Using machine but still waking up gasping for air

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
michael023
Posts: 2
Joined: Mon Jun 03, 2019 1:50 pm

Using machine but still waking up gasping for air

Post by michael023 » Mon Jun 03, 2019 2:46 pm

Hi.

I'm a 19-year-old male and I was diagnosed with moderate obstructive sleep apnea 3 months ago. I'm using a DreamStation Auto CPAP machine and a DreamWear Nasal Mask. As you can see in the image attached, I have a lot of events for the first one or two hours of sleep and then it normalizes. I wake up gasping for air many times in those two hours. I've tried everything, I've set the pressure to 6, 8, 10, enabled A-Flex, disabled it. Nothing seems to work.
What can be causing this? Thanks and sorry for my bad English.

_________________
MachineMask
Attachments
screenshot-20190603-152608.png
screenshot-20190603-152608.png (105.01 KiB) Viewed 858 times

User avatar
LSAT
Posts: 13399
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Using machine but still waking up gasping for air

Post by LSAT » Mon Jun 03, 2019 5:41 pm

I would set the pressure at 9/20 and leave it there for several days. Try to avoid turning off the machine so often. Set A flex to 2.

User avatar
djams
Posts: 927
Joined: Mon Jun 18, 2018 4:37 pm
Location: DFW area

Re: Using machine but still waking up gasping for air

Post by djams » Mon Jun 03, 2019 6:57 pm

michael023 wrote:
Mon Jun 03, 2019 2:46 pm
Hi.

I'm a 19-year-old male and I was diagnosed with moderate obstructive sleep apnea 3 months ago. I'm using a DreamStation Auto CPAP machine and a DreamWear Nasal Mask. As you can see in the image attached, I have a lot of events for the first one or two hours of sleep and then it normalizes. I wake up gasping for air many times in those two hours. I've tried everything, I've set the pressure to 6, 8, 10, enabled A-Flex, disabled it. Nothing seems to work.
What can be causing this? Thanks and sorry for my bad English.
Any idea what time you fell asleep last night?

Good for you getting started at such a young age. You're doing yourself a huge favor here. Assuming that you want to live a long, healthy life. :D

And your English is great, BTW. I saw your comment and it took me by complete surprise. Went back and re-read your entire post. You're good on the English. Maybe should say "poor English", but since you don't have poor English, you really don't even need to know that. :lol:

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins

michael023
Posts: 2
Joined: Mon Jun 03, 2019 1:50 pm

Re: Using machine but still waking up gasping for air

Post by michael023 » Wed Jun 05, 2019 3:32 pm

LSAT wrote:
Mon Jun 03, 2019 5:41 pm
I would set the pressure at 9/20 and leave it there for several days. Try to avoid turning off the machine so often. Set A flex to 2.
That's the pressure I use normally. In the image it's set to 6/15 because I was testing if I didn't wake up with that pressure.
The weird thing is that sometimes I wake up once or twice and sometimes I wake up many times before falling asleep.
djams wrote:
Mon Jun 03, 2019 6:57 pm
michael023 wrote:
Mon Jun 03, 2019 2:46 pm
Hi.

I'm a 19-year-old male and I was diagnosed with moderate obstructive sleep apnea 3 months ago. I'm using a DreamStation Auto CPAP machine and a DreamWear Nasal Mask. As you can see in the image attached, I have a lot of events for the first one or two hours of sleep and then it normalizes. I wake up gasping for air many times in those two hours. I've tried everything, I've set the pressure to 6, 8, 10, enabled A-Flex, disabled it. Nothing seems to work.
What can be causing this? Thanks and sorry for my bad English.
Any idea what time you fell asleep last night?

Good for you getting started at such a young age. You're doing yourself a huge favor here. Assuming that you want to live a long, healthy life. :D

And your English is great, BTW. I saw your comment and it took me by complete surprise. Went back and re-read your entire post. You're good on the English. Maybe should say "poor English", but since you don't have poor English, you really don't even need to know that. :lol:
I went to sleep at 3:00 AM but I kept waking up gasping for air till 4:30 AM. So I fell asleep at 4:30 AM.
And I said that my English is bad just in case since English is not my first language :lol:

_________________
MachineMask

User avatar
katestyles
Posts: 610
Joined: Sun Jan 06, 2019 9:08 am

Re: Using machine but still waking up gasping for air

Post by katestyles » Wed Jun 05, 2019 3:48 pm

michael023 wrote:
Wed Jun 05, 2019 3:32 pm
I went to sleep at 3:00 AM but I kept waking up gasping for air till 4:30 AM. So I fell asleep at 4:30 AM.
And I said that my English is bad just in case since English is not my first language
It looks as if all those events at the start are before you fall asleep properly.

Once you are asleep, you do well with 10cm of pressure.

I agree with LSAT. Set the machine to min 9, max 20, let it do its job. Your job is to work on falling asleep.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Back up mask - anything in the drawer