AHI goal.
-
Jcastellaw
- Posts: 37
- Joined: Wed Dec 26, 2018 1:20 am
AHI goal.
I’m sure that the goal for AHI is 0.0 but that’s an allowable ahi? I know doctors want under 5 but I’d like as low as possible. I’ve been playing with my minimum pressure and I’ve bounced all over the place.
My question is: how long should I leave my pressure at a certain lever before I get a good average at that level?
I’ll post some sleepyhead charts but just curious if I should leave my pressure the same for several days before I change it.
Thanks in advance.
My question is: how long should I leave my pressure at a certain lever before I get a good average at that level?
I’ll post some sleepyhead charts but just curious if I should leave my pressure the same for several days before I change it.
Thanks in advance.
_________________
| Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
| Additional Comments: ResMed S9 Autoset |
Re: AHI goal.
I tend to use pressure changes for a week to evaluate that change. In the beginning I was chasing pressure too often. As you are probably finding out we sleep different night to night due to the many other factors that affect our sleep.
_________________
| Mask: DreamWear Gel Nasal Pillow CPAP Mask with Headgear - Fit Pack (All Nasal Pillows with Medium Frame) |
| Additional Comments: Settings 11-20, EPR 2. Dreamstation Full Auto for back up |
-
Jcastellaw
- Posts: 37
- Joined: Wed Dec 26, 2018 1:20 am
Re: AHI goal.
That’s kinda what I was thinking. So say, 5-7 days and how much would you change the minimum pressure?
_________________
| Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
| Additional Comments: ResMed S9 Autoset |
-
realshelby
- Posts: 337
- Joined: Thu Jul 26, 2018 9:36 am
Re: AHI goal.
With help and knowledge from this site, I have tweeked my base pressure and what mask I use many times. I would go up and back 1/2 point of pressure, then go up 1 and so on. Now at 6.2-20.
My AHI averages under 1.
But it varies from a couple zeros to 1.8 or so on occasion.
So unless you see something in the Sleepyhead chart that indicates more starting pressure is needed, I would go at least 4-5 days before making slight tweaks to pressure. Or to exhalation relief. I am lucky in that I have found the humidity and temperature settings I am content with and never change those.
Different masks make a recordable difference in my AHI scores. But not more than 1 point.
My AHI averages under 1.
But it varies from a couple zeros to 1.8 or so on occasion.
So unless you see something in the Sleepyhead chart that indicates more starting pressure is needed, I would go at least 4-5 days before making slight tweaks to pressure. Or to exhalation relief. I am lucky in that I have found the humidity and temperature settings I am content with and never change those.
Different masks make a recordable difference in my AHI scores. But not more than 1 point.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Min 7 Max 20 (cmH2O)
Re: AHI goal.
Jcastellaw wrote: ↑Thu Feb 14, 2019 7:04 pmMy question is: how long should I leave my pressure at a certain lever before I get a good average at that level?
We have zero way to answer this question without seeing what is happening now and what that AHI is composed of with just guessing and I don't like just guessing.Jcastellaw wrote: ↑Thu Feb 14, 2019 7:04 pmjust curious if I should leave my pressure the same for several days before I change it.
You might need more pressure but you might not.
All depends on what that AHI is composed of.
More pressure won't do thing if the bulk of your AHI is central in nature. More pressure doesn't fix centrals.
Now if the bulk of your AHI is OA/hyponeas which are obstructive in nature then yes more pressure is usually what is advised.
So we don't know what your AHI is and we don't know what it is composed of...so can't really offer anything but a guess.
You don't want to go making any changes based on just 1 or 2 nights of data. We don't sleep the same each night and we can have fluke nights where there is a big change in results. Unless an urgent need is present for a change...a week's worth of data is a good baseline.
Urgent meaning a high obstructive in nature AHI and you know you were asleep when those obstructives got flagged.
Awake flagged events of any kind simply have to be removed from the evaluation of pressure needs.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- loggerhead12
- Posts: 585
- Joined: Fri Sep 01, 2017 4:00 pm
Re: AHI goal.
My goal is 1. I feel good around 1. 0 is better, but it's a unicorn that shows up on its own every once in a while. Don't chase unicorns.
2 doesn't sound all that far from 1, but it means you're waking up 16 times in an 8-hour night instead of just 8 times. That's huge.
Heck, people without sleep apnea wake up once per hour. Yeah, 1 is a good place to be.
2 doesn't sound all that far from 1, but it means you're waking up 16 times in an 8-hour night instead of just 8 times. That's huge.
Heck, people without sleep apnea wake up once per hour. Yeah, 1 is a good place to be.
Re: AHI goal.
Just don't lose sight of the "get a good night's sleep" goal along the way.Jcastellaw wrote: ↑Thu Feb 14, 2019 7:04 pmI’m sure that the goal for AHI is 0.0 but that’s an allowable ahi? I know doctors want under 5 but I’d like as low as possible. I’ve been playing with my minimum pressure and I’ve bounced all over the place.
