I tried searching for this and found nothing, but apologies if I missed it and it's been addressed previously.
I've recently begun using my PR S1 ASV again, and the Sleep, Wake, and Hours numbers in Sleepyhead aren't being shown correctly. Oddly, importing the data into Encore Pro yields correct numbers.
The bug is that several nights in a row will show the exact same time of sleep, right down to the second, which is obviously impossible. Strangely, the sleep and wake times move from night to night.
An example may be helpful:
1/19/19 Sleep 1:30:11 Wake 4:03:52 Hours 2:33:41
1/20/19 Sleep 1:54:29 Wake 4:28:10 Hours 2:33:41
1/21/19 Sleep 2:00:58 Wake 4:34:39 Hours 2:33:41
1/22/19 Sleep 3:10:00 Wake 5:43:41 Hours 2:33:41
Any ideas what's going on? If it were just affecting the times, I wouldn't care, but whatever is going also appears to be skewing the AHI calculations.
Thank you!
Sleepyhead Sleep/Wake/Hours Incorrect
Re: Sleepyhead Sleep/Wake/Hours Incorrect
Which version of SleepyHead?
I may have to RISE but I refuse to SHINE.
Re: Sleepyhead Sleep/Wake/Hours Incorrect
It's just buggy software when it comes to the PR S1 ASV machines and actually a couple of the other high end PR S1 models.
You might try this version...not officially released yet and not without some bugs of its own but I think the ASV bugs were ironed out.
https://sleepyhead.jedimark.net/testing/1.1.0/
You might try this version...not officially released yet and not without some bugs of its own but I think the ASV bugs were ironed out.
https://sleepyhead.jedimark.net/testing/1.1.0/
I may have to RISE but I refuse to SHINE.
Re: Sleepyhead Sleep/Wake/Hours Incorrect
I just installed that testing version and even started with a fresh profile, but unfortunately, that didn't change anything. The testing version looks nice though!
Re: Sleepyhead Sleep/Wake/Hours Incorrect
Bummer. I was hoping that would fix things for you.
Since Encore is reporting the correct time frames then we know the machine is writing to the SD card correctly.
That leaves a SleepyHead bug of some sort. Have you changed any of the little settings lately? I know when I was using the 960 there was some little something that I did that caused the reports to go whacko but I didn't pay enough attention back then to document whatever it was I did. It was something that I turned on or off that seemed to trigger the mess up in the SH reports. I didn't think too much of it until weeks later and I saw the same thing in someone else's 960 report...unfortunately I don't remember what it was.
We have another forum member using the 1060 model and he gets the same weirdo sleeping times like you are getting...never could figure out what it was. Thought it might be multiple sessions confusing things but it also just showed approx the same amount of time you are seeing and nothing else even when there was just one session.
Since Encore is reporting the correct time frames then we know the machine is writing to the SD card correctly.
That leaves a SleepyHead bug of some sort. Have you changed any of the little settings lately? I know when I was using the 960 there was some little something that I did that caused the reports to go whacko but I didn't pay enough attention back then to document whatever it was I did. It was something that I turned on or off that seemed to trigger the mess up in the SH reports. I didn't think too much of it until weeks later and I saw the same thing in someone else's 960 report...unfortunately I don't remember what it was.
We have another forum member using the 1060 model and he gets the same weirdo sleeping times like you are getting...never could figure out what it was. Thought it might be multiple sessions confusing things but it also just showed approx the same amount of time you are seeing and nothing else even when there was just one session.
I may have to RISE but I refuse to SHINE.
Re: Sleepyhead Sleep/Wake/Hours Incorrect
Yeah, I've tweaked both the humidity and the pressure settings lately. Looking at the session information, each night is showing up as a separate and different session, so at least it looks like it's not some cross-session merging bug.
I do see the bug going back to some very old data now too, and I think I would have noticed it back then...but given my brain power (or lack thereof!) these days, who knows.
I do see the bug going back to some very old data now too, and I think I would have noticed it back then...but given my brain power (or lack thereof!) these days, who knows.