Sleepyhead graph question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Woody
Posts: 591
Joined: Sat Mar 10, 2007 7:59 pm

Sleepyhead graph question

Post by Woody » Thu Jun 20, 2019 8:12 pm

Why would it raise my pressure for a couple of minutes if there were no events ?

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: SleapyHead software
Attachments
screenshot-20190620-220159.png
screenshot-20190620-220159.png (192.36 KiB) Viewed 2773 times

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepyhead graph question

Post by Pugsy » Thu Jun 20, 2019 8:28 pm

That looks like a zoomed in view of the normal test pressure probe that Respironics auto adjusting machines will do.

It's usually a 1.5 or there abouts increase in pressure just to sort of test things out. Has nothing to do with specific events happening or not happening.

This is old but it still applies today.
viewtopic/t80875/Respironics-APAP-press ... se-up.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Woody
Posts: 591
Joined: Sat Mar 10, 2007 7:59 pm

Re: Sleepyhead graph question

Post by Woody » Thu Jun 20, 2019 8:53 pm

Thanks for explaining that.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: SleapyHead software