Used petroleum jelly to better my mask seal..but then realized petroleum will break down my mask. Any alternatives?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
zoomzoom
Posts: 140
Joined: Tue Mar 05, 2019 7:39 pm

Used petroleum jelly to better my mask seal..but then realized petroleum will break down my mask. Any alternatives?

Post by zoomzoom » Thu Apr 18, 2019 12:41 am

I was getting annoyed at side sleeping and having leaks on my Brevida. I hate the sound of leaks and I don't like the air waking me up. I like it really quiet when I sleep, but I'm even able to hear the otherwise quiet Airsense 10.

I did some research and came across Ayr gel and something called CPAP Chap Moisture Therapy Face Cream (https://www.amazon.com/CPAP-Chap-Moistu ... r=1-3&th=1).

All I want is the seal to be better so I don't have to constantly readjust for the mask to go silent.

Which of the above two items (or something else altogether) would be best for my purpose?

I don't get bloody noses or anything.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Used petroleum jelly to better my mask seal..but then realized petroleum will break down my mask. Any alternatives?

Post by palerider » Thu Apr 18, 2019 12:50 am

zoomzoom wrote:
Thu Apr 18, 2019 12:41 am
I was getting annoyed at side sleeping and having leaks on my Brevida. I hate the sound of leaks and I don't like the air waking me up. I like it really quiet when I sleep, but I'm even able to hear the otherwise quiet Airsense 10.
Mask liners were a game changer for me when I was using a FFM.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.


User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: Used petroleum jelly to better my mask seal..but then realized petroleum will break down my mask. Any alternatives?

Post by Jay Aitchsee » Thu Apr 18, 2019 6:21 am

I'm not sure of your application, but if you want a substitute for petroleum jelly than won't harm you or your mask, you could try a lanolin cream. Lansinoh is favorite brand here. It's a nipple cream for nursing mothers. A little goes a very long way. It can be found online or in the baby section of stores.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

SnoringInOregon
Posts: 60
Joined: Sun Nov 26, 2017 6:47 pm
Location: Oregon

Re: Used petroleum jelly to better my mask seal..but then realized petroleum will break down my mask. Any alternatives?

Post by SnoringInOregon » Fri Apr 19, 2019 10:09 pm

You definitely don't want to use petroleum jelly in a situation where you could be breathing it in. Here's what could happen if any vaporizes and winds up in your lungs: https://www.mayoclinic.org/petroleum-je ... q-20057784

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Used petroleum jelly to better my mask seal..but then realized petroleum will break down my mask. Any alternatives?

Post by Wulfman... » Fri Apr 19, 2019 10:28 pm

zoomzoom wrote:
Thu Apr 18, 2019 12:41 am
I was getting annoyed at side sleeping and having leaks on my Brevida. I hate the sound of leaks and I don't like the air waking me up. I like it really quiet when I sleep, but I'm even able to hear the otherwise quiet Airsense 10.

I did some research and came across Ayr gel and something called CPAP Chap Moisture Therapy Face Cream (https://www.amazon.com/CPAP-Chap-Moistu ... r=1-3&th=1).

All I want is the seal to be better so I don't have to constantly readjust for the mask to go silent.

Which of the above two items (or something else altogether) would be best for my purpose?

I don't get bloody noses or anything.
Maybe your real problem is your bed pillow.

search.php?keywords=%22bed+pillow%22


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Used petroleum jelly to better my mask seal..but then realized petroleum will break down my mask. Any alternatives?

Post by chunkyfrog » Fri Apr 19, 2019 10:49 pm

I vote for the mask liner.
For immediate use, tear one out of a Viva paper towel.
Start big, trim down until it works.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

zoomzoom
Posts: 140
Joined: Tue Mar 05, 2019 7:39 pm

Re: Used petroleum jelly to better my mask seal..but then realized petroleum will break down my mask. Any alternatives?