My question is: how long should I leave my pressure at a certain lever before I get a good average at that level?
Like Thumper68, I would change my pressure weekly. There's no usually reason to rush it. But we don't know with no SH charts.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins |
-
Jcastellaw
- Posts: 37
- Joined: Wed Dec 26, 2018 1:20 am
Re: AHI goal.
Thanks guys. I’m gonna leave it where it’s at and post some charts in 7 days or so.
_________________
| Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
| Additional Comments: ResMed S9 Autoset |
Re: AHI goal.
Just to reinforce what djams said... when you're 'fine tuning', you have to work with *averages*, since sleep varies from night to night... don't diddle with the settings just because you had an off night... now, if you're having a string of off nights, then it warrants investigation.djams wrote: ↑Fri Feb 15, 2019 9:04 amJust don't lose sight of the "get a good night's sleep" goal along the way.Jcastellaw wrote: ↑Thu Feb 14, 2019 7:04 pmI’m sure that the goal for AHI is 0.0 but that’s an allowable ahi? I know doctors want under 5 but I’d like as low as possible. I’ve been playing with my minimum pressure and I’ve bounced all over the place.
My question is: how long should I leave my pressure at a certain lever before I get a good average at that level?
Like Thumper68, I would change my pressure weekly. There's no usually reason to rush it. But we don't know with no SH charts.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Dog Slobber
- Posts: 4336
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: AHI goal.
This will be obvious to p.r. and most experienced SleepyHead users, but many novices might benefit. I'm quoting his post for context.palerider wrote: ↑Fri Feb 15, 2019 11:38 am
Just to reinforce what djams said... when you're 'fine tuning', you have to work with *averages*, since sleep varies from night to night... don't diddle with the settings just because you had an off night... now, if you're having a string of off nights, then it warrants investigation.
SleepyHead has a good mechanism for comparing prescription changes. At the bottom of the Statistics Tab, SleepyHead will display a list include AHI with a demarc for every prescription change. Using this tool, you can easily see the Average AHI and days for every unique prescription.
Dark Red - Worse AHI
Light Red - 2nd worse AHI
Dark Green - Best AHI
Light Green - 2nd best AHI
You can also use it to create an artificial demarcation if you want to see if there is another change that might impact your AHI.
For example let say you are changing some medication and you want to see if it affects your AHI. Say your pressure is 12-20cm. Change your pressure to 12-19.8cm, this won't change your therapy, but it will create a demarc point on the SleepyHead chart where you started the new medication.
Now you can easily compare your AHI pre and post medication change. Of course, you need to give your averages some time to settle.
Last night I started some CBTi Therapy to help with night awakenings. I changed my max to create a demarcation point. Now I'll be able easily see if it impacts my AHI.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
-
Jcastellaw
- Posts: 37
- Joined: Wed Dec 26, 2018 1:20 am
Re: AHI goal.
This is an Idea where I’m at.
_________________
| Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
| Additional Comments: ResMed S9 Autoset |
- Attachments
-
- screenshot-20190213-104051.png (101.17 KiB) Viewed 3814 times
-
Jcastellaw
- Posts: 37
- Joined: Wed Dec 26, 2018 1:20 am
Re: AHI goal.
Puggy, let me know what you think.
_________________
| Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
| Additional Comments: ResMed S9 Autoset |
Re: AHI goal.
How are you sleeping and feeling in general?
I see one break in therapy for sure last night...why?
Were there more times where you woke up but you didn't turn the machine off?
How is your nasal congestion? Having much or is the nose clear for the most part?
Answer these questions and I can better offer an opinion.
I see one break in therapy for sure last night...why?
Were there more times where you woke up but you didn't turn the machine off?
How is your nasal congestion? Having much or is the nose clear for the most part?
Answer these questions and I can better offer an opinion.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: AHI goal.
let us know how it goes for you. i've heard of this therapy. in fact, i think there was some discussion about it the first year i got here.Dog Slobber wrote: ↑Fri Feb 15, 2019 12:12 pm<snip>
Last night I started some CBTi Therapy to help with night awakenings. I changed my max to create a demarcation point. Now I'll be able easily see if it impacts my AHI.
i'd be curious to see how it goes.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
-
Jcastellaw
- Posts: 37
- Joined: Wed Dec 26, 2018 1:20 am
Re: AHI goal.
Pugsy wrote: ↑Sun Feb 17, 2019 7:10 pmHow are you sleeping and feeling in general?
I see one break in therapy for sure last night...why?
Were there more times where you woke up but you didn't turn the machine off?
How is your nasal congestion? Having much or is the nose clear for the most part?
Answer these questions and I can better offer an opinion.
I feel decent when I wake up. So much better than when my AHI was 17. I’m guessing I woke up to go to the bathroom. No nasal congestion and I’m breathing really well. I use Zicam if I get congested. It’s awesome.
_________________
| Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
| Additional Comments: ResMed S9 Autoset |