret my data with this machine.

Am i ast risk of causing more health problems by using the wrong pressure, or is it more risky to not use the machine entirely. Are there helpful videos i can watch about how to get myself set up right and how to manage/monitor the pressure setting i need?
Start with a min pressure of 6,and a Max of 20, then post some charts,as directed in the "NEWBIES,READ THIS FIRST!" post.
i see oscar had actually replaced it. does the software still look the same? will reading this guide to sleepyhead be beneficial for using oscar?
Yes, Oscar is *very* similar at this point, that's why Pugsy hasn't bothered to update the instructions... Now, when the next version comes out, they'll probably need updating.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Airsense 10 auto

Post by palerider » Wed Jun 26, 2019 7:42 pm

curiousbetsy69 wrote: ↑
Wed Jun 26, 2019 5:40 pm
palerider wrote: ↑
Wed Jun 26, 2019 5:15 pm
curiousbetsy69 wrote: ↑
Wed Jun 26, 2019 5:02 pm
Undiagnosed- buying tommorow

when i buy the machine i suppose ill reset it then set the date and time. Apparently it comes with mask hose cord and a card in it.

i know nothing about how this machine or sleep apnea works, and so i was wondering where i can read up on pressures, what pressure i should start with, and how to get and interpret my data with this machine.

Am i ast risk of causing more health problems by using the wrong pressure, or is it more risky to not use the machine entirely. Are there helpful videos i can watch about how to get myself set up right and how to manage/monitor the pressure setting i need?
Start with a min pressure of 6,and a Max of 20, then post some charts,as directed in the "NEWBIES,READ THIS FIRST!" post.


Should i use sleepy head or oscar
Might as well use Oscar, it's supported now, whereas sleepyhead was strangled in it's sleep by it's creator in a fit of rage.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11501
Joined: Fri Jun 19, 2015 4:36 pm

Re: Airsense 10 auto

Post by zonker » Wed Jun 26, 2019 9:14 pm

palerider wrote: ↑
Wed Jun 26, 2019 7:41 pm

Yes, Oscar is *very* similar at this point, that's why Pugsy hasn't bothered to update the instructions... Now, when the next version comes out, they'll probably need updating.
ooohhh??
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Airsense 10 auto

Post by Miss Emerita » Wed Jun 26, 2019 9:30 pm

I think it’s great you are going to try CPAP therapy! The Oscar link is in palerider’s signature block, and here it is again:

https://www.sleepfiles.com/OSCAR/

When you get the machine, let us know if you need help actually setting the pressures. There are some other settings too, e.g., humidity, ramp, and EPR (basically lowers the pressure a bit every time you breathe out).

Have you thought about what kind of mask you’d like to try? If not, I guess the first question is whether you need to breath through you mouth at night.

And a tip: try using the machine and mask for the first time during the day, ideally while you’re doing something enjoyable, like reading or watching TV. This will help you get used to the general feel of it all without the extra anxiety of trying to get some sleep.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Airsense 10 auto

Post by palerider » Thu Jun 27, 2019 12:37 am

zonker wrote: ↑
Wed Jun 26, 2019 9:14 pm
palerider wrote: ↑
Wed Jun 26, 2019 7:41 pm

Yes, Oscar is *very* similar at this point, that's why Pugsy hasn't bothered to update the instructions... Now, when the next version comes out, they'll probably need updating.
ooohhh??
ooohhh? what? software evolves, changes, there's some crap in sleepyhead that's being eliminated, there's some stuff that confuses most people that's been clarified, and who knows what else.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65416
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Airsense 10 auto

Post by Pugsy » Thu Jun 27, 2019 9:23 am

palerider wrote: ↑
Wed Jun 26, 2019 7:41 pm
Yes, Oscar is *very* similar at this point, that's why Pugsy hasn't bothered to update the instructions... Now, when the next version comes out, they'll probably need updating.
This may come as a big surprise (or it may not) to everyone but redoing that ancient tutorial is the last thing on my to do list. :lol: :shock:

It needs to been done though....but no sense in doing it until OSCAR is more stable....and I have a larger supply of Valium and maybe a gallon of my homemade Irish Creme stuff.

Fair warning though...it will be read only. I will force people who have questions or problems or want something "analyzed" to create a new topic/thread. :lol: :lol: :lol: It is commonly known that I hate it when I say "don't ask questions here but create your own topic" and what do they do....they post in the tutorial. That will stop. :shock:

For the time being OSCAR is pretty much a clone of SleepyHead with some minor bug fixes and minor changes...the outdated tutorial for SleepyHead will work well enough for OSCAR for the time being.

In the meantime....I might be up for donations of suitable drugs to help me get through another tutorial. :D

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.