Setting Min and Max

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
babybear
Posts: 54
Joined: Sun Sep 09, 2018 5:49 am

Setting Min and Max

Post by babybear » Tue May 21, 2019 11:55 am

Hello

I don't how much of a difference I need to have with the min and max settings. Would a min of 10 and a max of 12 (epr 3) work or do I need a wider range? Thank you for your time and advice in advance. :)

User avatar
Pugsy
Posts: 65418
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Setting Min and Max

Post by Pugsy » Tue May 21, 2019 12:06 pm

As long as your maximum is never ever reached it doesn't really matter what it is.
There's nothing magical about a tight range or a wide range.
The minimum is the most critical setting...not the maximum.
If you never ever ever reach 12....won't hurt to use 12 at the max....but just bear in mind that when you restrict the max if something changes in terms of pressure needs the machine will have its hands tied.
If you have a wider range it doesn't tie the machines hands plus it won't go anywhere without a good reason anyway.

Now if going higher causes problems like aerophagia issues...then by all means limit the max somewhat.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Setting Min and Max

Post by palerider » Tue May 21, 2019 3:28 pm

babybear wrote: ↑
Tue May 21, 2019 11:55 am
Hello

I don't how much of a difference I need to have with the min and max settings. Would a min of 10 and a max of 12 (epr 3) work or do I need a wider range? Thank you for your time and advice in advance. :)
Anybody that says you 'need' a 'tight range' doesn't understand what they're doing.

What is important, as Pugsy says, is the MINIMUM pressure. the Max pressure is immaterial for most people, and is best left at whatever the machine max setting is, (20 or 25 for most of them).

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.