Autoset 10 by Resmed stays on 4

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
MarkDeLange
Posts: 14
Joined: Wed Sep 19, 2018 1:03 pm

Autoset 10 by Resmed stays on 4

Post by MarkDeLange » Sat Jun 22, 2019 10:58 am

Changed from cpap to auto, set high and low

8. And 17
Machine will not move off of 4

Ramp off.
Perplexed

Mark

_________________
Mask: Amara View Full Face CPAP Mask with Headgear - Fit Pack
Additional Comments: Sleepyhead software

User avatar
zonker
Posts: 11288
Joined: Fri Jun 19, 2015 4:36 pm

Re: Autoset 10 by Resmed stays on 4

Post by zonker » Sat Jun 22, 2019 11:12 am

MarkDeLange wrote:
Sat Jun 22, 2019 10:58 am
Changed from cpap to auto, set high and low

8. And 17
Machine will not move off of 4

Ramp off.
Perplexed

Mark
do you know this through the machines screen or via software, such as sleepyhead or oscar?

if the later, please post some charts so experts can see.

not being an expert, i'm stymied as to how your auto machine won't move off of a minimum of four.

has it always done this?

silly question-have you accidentally set the machine to cpap mode?
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Dog Slobber
Posts: 4215
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Autoset 10 by Resmed stays on 4

Post by Dog Slobber » Sat Jun 22, 2019 11:17 am

Occasionally people make claims about machines not behaving how the settings say they should.

How do I put this gently?

A lot of the time it's not the machine.

Take some pics of the settings.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Autoset 10 by Resmed stays on 4

Post by Wulfman... » Sat Jun 22, 2019 7:07 pm

MarkDeLange wrote:
Sat Jun 22, 2019 10:58 am
Changed from cpap to auto, set high and low

8. And 17
Machine will not move off of 4

Ramp off.
Perplexed

Mark
If you don't have any flow limitations or snoring.......it won't increase pressures.
Those two are what primarily trigger pressure increases.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Dog Slobber
Posts: 4215
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Autoset 10 by Resmed stays on 4

Post by Dog Slobber » Sat Jun 22, 2019 7:15 pm

Wulfman... wrote:
Sat Jun 22, 2019 7:07 pm
MarkDeLange wrote:
Sat Jun 22, 2019 10:58 am
Changed from cpap to auto, set high and low

8. And 17
Machine will not move off of 4

Ramp off.
Perplexed

Mark
If you don't have any flow limitations or snoring.......it won't increase pressures.
Those two are what primarily trigger pressure increases.


Den

.
The OP is making the claim that the range is set to 8-17, but won't move off of 4. Ramp is off.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Autoset 10 by Resmed stays on 4

Post by Wulfman... » Sat Jun 22, 2019 7:27 pm

Dog Slobber wrote:
Sat Jun 22, 2019 7:15 pm
Wulfman... wrote:
Sat Jun 22, 2019 7:07 pm
MarkDeLange wrote:
Sat Jun 22, 2019 10:58 am
Changed from cpap to auto, set high and low

8. And 17
Machine will not move off of 4

Ramp off.
Perplexed

Mark
If you don't have any flow limitations or snoring.......it won't increase pressures.
Those two are what primarily trigger pressure increases.


Den

.
The OP is making the claim that the range is set to 8-17, but won't move off of 4. Ramp is off.
Yeah. Something is not right with the setup. If the OP is using software, the problem SHOULD be visible.

Edit to add:

Maybe consulting the clinical manual would help (if the OP doesn't have a copy).

https://www.respshop.com/manuals/ResMed ... %20her.pdf


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05