AHI/frequent waking

Post by midnightprowler1960 » Sat May 04, 2019 7:26 am

My clinic set my machine after the first month at 8-12 cmH2O. Since then I have had maybe a couple dozen AHI readings under 10. Most are in the teens. Also I wake up after most dreams, probably 4-6 times a night. Sometimes a little less. Hav a appt at the clinic the 30th. Wonder if I should bump it up?

_________________
MachineHumidifier
Additional Comments: Airfit F30 mask

User avatar
LSAT
Posts: 13404
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: AHI/frequent waking

Post by LSAT » Sat May 04, 2019 7:39 am

Hard to tell what to set the pressures to without data. You could start with raising the max pressure to 20...that can only help AHI. Don't know about wake-ups without data.

User avatar
zonker
Posts: 11451
Joined: Fri Jun 19, 2015 4:36 pm

Re: AHI/frequent waking

Post by zonker » Sat May 04, 2019 10:54 am

midnightprowler1960 wrote:
Sat May 04, 2019 7:26 am
My clinic set my machine after the first month at 8-12 cmH2O. Since then I have had maybe a couple dozen AHI readings under 10. Most are in the teens. Also I wake up after most dreams, probably 4-6 times a night. Sometimes a little less. Hav a appt at the clinic the 30th. Wonder if I should bump it up?
are you using software to track your data? that is, either sleepyhead or oscar? please see my sig for download of sleepyhead. someone else will have to supply the oscar link.

these are real software programs, not data apps. what you can do is screen grab of the charts it produces and then post them here for perusal. several people here can make suggestions on how to get your settings optimized.

without the data, it's guesswork.

oh! also, see-

viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
midnightprowler1960
Posts: 14
Joined: Thu May 02, 2019 6:20 am
Location: Cedar, MN

Re: AHI/frequent waking

Post by midnightprowler1960 » Sat May 04, 2019 2:13 pm

Not sure how I would use any software with my machine. I'm not the most tech savvy

_________________
MachineHumidifier
Additional Comments: Airfit F30 mask

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AHI/frequent waking

Post by palerider » Sat May 04, 2019 2:36 pm

midnightprowler1960 wrote:
Sat May 04, 2019 2:13 pm
Not sure how I would use any software with my machine. I'm not the most tech savvy
Instead of insulting an entire forum, you could try READING and LEARNING A much more productive use of your time.

Start here:

Sticky: Newbies PLEASE READ BEFORE POSTING

Very first post on the front page.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
midnightprowler1960
Posts: 14
Joined: Thu May 02, 2019 6:20 am
Location: Cedar, MN

Re: AHI/frequent waking

Post by midnightprowler1960 » Sat May 04, 2019 2:47 pm

Maybe YOU should read it bully boy.

_________________
MachineHumidifier
Additional Comments: Airfit F30 mask

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: AHI/frequent waking

Post by Wulfman... » Sat May 04, 2019 2:52 pm

midnightprowler1960 wrote:
Sat May 04, 2019 7:26 am
My clinic set my machine after the first month at 8-12 cmH2O. Since then I have had maybe a couple dozen AHI readings under 10. Most are in the teens. Also I wake up after most dreams, probably 4-6 times a night. Sometimes a little less. Hav a appt at the clinic the 30th. Wonder if I should bump it up?
Maybe you're sensitive to pressure changes. (some of us on the forum are) You could try a fixed/straight pressure for awhile and see how that goes. Just because you HAVE an APAP doesn't necessarily mean you have to use it in a range of pressures.
Sounds like you're getting bumped out of the deep sleep stages and into lighter ones. (REM is not a deep sleep stage, but we can dream in any sleep stage)

Are you using software to monitor your therapy or just using the information from the LCD?
Never mind.......I just read your previous post.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: AHI/frequent waking

Post by Miss Emerita » Sat May 04, 2019 2:58 pm

Palerider, I think midnightprowler was just worrying about whether she or he had the tech chops needed to use Sleepyhead.

Midnightprowler, if you have a laptop, give it a try. I’m not very tech-savvy myself, but I managed to use SH (and now Oscar), and I rely on the information heavily.

Plan B would be to ask a tech-savvy friend or relative for help. Once you got the software downloaded and saw how to import your nightly data into it, I think you’d find it easy to continue on your own.

Waking up after dreams is pretty normal, but an AHI that is in the teens isn’t. If you can’t deal with SH, can you get an earlier appointment at the clinic? Your settings just aren’t working for you, and someone needs to look at what kinds of events you’re having and make changes in light of that information.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
Lee Rilea
Posts: 50
Joined: Fri Apr 26, 2019 3:10 am

Re: AHI/frequent waking

Post by Lee Rilea » Sat May 04, 2019 3:45 pm

Wulfman... wrote:
Sat May 04, 2019 2:52 pm
midnightprowler1960 wrote:
Sat May 04, 2019 7:26 am
My clinic set my machine after the first month at 8-12 cmH2O. Since then I have had maybe a couple dozen AHI readings under 10. Most are in the teens. Also I wake up after most dreams, probably 4-6 times a night. Sometimes a little less. Hav a appt at the clinic the 30th. Wonder if I should bump it up?
Maybe you're sensitive to pressure changes. (some of us on the forum are) You could try a fixed/straight pressure for awhile and see how that goes. Just because you HAVE an APAP doesn't necessarily mean you have to use it in a range of pressures.
Sounds like you're getting bumped out of the deep sleep stages and into lighter ones. (REM is not a deep sleep stage, but we can dream in any sleep stage)

Are you using software to monitor your therapy or just using the information from the LCD?
Never mind.......I just read your previous post.



