Need Some Help with SleepyHead
Need Some Help with SleepyHead
Hey Everyone,
I've been using my CPAP machine for about 10 days now (since March 15th), and while I keep transferring all of my data over to SleepyHead I honestly have no idea what any of it means.
A little bit about me:
* I do take a sleep medication, Zopiclone (5mg) every night before bed as well as 20mg of Cipralex (Lexapro in the USA), and 150mg of Welbutrin
* I did a home sleep study and the specialist determined I had sleep apnea, but my doctor disagreed and did not give me the results or recommend investigating further. I then saw an ENT who again agreed that I had sleep apnea and simply told me to get a CPAP machine. I was never given a recommended pressure setting or any real information, so I am trying to figure it out as I go.
* I am using the AireSense 10 AutoSet for Her and the F30 mask
I am including two different nights of data to give a better overview of what I'm experiencing in the hopes of getting some help with what my pressure should be at, if I'm doing things correctly/incorrectly, etc. The first image is from March 23rd (8 days after starting) and the second image is from March 19th (4 days after starting).
Thank you to everyone who responds!
I've been using my CPAP machine for about 10 days now (since March 15th), and while I keep transferring all of my data over to SleepyHead I honestly have no idea what any of it means.
A little bit about me:
* I do take a sleep medication, Zopiclone (5mg) every night before bed as well as 20mg of Cipralex (Lexapro in the USA), and 150mg of Welbutrin
* I did a home sleep study and the specialist determined I had sleep apnea, but my doctor disagreed and did not give me the results or recommend investigating further. I then saw an ENT who again agreed that I had sleep apnea and simply told me to get a CPAP machine. I was never given a recommended pressure setting or any real information, so I am trying to figure it out as I go.
* I am using the AireSense 10 AutoSet for Her and the F30 mask
I am including two different nights of data to give a better overview of what I'm experiencing in the hopes of getting some help with what my pressure should be at, if I'm doing things correctly/incorrectly, etc. The first image is from March 23rd (8 days after starting) and the second image is from March 19th (4 days after starting).
Thank you to everyone who responds!
- Attachments
-
- March 23rd
- Saturday March 23.png (423.81 KiB) Viewed 5596 times
-
- March 19th
- Tuesday March 19.png (418.25 KiB) Viewed 5596 times
Re: Need Some Help with SleepyHead
Welcome to the forum. 1st off you are doing great! I think that you need to kill the obstructive events first. You do that with more minimum pressure so you can prevent them from ever happening. If those were my charts I would bump the minimum pressure up to 10 and see how the machine performed. You picked a good machine by the way. If you don't want to make that big of a minimum pressure increase, you can increase in steps. How are you doing with the pressures now? On the 2 charts that you posted you seem to be using the machine for 6-7 hours, which is great (especially for just starting out). I predict that your final minimum pressure will be closer to 12 than 10 in the end.
_________________
| Mask: DreamWear Gel Nasal Pillow CPAP Mask with Headgear - Fit Pack (All Nasal Pillows with Medium Frame) |
| Additional Comments: Settings 11-20, EPR 2. Dreamstation Full Auto for back up |
Re: Need Some Help with SleepyHead
watch all the videos here first.
http://freecpapadvice.com/sleepyhead-free-software
http://freecpapadvice.com/sleepyhead-free-software
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Need Some Help with SleepyHead
Right now my pressure is 5.6 which I know is quite low and it's no issue. In terms of higher pressures, for the first week, I would wake up any time my machine bumped me up to 10+ but now I seem to sleep through until I wake up for work (for the most part).Thumper68 wrote: ↑Tue Mar 26, 2019 6:02 amWelcome to the forum. 1st off you are doing great! I think that you need to kill the obstructive events first. You do that with more minimum pressure so you can prevent them from ever happening. If those were my charts I would bump the minimum pressure up to 10 and see how the machine performed. You picked a good machine by the way. If you don't want to make that big of a minimum pressure increase, you can increase in steps. How are you doing with the pressures now? On the 2 charts that you posted you seem to be using the machine for 6-7 hours, which is great (especially for just starting out). I predict that your final minimum pressure will be closer to 12 than 10 in the end.
With increasing the minimum should I set up that EPR feature? Or just try without it?
Thanks!
Re: Need Some Help with SleepyHead
That you for the link - will watch through them today. I've been doing a lot of reading, both on here and off, but the technical information currently seems to go right over my head! Very motivated to learn, however, so I appreciate the resources.Pugsy wrote: ↑Tue Mar 26, 2019 6:15 amwatch all the videos here first.
http://freecpapadvice.com/sleepyhead-free-software
Re: Need Some Help with SleepyHead
EPR is a personal choice. It is basically a comfort setting. You will have to try it with and without to know what you prefer. I do use a EPR of 2. My minimum pressure is currently 12.4. EPR drops my exhale pressure to 10.4 on exhale.
_________________
| Mask: DreamWear Gel Nasal Pillow CPAP Mask with Headgear - Fit Pack (All Nasal Pillows with Medium Frame) |
| Additional Comments: Settings 11-20, EPR 2. Dreamstation Full Auto for back up |
Re: Need Some Help with SleepyHead
I took the advice given and upped my starting pressure to 7, which seems to have worked like a charm and left me with my lowest AHI since starting (0.19).