Post by zoomzoom » Sun Apr 21, 2019 3:19 am

Wulfman... wrote:
Fri Apr 19, 2019 10:28 pm
Maybe your real problem is your bed pillow.

search.php?keywords=%22bed+pillow%22


Den

.
You might very well be right. I am using a MyPillow. I like it because I can wash it every other week. I am willing to try something like this (https://www.apriadirect.com/contour-cpa ... leep-apnea) -- linking to Apria, I hope that's OK! There is also a buckwheat pillow for the same price: https://www.apriadirect.com/snugglehose ... illow-case . The cpap Contour pillow has little cutouts for the hose so perhaps it won't get snagged as I'm tossing or turning, so that might help keep everything in place.

chunkyfrog wrote:
Fri Apr 19, 2019 10:49 pm
I vote for the mask liner.
For immediate use, tear one out of a Viva paper towel.
Start big, trim down until it works.
What exactly does the liner do? It is my understanding it just a comfort thing for your cheeks so the straps don't bother you. Am I mistaken? Is it supposed to hold the mask in place ("sealed") or something?
SnoringInOregon wrote:
Fri Apr 19, 2019 10:09 pm
You definitely don't want to use petroleum jelly in a situation where you could be breathing it in. Here's what could happen if any vaporizes and winds up in your lungs: https://www.mayoclinic.org/petroleum-je ... q-20057784
Wow, glad I only used it once. Jeez, that's crazy!
Jay Aitchsee wrote:
Thu Apr 18, 2019 6:21 am
I'm not sure of your application, but if you want a substitute for petroleum jelly than won't harm you or your mask, you could try a lanolin cream. Lansinoh is favorite brand here. It's a nipple cream for nursing mothers. A little goes a very long way. It can be found online or in the baby section of stores.
I have read this recommendation a few times. Do you think they'd work better than the two products I mentioned in the OP. The Ayr gel I've also seen recommended and the Chap cream was just something I found on amazon.
palerider wrote:
Thu Apr 18, 2019 12:50 am
Mask liners were a game changer for me when I was using a FFM.
They shouldn't have any impact for sealing for the brevida, though, right? Maybe I am mistaking what mask liners actually do (I think they just provide comfort to the cheek area, right?)

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Used petroleum jelly to better my mask seal..but then realized petroleum will break down my mask. Any alternatives?

Post by palerider » Sun Apr 21, 2019 3:34 am

zoomzoom wrote:
Sun Apr 21, 2019 3:19 am
They shouldn't have any impact for sealing for the brevida, though, right? Maybe I am mistaking what mask liners actually do (I think they just provide comfort to the cheek area, right?)
No, sorry, not on a nasal pillow.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

zoomzoom
Posts: 140
Joined: Tue Mar 05, 2019 7:39 pm

Re: Used petroleum jelly to better my mask seal..but then realized petroleum will break down my mask. Any alternatives?

Post by zoomzoom » Sun Apr 21, 2019 4:09 am

palerider wrote:
Sun Apr 21, 2019 3:34 am
zoomzoom wrote:
Sun Apr 21, 2019 3:19 am
They shouldn't have any impact for sealing for the brevida, though, right? Maybe I am mistaking what mask liners actually do (I think they just provide comfort to the cheek area, right?)
No, sorry, not on a nasal pillow.
Ah shucks

User avatar
katestyles
Posts: 610
Joined: Sun Jan 06, 2019 9:08 am

Re: Used petroleum jelly to better my mask seal..but then realized petroleum will break down my mask. Any alternatives?

Post by katestyles » Sun Apr 21, 2019 7:29 am

zoomzoom wrote:
Sun Apr 21, 2019 4:09 am
palerider wrote:
Sun Apr 21, 2019 3:34 am
zoomzoom wrote:
Sun Apr 21, 2019 3:19 am
They shouldn't have any impact for sealing for the brevida, though, right? Maybe I am mistaking what mask liners actually do (I think they just provide comfort to the cheek area, right?)
No, sorry, not on a nasal pillow.
Ah shucks
There are some ..."things"...that can be wrapped round nasal interfaces to help with comfort/seal - I'm not familiar with your mask, but you might be able to construct something out of T-shirt fabric that helps. The trick is to put the fabric like a gasket between the silicone and skin.