Den

.
When I started last December I was at a steady 8. I improved from my sleep study, but not enough. They set it up for the variable 8-12. I may try to go back in early. I have had the frequent waking since the beginning.

_________________
MachineHumidifier
Additional Comments: I use a Airfit F30 mask, no idea what software. I have dreammapper app on my phone
1 Corinthians: 15:51-54
1 Thessalonians 4:13-18

User avatar
Lee Rilea
Posts: 50
Joined: Fri Apr 26, 2019 3:10 am

Re: AHI/frequent waking

Post by Lee Rilea » Sat May 04, 2019 3:49 pm

Miss Emerita wrote:
Sat May 04, 2019 2:58 pm
Palerider, I think midnightprowler was just worrying about whether she or he had the tech chops needed to use Sleepyhead.

Midnightprowler, if you have a laptop, give it a try. I’m not very tech-savvy myself, but I managed to use SH (and now Oscar), and I rely on the information heavily.

Plan B would be to ask a tech-savvy friend or relative for help. Once you got the software downloaded and saw how to import your nightly data into it, I think you’d find it easy to continue on your own.

Waking up after dreams is pretty normal, but an AHI that is in the teens isn’t. If you can’t deal with SH, can you get an earlier appointment at the clinic? Your settings just aren’t working for you, and someone needs to look at what kinds of events you’re having and make changes in light of that information.
I'm not sure my machine is compatible with software.

_________________
MachineHumidifier
Additional Comments: I use a Airfit F30 mask, no idea what software. I have dreammapper app on my phone
1 Corinthians: 15:51-54
1 Thessalonians 4:13-18

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: AHI/frequent waking

Post by Miss Emerita » Sat May 04, 2019 5:43 pm

Could you check whether you have the right machine in your profile? I followed the link for the one you have there, and it is fixed pressure only, which doesn’t fit with your reference to a pressure range. For what it’s worth, the one you have listed comes with an SD card, which suggests you could use SH or Oscar.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
Lee Rilea
Posts: 50
Joined: Fri Apr 26, 2019 3:10 am

Re: AHI/frequent waking

Post by Lee Rilea » Sat May 04, 2019 5:49 pm

Miss Emerita wrote:
Sat May 04, 2019 5:43 pm
Could you check whether you have the right machine in your profile? I followed the link for the one you have there, and it is fixed pressure only, which doesn’t fit with your reference to a pressure range. For what it’s worth, the one you have listed comes with an SD card, which suggests you could use SH or Oscar.
Yes it is correct, the clinic made some adjustment and now it starts ramp up at 4 for 10 minutes, then goes to 8 and as high as 12. It has a sd card.

_________________
MachineHumidifier
Additional Comments: I use a Airfit F30 mask, no idea what software. I have dreammapper app on my phone
1 Corinthians: 15:51-54
1 Thessalonians 4:13-18

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AHI/frequent waking

Post by palerider » Sat May 04, 2019 5:51 pm

Lee Rilea wrote:
Sat May 04, 2019 5:49 pm
Miss Emerita wrote:
Sat May 04, 2019 5:43 pm
Could you check whether you have the right machine in your profile? I followed the link for the one you have there, and it is fixed pressure only, which doesn’t fit with your reference to a pressure range. For what it’s worth, the one you have listed comes with an SD card, which suggests you could use SH or Oscar.
Yes it is correct, the clinic made some adjustment and now it starts ramp up at 4 for 10 minutes, then goes to 8 and as high as 12. It has a sd card.
The machine in your profile *does not change pressures*, so either it is the wrong machine listed, or...

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Lee Rilea
Posts: 50
Joined: Fri Apr 26, 2019 3:10 am

Re: AHI/frequent waking

Post by Lee Rilea » Sat May 04, 2019 5:52 pm

It's the one listed hoser. Just stay off my threads. I bought it, I was there when they set it up you were not. Besides I thought you were ignoring me.

_________________
MachineHumidifier
Additional Comments: I use a Airfit F30 mask, no idea what software. I have dreammapper app on my phone
1 Corinthians: 15:51-54
1 Thessalonians 4:13-18

User avatar
zonker
Posts: 11451
Joined: Fri Jun 19, 2015 4:36 pm

Re: AHI/frequent waking

Post by zonker » Sat May 04, 2019 6:05 pm

palerider wrote:
Sat May 04, 2019 5:51 pm
Lee Rilea wrote:
Sat May 04, 2019 5:49 pm
Miss Emerita wrote:
Sat May 04, 2019 5:43 pm
Could you check whether you have the right machine in your profile? I followed the link for the one you have there, and it is fixed pressure only, which doesn’t fit with your reference to a pressure range. For what it’s worth, the one you have listed comes with an SD card, which suggests you could use SH or Oscar.
Yes it is correct, the clinic made some adjustment and now it starts ramp up at 4 for 10 minutes, then goes to 8 and as high as 12. It has a sd card.
The machine in your profile *does not change pressures*, so either it is the wrong machine listed, or...
you have nothing to worry about; knightsleeper is helping via pm.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win