With that said... I woke up this morning because of terrible stomach pains and very loud, large burps. After doing some reading online I have a hunch that it could be aerophagia. My sleep was a lot more restful, but it was also a lot shorter.
So my question is, will I just have to make a choice between being quite gassy and having a low AHI, or, having a higher AHI but no stomach pains? Or, are there things I can do to combat the gas and pain?
Thank you!!
With that said... I woke up this morning because of terrible stomach pains and very loud, large burps. After doing some reading online I have a hunch that it could be aerophagia. My sleep was a lot more restful, but it was also a lot shorter.
So my question is, will I just have to make a choice between being quite gassy and having a low AHI, or, having a higher AHI but no stomach pains? Or, are there things I can do to combat the gas and pain?
Thank you!!
Re: Need Some Help with SleepyHead
Did you add EPR when you increased the pressure minimum to 7cm? If not that would be the first thing I did. If you did...increase it if you can.
wiki/index.php/Aerophagia
wiki/index.php/Aerophagia
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Need Some Help with SleepyHead
Hi Pugsy - No, I didn't touch the EPR setting because I wasn't too sure what I was doing and it had two options, Ramp Only and Full-Time.Pugsy wrote: ↑Wed Mar 27, 2019 8:41 amDid you add EPR when you increased the pressure minimum to 7cm? If not that would be the first thing I did. If you did...increase it if you can.
wiki/index.php/Aerophagia
The way I have my settings right now is that I am using a starting pressure of 5.6 with Auto-Ramp, and then a minimum pressure of 7. So would it be smarter to turn off my ramp setting and just start at 7 with the EPR on full-time?
Re: Need Some Help with SleepyHead
Use the ramp if you need it to ease up to the 7 cm starting minimum.
Set EPR to full time (you can still use ramp if you wish) and set it to 2 for now and see what happens in terms of aerophagia issues.
The idea is hopefully the lower pressure on exhale will stop the air from entering the stomach so much or so often. Less consistent pressure on the LES lower esophageal sphincter.
Set EPR to full time (you can still use ramp if you wish) and set it to 2 for now and see what happens in terms of aerophagia issues.
The idea is hopefully the lower pressure on exhale will stop the air from entering the stomach so much or so often. Less consistent pressure on the LES lower esophageal sphincter.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Need Some Help with SleepyHead
Perfect - I will try this tonight and see if I wake up less... bubbly...Pugsy wrote: ↑Wed Mar 27, 2019 8:59 amUse the ramp if you need it to ease up to the 7 cm starting minimum.
Set EPR to full time (you can still use ramp if you wish) and set it to 2 for now and see what happens in terms of aerophagia issues.
The idea is hopefully the lower pressure on exhale will stop the air from entering the stomach so much or so often. Less consistent pressure on the LES lower esophageal sphincter.
Thank you for taking the time to help!!
Re: Need Some Help with SleepyHead
Hey Everyone!
So I gave it about a week after moving my minimum pressure up to 7, but now it seems my AHI has gone up along with it. The first few days were great, but the last few not so much.
I'm attaching a sleepyhead screenshot from last night hoping one of you might be able to decipher it for me and pinpoint where I'm having any issues? I know that since I am under 5 it is considered good and therapeutic - but it is more me being worried that my settings aren't correct and that I could be doing something better.
Thank you for taking the time to read this and help, it is greatly appreciated.
So I gave it about a week after moving my minimum pressure up to 7, but now it seems my AHI has gone up along with it. The first few days were great, but the last few not so much.
I'm attaching a sleepyhead screenshot from last night hoping one of you might be able to decipher it for me and pinpoint where I'm having any issues? I know that since I am under 5 it is considered good and therapeutic - but it is more me being worried that my settings aren't correct and that I could be doing something better.
Thank you for taking the time to read this and help, it is greatly appreciated.
- Attachments
-
- March 30th.png (406.38 KiB) Viewed 5476 times
Re: Need Some Help with SleepyHead
How is your sleep quality? Waking often or sleeping soundly?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Need Some Help with SleepyHead
My sleep quality is significantly better than before I started CPAP. I used to wake up at least 10+ times a night, and now I wake up maybe once. Even last night, with the higher AHI, I still feel quite rested this morning.
Re: Need Some Help with SleepyHead
The bulk of your AHI is clear airway/central related. Can’t do much about those with this machine and you really aren’t having enough of them to worry about.
I wouldn’t change anything. Give yourself some time and just get used to using Cpap. A few centrals are normal anyway.
The main thing is you are sleeping good and feeling good. I suspect with time the central numbers will drop on their own.....and since we can’t do anything about them anyway with this machine....give it some time.
Actually for a newbie report......pretty darn good. Excellent job.
I wouldn’t change anything. Give yourself some time and just get used to using Cpap. A few centrals are normal anyway.
The main thing is you are sleeping good and feeling good. I suspect with time the central numbers will drop on their own.....and since we can’t do anything about them anyway with this machine....give it some time.
Actually for a newbie report......pretty darn good. Excellent job.