Resmed S9 APAP dying?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Arizona-Willie
Posts: 703
Joined: Sun Jul 02, 2006 2:27 pm
Location: Mesa AZ

Resmed S9 APAP dying?

Post by Arizona-Willie » Thu Feb 28, 2019 10:44 am

My S9 APAP machine is giving me trouble.
When I put the3 data card in it says it is reading the card and never finishes.
In order to get it to work I have to remove the card and then it will blow air for me.
I have tried a 2nd card with no luck.

It is 9 years old.

Is there a way to perhaps reset the machine or something that I might save it and not have to buy a new one out of pocket?

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleepyhead ver 1.0.0 Beta 2

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Resmed S9 APAP dying?

Post by Okie bipap » Thu Feb 28, 2019 11:14 am

I would get a can of electrical terminal cleaner, unplug the machine, spray the electrical contacts in the card slot, blow it out good with canned air, let it dry for several hours, and try again. If that fails, maybe it is time to look for a different machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed S9 APAP dying?

Post by palerider » Thu Feb 28, 2019 2:09 pm

Okie bipap wrote:
Thu Feb 28, 2019 11:14 am
I would get a can of electrical terminal cleaner, unplug the machine, spray the electrical contacts in the card slot, blow it out good with canned air, let it dry for several hours, and try again. If that fails, maybe it is time to look for a different machine.
blow it out first, cruft tends to accumulate in SD card slots.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11432
Joined: Fri Jun 19, 2015 4:36 pm

Re: Resmed S9 APAP dying?

Post by zonker » Thu Feb 28, 2019 5:27 pm

Okie bipap wrote:
Thu Feb 28, 2019 11:14 am
I would get a can of electrical terminal cleaner, unplug the machine, spray the electrical contacts in the card slot, blow it out good with canned air, let it dry for several hours, and try again. If that fails, maybe it is time to look for a different machine.
perhaps a dry eraser used on the contacts of the sd card?

(sorry, meant to reply to the op.)
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed S9 APAP dying?

Post by palerider » Thu Feb 28, 2019 8:38 pm

zonker wrote:
Thu Feb 28, 2019 5:27 pm
Okie bipap wrote:
Thu Feb 28, 2019 11:14 am
I would get a can of electrical terminal cleaner, unplug the machine, spray the electrical contacts in the card slot, blow it out good with canned air, let it dry for several hours, and try again. If that fails, maybe it is time to look for a different machine.
perhaps a dry eraser used on the contacts of the sd card?

(sorry, meant to reply to the op.)
I usually use a pencil eraser on electrical contacts, provides just enough abrasion to clean them, without any damage.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.