Cpap mask exhaust diffuser filter developing black spots overnight.
Cpap mask exhaust diffuser filter developing black spots overnight.
Hi, I’m looking for some help/ answer to a concern. I have a Resmed airsense10 machine and use a F&P Eson 2 nasal mask. My mask diffuser filter has black dots that match the small “holes” in the plastic clip that holds it in place. I have removed the original “factory” filter because of this reason and use replacement filters. This occurs nightly and has raised a concern of where it’s coming from and is it harmful? We are non-smokers in our household. This has been occurring for several months. I have only used a Cpap for under 2 yrs and use my machine every night. My humidity level usually stays set at “4”. My machine was new and I find it hard to believe it’s damaged. I’ve read another feed that someone considered this may be mildew forming. Is it possible that mildew could form that quickly (roughly 8 hrs) even on a new filter. Looking for any other cpap user that my have this issue and have you found an answer? Thank you.
- Sheriff Buford
- Posts: 4111
- Joined: Mon Aug 09, 2010 8:01 am
- Location: Kingwood, Texas
Re: Cpap mask exhaust diffuser filter developing black spots overnight.
Sounds like it is mildew or mold. I would clean the area real good with a splash of bleach or vinegar. Can you go one night without your humidifier? If it was me, I'd go one night without the humidifier and see if the black spots appear the next morning.
Sheriff
Sheriff
_________________
| Machine: AirSense 11 Autoset |
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
Re: Cpap mask exhaust diffuser filter developing black spots overnight.
Is it possible that mildew/mold to form so quickly? I keep my equipment clean. My masks are washed daily and I equip a new/clean one every night. After 7 hrs of use it occurred again. I’m trying to link photos to see the issue but having issue with uploading to site. Thanks for the input.
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: Cpap mask exhaust diffuser filter developing black spots overnight.
Any chance there has been a candle burning?
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
- BlueDragon
- Posts: 563
- Joined: Sun May 27, 2018 10:20 am
- Location: Sonoran Desert
Re: Cpap mask exhaust diffuser filter developing black spots overnight.
How does the filter in your CPAP machine look? The machine filter will be filtering air before it comes into the machine.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020. |
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
Re: Cpap mask exhaust diffuser filter developing black spots overnight.
As for the the reply about candles, yes my wife burns a candle in the living room every now and then. Nothing remotely close to the bedrooms which are in the back of the house. Also my machine filter is clean and is changed regularly. I’ve never seen this black on the machine filters.
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Cpap mask exhaust diffuser filter developing black spots overnight.
If deposits are on the inside of the filter, take said filter to your doctor.
Are you exposed to fine particulate at work?
OSHA may be able to test the deposits for you.
Your work environment may be unsafe--or harmless . . . .
The bullfrog used to be a factory spray painter.
He wore a mask at work, and I could see overspray on his face
where his goggles and breathing mask did not cover.
The only time he did not have solvent on his breath at night
was at least a WEEK into vacation.
Are you exposed to fine particulate at work?
OSHA may be able to test the deposits for you.
Your work environment may be unsafe--or harmless . . . .
The bullfrog used to be a factory spray painter.
He wore a mask at work, and I could see overspray on his face
where his goggles and breathing mask did not cover.
The only time he did not have solvent on his breath at night
was at least a WEEK into vacation.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
-
Fatterthanever
- Posts: 1
- Joined: Wed Jul 07, 2021 12:05 am
Re: Cpap mask exhaust diffuser filter developing black spots overnight.
I believe this is from the black foam filter particles originating whithin the motor housing. This foam filter is installed for noise reduction and has been the cause of a huge recall in the Phillips brand dreamstation cpap model. The foam breaks down and is sent along with air to the mask. This is why your external intake filter is clean but your diffuser is showing this material, as it is being expelled from inside the blower motor housing. This is not mold, it's likey carcinogenic polyester foam particles from the machine's sound reducing internal foam filter.
Re: Cpap mask exhaust diffuser filter developing black spots overnight.
The particles are not carcinogenic - that they are has been repeated over and over without basis in fact and I'd remove that diagnosis from the argument.
There may be a mold problem, but only you are in a position to judge that. In any case, can you possibly e.g. rent another machine (same type but different mfgr), adjust pressures etc. to fit the new machines algorithms and examine your set up closely for possible mold issues (in your home, room, use of humidifier, etc) and see if the problem persists.
There may be a mold problem, but only you are in a position to judge that. In any case, can you possibly e.g. rent another machine (same type but different mfgr), adjust pressures etc. to fit the new machines algorithms and examine your set up closely for possible mold issues (in your home, room, use of humidifier, etc) and see if the problem persists.
Re: Cpap mask exhaust diffuser filter developing black spots overnight.
OP in this thread is using a ResMed machine anyway....and it's a 2 year old post.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Cpap mask exhaust diffuser filter developing black spots overnight.
Fatterthanever, (first time poster) revived this zombie thread.
Went all the way back to early 2019 to find this oddball problem.
FIRST POST--EVER
Newbie mistake?
Went all the way back to early 2019 to find this oddball problem.
FIRST POST--EVER
Newbie mistake?
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: Cpap mask exhaust diffuser filter developing black spots overnight.
I believe in little green men visiting in a flying saucer!
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
ILoveFlowers
- Posts: 159
- Joined: Sat Feb 20, 2021 1:58 pm
- Location: Pacific Northwest
Re: Cpap mask exhaust diffuser filter developing black spots overnight.
Benefit of the doubt/ the post may have come up in response to a google search and Fatter didn't notice the date on it. Cpaptalk posts are very visible on the google! That said, I wonder what it turned out to be?chunkyfrog wrote: ↑Wed Jul 07, 2021 9:01 amFatterthanever, (first time poster) revived this zombie thread.
Went all the way back to early 2019 to find this oddball problem.
FIRST POST--EVER
Newbie mistake?
AirSense 10 AutoSet For Her
Airfit P10 mask
Airfit P10 mask
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Cpap mask exhaust diffuser filter developing black spots overnight.
Dust motes, body lice, bedbugs, ashes from a fire, or
even floaters inside his eyes?
Probably nothing of consequence--as usual.
even floaters inside his eyes?
Probably nothing of consequence--as usual.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: Cpap mask exhaust diffuser filter developing black spots overnight.
No, experience has taught us that first posters that come in and make some declaration are almost always full of crap, or trolls, and need to be ignored.ILoveFlowers wrote: ↑Sun Jul 11, 2021 1:37 amBenefit of the doubt/chunkyfrog wrote: ↑Wed Jul 07, 2021 9:01 amFatterthanever, (first time poster) revived this zombie thread.
Went all the way back to early 2019 to find this oddball problem.
FIRST POST--EVER
Newbie mistake?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.