Am I not understanding sleepyhead or a bipap machine settings?
Hi All,
I'm trying to understand my settings on the Dreamstation AutoBipap on basic Fixed BIPAP mode. Sorry for not setting up all the graphs the right way but I just want you to notice the "Fixed Bi-Level" and EPAP 11, IPAP 18, then notice the pressure graph. I am not getting it. The IPAP line is static at 18 as I expect, but the EPAP is ramping up from 11 -> 13 (when 11 is selected as EPAP!). I checked all the settings and see nothing related to PS, Ramp, Flex is turned off, nothing obvious. Can anyone tell me what is going on???
I'm trying to understand my settings on the Dreamstation AutoBipap on basic Fixed BIPAP mode. Sorry for not setting up all the graphs the right way but I just want you to notice the "Fixed Bi-Level" and EPAP 11, IPAP 18, then notice the pressure graph. I am not getting it. The IPAP line is static at 18 as I expect, but the EPAP is ramping up from 11 -> 13 (when 11 is selected as EPAP!). I checked all the settings and see nothing related to PS, Ramp, Flex is turned off, nothing obvious. Can anyone tell me what is going on???
Re: Am I not understanding sleepyhead or a bipap machine settings?
That is weird. I wonder if it is a SleepyHead bug with this version.
I suggest you try version 1.1 and see what it shows.
https://www.apneaboard.com/sleepyhead/
I suggest you try version 1.1 and see what it shows.
https://www.apneaboard.com/sleepyhead/
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Am I not understanding sleepyhead or a bipap machine settings?
Also....go look in the machine's clinical setup menu for verification of settings.
Sometimes the Settings part of SH is buggy so always look at the machine itself also.
Sometimes the Settings part of SH is buggy so always look at the machine itself also.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
Sleepygal1
- Posts: 40
- Joined: Tue Nov 20, 2018 2:13 pm
Re: Am I not understanding sleepyhead or a bipap machine settings?
Huh, that's interesting, didn't know there was a new version. Same thing though. i actually set the EPAP pressure down to 9 to try to get it around where I want it at 11->11.5. I've been relying on this data for a while so I went from auto Bipap to using the statistics in SH to select 11 and 18. Now I'm not sure what to do! I'm assuming if its a bug, then it should be down around 9ish instead of 11 if its always off by 2? Hard to know.
I did look through all the settings on the clinical menu, that's what is so baffling to me. Does 9 and 18 sound like a safe BiPAP level or is the EPAP and IPAP too far apart for normal?
I did look through all the settings on the clinical menu, that's what is so baffling to me. Does 9 and 18 sound like a safe BiPAP level or is the EPAP and IPAP too far apart for normal?
Re: Am I not understanding sleepyhead or a bipap machine settings?
biflex. it causes the EPAP to wander some.Sleepygal1 wrote: ↑Sun Feb 17, 2019 7:30 pm. I am not getting it. The IPAP line is static at 18 as I expect, but the EPAP is ramping up from 11 -> 13 (when 11 is selected as EPAP!). I checked all the settings and see nothing related to PS, Ramp, Flex is turned off, nothing obvious. Can anyone tell me what is going on???
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
Sleepygal1
- Posts: 40
- Joined: Tue Nov 20, 2018 2:13 pm
Re: Am I not understanding sleepyhead or a bipap machine settings?
I turned off Flex, so there shouldn't be any problems from that. Here's a pic to prove it:
Last edited by Sleepygal1 on Sun Feb 17, 2019 8:00 pm, edited 1 time in total.
Re: Am I not understanding sleepyhead or a bipap machine settings?
Yes it does cause some roaming but she says Flex is turned off.
And when Flex is used it sort of causes some bouncing around a little...not one straight line and certainly not what is shown on this EPAP graph.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Am I not understanding sleepyhead or a bipap machine settings?
I actually don't like to advise using PS over say 6...too much PS can cause problems unless the person has some specific ventilation needs for some reason.
My own personal preference for PS...4 or 5 to start with.
If you don't have issues with your lungs or problems breathing for some reason...I wouldn't go over 6 PS.
My own personal preference for PS...4 or 5 to start with.
If you don't have issues with your lungs or problems breathing for some reason...I wouldn't go over 6 PS.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Am I not understanding sleepyhead or a bipap machine settings?
I'm pretty sure *flex shows up on exhale only.. it used to not show up at all, then that changed.
If I'm wrong, then I've got no idea.
Last edited by palerider on Sun Feb 17, 2019 8:08 pm, edited 1 time in total.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Am I not understanding sleepyhead or a bipap machine settings?
I changed what I said because I just don't remember and I am looking for some of my old bilevel reports from the 760 to confirm or not.
Going to take me a while...they are way old..
I seem to remember it showing up on inhale but I can't swear to it. I am old and forgetful and we are talking years here.
But shouldn't matter...here with this OP....Flex is obviously off so can't blame that weird EPAP line on Flex.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Am I not understanding sleepyhead or a bipap machine settings?
This might be where I might like to run the data through Encore Pro for verification as to is SH buggy or not.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
Sleepygal1
- Posts: 40
- Joined: Tue Nov 20, 2018 2:13 pm
Re: Am I not understanding sleepyhead or a bipap machine settings?
Thanks Pugsy for the info. I've been using it in Auto bipap mode for quite a while, and this is what the 90% EPAP/IPAP results were from Sleepyhead. I am quite obese and I probably suffer from Obesity Hypoventilation syndrome, but not clinically diagnosed.Pugsy wrote: ↑Sun Feb 17, 2019 8:02 pmI actually don't like to advise using PS over say 6...too much PS can cause problems unless the person has some specific ventilation needs for some reason.
My own personal preference for PS...4 or 5 to start with.
If you don't have issues with your lungs or problems breathing for some reason...I wouldn't go over 6 PS.
I was mostly just surprised the PS was so large, and then after questioning the pressure levels that SleepyHead is showing me for fixed bipap mode last night, where I felt like it was giving me way too much air, I'm kinda questioning all its results now.
My best results were when i was running at: MinEPAP: 11, MaxIPAP 20, PS 5-8, so I was targetting what looked best in there.
-
Sleepygal1
- Posts: 40
- Joined: Tue Nov 20, 2018 2:13 pm
Re: Am I not understanding sleepyhead or a bipap machine settings?
I am not familiar with that program, is it something I can download?
Re: Am I not understanding sleepyhead or a bipap machine settings?
90 % numbers aren't necessarily what you should be using for fixed numbers.
It's it NOT where you were at for 90% of the night...people think this but it isn't.
90/95 % number are where we were AT OR BELOW for 90/95% of the night...people don't often know about the "or below" part of the definition.
I prefer to maybe start with the median average numbers if I am going to go with fixed mode.
What was the auto mode median numbers??
What kind of operating system is on your computer? Encore Pro is Windows only.
It's it NOT where you were at for 90% of the night...people think this but it isn't.
90/95 % number are where we were AT OR BELOW for 90/95% of the night...people don't often know about the "or below" part of the definition.
I prefer to maybe start with the median average numbers if I am going to go with fixed mode.
What was the auto mode median numbers??
What kind of operating system is on your computer? Encore Pro is Windows only.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
Sleepygal1
- Posts: 40
- Joined: Tue Nov 20, 2018 2:13 pm
Re: Am I not understanding sleepyhead or a bipap machine settings?
I am on Windows 10.Pugsy wrote: ↑Sun Feb 17, 2019 8:15 pm90 % numbers aren't necessarily what you should be using for fixed numbers.
It's it NOT where you were at for 90% of the night...people think this but it isn't.
90/95 % number are where we were AT OR BELOW for 90/95% of the night...people don't often know about the "or below" part of the definition.
I prefer to maybe start with the median average numbers if I am going to go with fixed mode.
What was the auto mode median numbers??
What kind of operating system is on your computer? Encore Pro is Windows only.
As you can see I've been trying to dial it in for some time