Using version 1.0.0 beta 2 of Sleephead.
Used to work fine, but now I'm not getting any new data after March 5th.
Any suggestions? Data IS getting to ResMed.
Sleepyhead problem
Sleepyhead problem
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Airsense 10 Autoset
AirTouch and AirFit F10 masks

AirTouch and AirFit F10 masks

- Dog Slobber
- Posts: 4236
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Sleepyhead problem
The data that's going to ResMed is summary and compliance not the detail data.
- Is the write-protect tab on the SD Card engaged
- Reformat the SD Card
- Replace the SD Card, anyone will do buy the most inexpensive you can get.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
- Dog Slobber
- Posts: 4236
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Sleepyhead problem
The AirSense 10 line is not impacted by the foreign file / reformatting card problem the S9 had, but it will not write to an SD Card that is write protected. Though, I haven't verified this for myself.KnightSleeper wrote: ↑Mon Mar 18, 2019 10:16 amGood points Dog Slobber, but the Resmed Airsense 10 Autoset is not impacted by a locked SD card.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Sleepyhead problem
After playing around last night, seems Sleepyhead had changed where it wanted to find the data.Dog Slobber wrote: ↑Mon Mar 18, 2019 7:03 amThe data that's going to ResMed is summary and compliance not the detail data.
- Is the write-protect tab on the SD Card engaged
- Reformat the SD Card
- Replace the SD Card, anyone will do buy the most inexpensive you can get.
Pointed it back at the card, and all is now well again.
Now at 10 days straight of AHI less than 2, and loving it!
Thanks all, for the help.