extremely low tidal volume

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

extremely low tidal volume

Post by yearning4zs » Tue May 14, 2019 9:43 pm

I posted yesterday about my respiratory rate looking too high. Thanks everyone for the answers, the machine was definitely overestimating my respiratory rate. However, this led me to find another issue. In the areas of artificially high respiratory rate, my tidal volume dips down below 100, and minute ventilation drops below 5, sometimes even to 3.7. Does anyone understand what's going on here???

I previously was diagnosed with epiglottic collapse and had surgery to partially excise the epiglottis. I'm starting to wonder if this is still a manifestation of an epiglottis problem. The strange thing is, the machine does not pick up any flow limit during these times...

Please see the below graphs.

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable
Attachments
tidal volume 3.jpg
tidal volume 3.jpg (524.48 KiB) Viewed 2450 times
tidal volume 2.jpg
tidal volume 2.jpg (422.2 KiB) Viewed 2450 times
tidal volume.jpg
tidal volume.jpg (409.27 KiB) Viewed 2450 times

Snoregone Conclusion
Posts: 210
Joined: Thu Jun 07, 2018 10:28 pm
Location: Issaquah, WA

Re: extremely low tidal volume

Post by Snoregone Conclusion » Wed May 15, 2019 2:05 am

Minute vent is tidal volume multiplied by respiration rate, and as long as you get enough minute vent, you won’t be starving for air.

You need to get your leaks down, from what’s visible, but otherwise, doesn’t seem crazy.

_________________
Machine
Additional Comments: 3B Elara FFM
Sleep, sleep monster, sleep!

yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

Re: extremely low tidal volume

Post by yearning4zs » Wed May 15, 2019 2:32 am

I guess was concerned that minute vent looked a little low too. This is not considered good in terms of overall leak rate? (see graph) resmed gave me a smiley face!

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable
Attachments
tidal volume 4.png
tidal volume 4.png (115.18 KiB) Viewed 2421 times

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: extremely low tidal volume

Post by Jas_williams » Wed May 15, 2019 4:11 am

yearning4zs wrote:
Wed May 15, 2019 2:32 am
I guess was concerned that minute vent looked a little low too. This is not considered good in terms of overall leak rate? (see graph) resmed gave me a smiley face!
Your leak rate is fine, anything less than 24l/m is fine

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

User avatar
LSAT
Posts: 13412
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: extremely low tidal volume

Post by LSAT » Wed May 15, 2019 5:52 am

The machine shown on Sleepyhead does not correspond to the machine in the profile.

yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

Re: extremely low tidal volume

Post by yearning4zs » Wed May 15, 2019 6:21 am

LSAT wrote:
Wed May 15, 2019 5:52 am
The machine shown on Sleepyhead does not correspond to the machine in the profile.
sorry, I just updated it. I've been using for her for a few weeks now. Was originally renting and then just recently bought one. The newest one I've been using for 3-4 days. The high artificial respiration rate started then.

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable

ajack
Posts: 977
Joined: Thu Mar 02, 2017 2:54 am
Location: australia

Re: extremely low tidal volume

Post by ajack » Wed May 15, 2019 6:22 am

You may be mouth breathing a bit with a nasal mask? Keep and eye on it and I would fix leaks where I can, if for no other reason than they annoy me and arouse me from sleep. I would turn the ramp off, have the start and min pressure on 6, max 10. It's not taking much pressure to resolve any apnea.

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 ST-A iVAPS and adapt ASV

yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

Re: extremely low tidal volume

Post by yearning4zs » Wed May 15, 2019 7:19 am

ajack wrote:
Wed May 15, 2019 6:22 am
You may be mouth breathing a bit with a nasal mask? Keep and eye on it and I would fix leaks where I can, if for no other reason than they annoy me and arouse me from sleep. I would turn the ramp off, have the start and min pressure on 6, max 10. It's not taking much pressure to resolve any apnea.
I tape my mouth, so probably no mouth breathing (doesn't seem to be leaking). But definitely leaks maybe from the mask. I could tighten it up a bit I think. The odd thing to me is there is basically no apnea recorded and even no flow limitations sometimes despite the really low tidal volume and minute ventilation. That's also why I'd be reluctant to increase the min-max pressure. My pressure never gets up to my max these days.

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable

ajack
Posts: 977
Joined: Thu Mar 02, 2017 2:54 am
Location: australia

Re: extremely low tidal volume

Post by ajack » Wed May 15, 2019 8:27 am

You may find, 6 isn't considered anything more than a starting pressure with the FFM, so it vents properly. You actually might find it more comfortable, as the EPR will be 2cm lower and 4.

See how the leaks go.

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 ST-A iVAPS and adapt ASV

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: extremely low tidal volume

Post by Jay Aitchsee » Wed May 15, 2019 8:41 am

Tidal Volume is the volume of each breath. Your erroneous respiration rate is high because each breath is being counted as more than one smaller breath. These erroneous smaller breaths each have less volume than the correct whole breath; therefore, the reported Tidal Volume for each breath is smaller.
In other words, the Tidal Volume reported during periods of erroneously high Respiration Rates is based on much smaller breaths then you are actually taking.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

Re: extremely low tidal volume

Post by yearning4zs » Wed May 15, 2019 8:42 pm

Jay Aitchsee wrote:
Wed May 15, 2019 8:41 am
Tidal Volume is the volume of each breath. Your erroneous respiration rate is high because each breath is being counted as more than one smaller breath. These erroneous smaller breaths each have less volume than the correct whole breath; therefore, the reported Tidal Volume for each breath is smaller.
In other words, the Tidal Volume reported during periods of erroneously high Respiration Rates is based on much smaller breaths then you are actually taking.
very interesting and well noted!

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable