yearning 4zs therapy -- not going well!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

yearning 4zs therapy -- not going well!

Post by yearning4zs » Thu May 16, 2019 12:12 am

I have low recorded ahi, but still wake up in the morning with headache and strong heartbeat and a sort of "hypoxic" feeling, like perhaps a bit dizzy. I've asked a bunch of doctors and they say your data looks great. I'm going in for a titration soon, but would like to figure out some stuff in the mean time. One strange thing I've noticed is that the inspiration on my wave form will have a very obvious uptick that would seem to indicate resistance. However, when I look at flow limit, there is usually no flow limit apparent. Has anyone seen waveforms like this before and know what they may indicate?

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable
Attachments
waveform.jpg
waveform.jpg (422.91 KiB) Viewed 5084 times
Last edited by yearning4zs on Fri May 24, 2019 12:29 am, edited 3 times in total.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: strange flow rate waveform question

Post by palerider » Thu May 16, 2019 12:15 am


_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

Re: strange flow rate waveform question

Post by yearning4zs » Thu May 16, 2019 12:22 am

palerider wrote:
Thu May 16, 2019 12:15 am
Yup... search.php?keywords=ballistocardiographic
Great, thanks so much. So I don't really need to worry about that then? It's strange because previously they never showed up. Only since I've started using the resmed 10 for her have I seen them.

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: strange flow rate waveform question

Post by palerider » Thu May 16, 2019 12:44 am

yearning4zs wrote:
Thu May 16, 2019 12:22 am
palerider wrote:
Thu May 16, 2019 12:15 am
Yup... search.php?keywords=ballistocardiographic
Great, thanks so much. So I don't really need to worry about that then? It's strange because previously they never showed up. Only since I've started using the resmed 10 for her have I seen them.
Could be you're sleeping differently, or the machine is just a bit more sensitive... what did you have before?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

Re: strange flow rate waveform question

Post by yearning4zs » Thu May 16, 2019 12:48 am

looked more like this. I would say overall I feel worse on the new machine.

edit: I had a resmed 10 for her before as well, but it was a rental. Perhaps had been used longer?

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable
Attachments
waveform2.jpg
waveform2.jpg (434.22 KiB) Viewed 5073 times

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: strange flow rate waveform question

Post by palerider » Thu May 16, 2019 1:16 am

yearning4zs wrote:
Thu May 16, 2019 12:48 am
looked more like this. I would say overall I feel worse on the new machine.

edit: I had a resmed 10 for her before as well, but it was a rental. Perhaps had been used longer?
I see the same artifacts in that trace..

I think as you scroll through the night, you'll see variations... what you don't have is a nice smooth exhalation... you're seeing the cardiac effects in that, and it just depends on *when* you inhale in or out of time with your heartbeat determines what kind of strange shape the top of your inhalation curve looks like.

Are the settings on the new machine the same as the old one?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

Re: strange flow rate waveform question

Post by yearning4zs » Thu May 16, 2019 1:21 am

palerider wrote:
Thu May 16, 2019 1:16 am
yearning4zs wrote:
Thu May 16, 2019 12:48 am
looked more like this. I would say overall I feel worse on the new machine.

edit: I had a resmed 10 for her before as well, but it was a rental. Perhaps had been used longer?
I see the same artifacts in that trace..

I think as you scroll through the night, you'll see variations... what you don't have is a nice smooth exhalation... you're seeing the cardiac effects in that, and it just depends on *when* you inhale in or out of time with your heartbeat determines what kind of strange shape the top of your inhalation curve looks like.

Are the settings on the new machine the same as the old one?
I believe the main difference is that I upped the max to 8.6, whereas it was at 6.8 before (it seemed to want to go higher). I also have the machine on soft response, which I might not have before. I think everything else is more or less the same. I have it on auto-climate control now. I think before it was on manual with humidity at 3.

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: strange flow rate waveform question

Post by palerider » Thu May 16, 2019 1:58 am

Post a whole recent night, including the flow limitation graph.. let's see what things look like.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

Re: strange flow rate waveform question

Post by yearning4zs » Thu May 16, 2019 2:40 am

palerider wrote:
Thu May 16, 2019 1:58 am
Post a whole recent night, including the flow limitation graph.. let's see what things look like.
Ok, here is one night ago.

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable
Attachments
whole night.jpg
whole night.jpg (481.16 KiB) Viewed 5055 times

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: strange flow rate waveform question

Post by palerider » Thu May 16, 2019 2:43 am

I'd raise the min pressure to 6.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

Re: strange flow rate waveform question

Post by yearning4zs » Thu May 16, 2019 2:56 am

palerider wrote:
Thu May 16, 2019 2:43 am
I'd raise the min pressure to 6.
Great, thanks for the advice! Can I humbly ask why the min pressure should be raised? Someone else also suggested that. Just would like to learn a bit and understand the logic. Thanks.

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable
Last edited by yearning4zs on Thu May 16, 2019 3:18 am, edited 1 time in total.

ajack
Posts: 977
Joined: Thu Mar 02, 2017 2:54 am
Location: australia

Re: strange flow rate waveform question

Post by ajack » Thu May 16, 2019 3:10 am

I'd try min 7 for one night and see is pressure helps.

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 ST-A iVAPS and adapt ASV

yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

Re: strange flow rate waveform question

Post by yearning4zs » Thu May 16, 2019 3:20 am

ajack wrote:
Thu May 16, 2019 3:10 am
I'd try min 7 for one night and see is pressure helps.
Is the idea of raising pressure to bring down the flow limitations? Would my amount of flow limitations be considered a lot? Just curious, because 7 seems to be pretty close to my 95%

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable

User avatar
Pugsy
Posts: 64983
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: strange flow rate waveform question

Post by Pugsy » Thu May 16, 2019 6:21 am

The amount of flow limitations shown isn't horrible but here's one of mine from some time back. Pretty boring.
You do have some activity showing.
More minimum pressure helps stent or hold the airway open better so as to help prevent those FLs from happening as long as the FLs aren't nasal congestion related. More pressure won't usually help nasal mucosa related FLs....you have to do something else for those.

Image

or here is someone else's report I had handy.

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

Re: strange flow rate waveform question

Post by yearning4zs » Thu May 16, 2019 7:49 am

Ok, good to know, thanks so much~~

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable