I am getting an error message in the window of my Airsense 10 Auto. It says "Attention: The electrical connection at the rear of your device is not fully inserted".
I have checked it and unplugged and re-plugged several times. I am sure it is fully inserted but I cannot find a way to make the message go away. The machine works as normal except the message is always there and the screen and home buttons are fully lit (distracting at night). Does anyone have a clue for how I can clear this message? All suggestions are appreciated.
Airsense 10 Auto with Error Message
- katestyles
- Posts: 610
- Joined: Sun Jan 06, 2019 9:08 am
Re: Airsense 10 Auto with Error Message
I have no idea why this should be happening - You can call ResMed, or you can go to your DME, but it sounds like a fault to me.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Back up mask - anything in the drawer |
-
prodigyplace
- Posts: 1795
- Joined: Mon Mar 27, 2017 5:45 pm
- Location: Central Virginia
Re: Airsense 10 Auto with Error Message
If unplugging & re-plugging does not help, the issue could be either with the power brick or the machine.
One other thing you could try. Unplug from the wall for 10 minutes & then reconnect. Perhaps that can reset something that is "confused"
One other thing you could try. Unplug from the wall for 10 minutes & then reconnect. Perhaps that can reset something that is "confused"
_________________
| Machine: AirSense 11 Autoset |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software |
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.
- Dog Slobber
- Posts: 4287
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Airsense 10 Auto with Error Message
This occurs when the centre pin (3.3 volt) isn't making [good] contact with the pin receptacle on the back of the device.
Some possible causes:
The pin and the outer shell should measure 3.3v, the inner shell and outer shell should measure 24v.
If everything checks out ok, it's likely something wrong with the internal connectors on the device. If not the power supply needs to be replaced.
Some possible causes:
- Not inserted correctly
- Water damage
- Missing or bent middle pin
The pin and the outer shell should measure 3.3v, the inner shell and outer shell should measure 24v.
If everything checks out ok, it's likely something wrong with the internal connectors on the device. If not the power supply needs to be replaced.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Last edited by Dog Slobber on Tue Apr 30, 2019 9:19 am, edited 2 times in total.
Battery Backup: EcoFlow Delta 2
Re: Airsense 10 Auto with Error Message
Try calling https://acbio.com/ ..they may have an answer
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
Re: Airsense 10 Auto with Error Message
did you try reattach the cable that attaches to the outboard power supply?
-tino
-tino
_________________
| Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Autoset Pressure 9-15.6, EPR 1, no ramp |
Tino
Re: Airsense 10 Auto with Error Message
Thanks to all who replied. The center pin looks ok and I think the power supply is good because everything operates as normal except for the error message and the "window" staying lit all the time. I tried connecting my 12V power supply and got the same error message so I guess the problem is in the device board. Guess I'll live with it for a while and see if anything changes.
Re: Airsense 10 Auto with Error Message
Resmeds run on 24 volts.