Precision, pressure stability and pump performance test on 10 popular CPAP devices

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
pcstud
Posts: 38
Joined: Tue Feb 19, 2019 2:40 pm

Precision, pressure stability and pump performance test on 10 popular CPAP devices

Post by pcstud » Fri Jun 21, 2019 7:10 am

I translated my last test for you. I think it's worth it.

Are you sure that your machine works well with a high leak? What leaks are acceptable? Are the Chinese machines as good as the leading manufacturers? Which CPAP is the best having regard to those criteria?

https://www.cpapblog.pl/2019/05/precisi ... -CPAP.html
cpaptest.png
cpaptest.png (944.2 KiB) Viewed 3354 times

_________________
MachineMaskHumidifier

yrnkrn
Posts: 172
Joined: Sat Feb 12, 2011 1:07 pm

Re: Precision, pressure stability and pump performance test on 10 popular CPAP devices

Post by yrnkrn » Fri Jun 21, 2019 8:52 am

Nice work!
Do you know why the high-quality full-sized machines all output 14.65 instead of 15 and 19.55 instead of 20?
Could this be systematic measurement error?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Gel Nasal Pillow CPAP Mask with Headgear - Fit Pack (All Nasal Pillows with Medium Frame)
Additional Comments: OSCAR, CPAP=6, https://sites.google.com/view/palatal-prolapse/

pcstud
Posts: 38
Joined: Tue Feb 19, 2019 2:40 pm

Re: Precision, pressure stability and pump performance test on 10 popular CPAP devices

Post by pcstud » Fri Jun 21, 2019 9:18 am

yrnkrn wrote:
Fri Jun 21, 2019 8:52 am
Nice work!
Do you know why the high-quality full-sized machines all output 14.65 instead of 15 and 19.55 instead of 20?
Could this be systematic measurement error?
maybe… each CPAP has a 0.4 to 1 cm possible error according to the specifications (ResMed and Philips 0.5 cm H2O)

the measuring device was calibrated and its accuracy is 0.1, when the indications were fluctuating, then we note 0.05

_________________
MachineMaskHumidifier

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Precision, pressure stability and pump performance test on 10 popular CPAP devices

Post by palerider » Fri Jun 21, 2019 7:19 pm

Your 'large leak' test is absurd.

Sure, if you have a leak that's higher than the machine is spec'd to tolerate, it's *not going to tolerate it*.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR