Started CPAP with AirSense 10 AutoSet. SleepyHead showing a lot of clear airway apneas. Is this normal?

Post by Highlander » Fri Jan 25, 2019 2:32 am

I have been having really bad sleep for years and finally I did a sleep study where my AHI was 9 with no central apneas detected. I started CPAP with an AirSense 10 AutoSet. I got an SD card and have been downloading the data every day into SleepyHead for almost a month now without modifying anything. My AHI is now between 2 and 5 but the majority of the remaining apnea events tend to be clear airways (CA). Overall I feel like I am sleeping and feeling better but I am concerned about the CAs taking the majority of the apnea times. Is this normal or should I be worried and consider switching to an ASV device? I tried turning off the EPR (currently set to 2) thinking it was causing the CAs for a couple of nights but that did not decrease my CAs and instead increased my hypopneas and overall AHI so I turned it back on. My follow up appointment with my doctor is in a couple of weeks. I attached pictures of a few days of data from SleepyHead.

Thank you for any advice.
Attachments
screenshot-20190124-235344.png
screenshot-20190124-235344.png (157.97 KiB) Viewed 2768 times
screenshot-20190124-235249.png
screenshot-20190124-235249.png (155.32 KiB) Viewed 2768 times
screenshot-20190124-235134.png
screenshot-20190124-235134.png (156.2 KiB) Viewed 2768 times

User avatar
Pugsy
Posts: 65144
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Started CPAP with AirSense 10 AutoSet. SleepyHead showing a lot of clear airway apneas. Is this normal?

Post by Pugsy » Fri Jan 25, 2019 8:20 am

Welcome to the forum.

Go here and learn how to figure out of the centrals are real or not
http://freecpapadvice.com/sleepyhead-free-software

If not real as in you weren't asleep...they simply don't count and have to be mentally removed from the evaluation process.

An occasion central is normal and no big deal....and even if every single one of your centrals was the real deal (which I doubt) you aren't having enough of them all night long to warrant going to ASV.

Even if related to cpap use...they might reduce with time.
So doctors won't usually go to ASV right away unless the number of centrals is a LOT higher than what you are seeing...assuming they are the real deal.

Do you remember waking up during the night?
Do you take any meds of any kind? If so, what?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

zheshi14
Posts: 3
Joined: Tue Oct 09, 2018 8:19 pm

Re: Started CPAP with AirSense 10 AutoSet. SleepyHead showing a lot of clear airway apneas. Is this normal?

Post by zheshi14 » Fri Jan 25, 2019 10:46 am

What do you mean if they are real or not? And how would I go about determining so?

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Started CPAP with AirSense 10 AutoSet. SleepyHead showing a lot of clear airway apneas. Is this normal?

Post by palerider » Fri Jan 25, 2019 12:20 pm

Also, please fix your charts.
(instructions in the link below).

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

nicholasjh1
Posts: 517
Joined: Wed Jun 14, 2017 9:07 am

Re: Started CPAP with AirSense 10 AutoSet. SleepyHead showing a lot of clear airway apneas. Is this normal?

Post by nicholasjh1 » Fri Jan 25, 2019 12:26 pm

I had quite a few when I started, they are gone now. but of course your milage may vary. It's not unusual to start with some though.
Instead of Sleep apnea it should be called "Sleep deprivation, starving of oxygen, being poisoned by high CO2 levels, damaging the body and brain while it's supposed to be healing so that you constantly get worse and can never get healthy Apnea"

User avatar
Pugsy
Posts: 65144
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Started CPAP with AirSense 10 AutoSet. SleepyHead showing a lot of clear airway apneas. Is this normal?

Post by Pugsy » Fri Jan 25, 2019 12:37 pm

zheshi14 wrote:
Fri Jan 25, 2019 10:46 am
What do you mean if they are real or not? And how would I go about determining so?
Real...as in asleep

Not real....as in not fully asleep

All explained in the link I posted.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Started CPAP with AirSense 10 AutoSet. SleepyHead showing a lot of clear airway apneas. Is this normal?

Post by palerider » Fri Jan 25, 2019 1:27 pm

A few "centrals" are common, and nothing to be concerned about.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Highlander
Posts: 15
Joined: Fri Jan 25, 2019 1:58 am

Re: Started CPAP with AirSense 10 AutoSet. SleepyHead showing a lot of clear airway apneas. Is this normal?

Post by Highlander » Fri Jan 25, 2019 11:53 pm

I am not taking any medication. I may be awake during some of those but I am not 100% sure. I need to find a better way to track when exactly I am awaken to be able to correlate it.

Here are updated screenshots with better layout.

Thanks!
Attachments
screenshot-20190125-214314.png
screenshot-20190125-214314.png (138.72 KiB) Viewed 2671 times
screenshot-20190125-214231.png
screenshot-20190125-214231.png (137.03 KiB) Viewed 2671 times
screenshot-20190125-214214.png
screenshot-20190125-214214.png (139.48 KiB) Viewed 2671 times

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Started CPAP with AirSense 10 AutoSet. SleepyHead showing a lot of clear airway apneas. Is this normal?

Post by palerider » Sat Jan 26, 2019 12:37 am

Highlander wrote:
Fri Jan 25, 2019 11:53 pm
I need to find a better way to track when exactly I am awaken to be able to correlate it.
Actually, what you need to do is to stop worrying about a few "centrals".

And maybe turn down EPR.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Started CPAP with AirSense 10 AutoSet. SleepyHead showing a lot of clear airway apneas. Is this normal?

Post by chunkyfrog » Sat Jan 26, 2019 12:48 am

We breathe differently, whether sound asleep or in transition (falling asleep or waking up)
Transitional stuff may look like centrals, but we call them sleep/wake/junk--"SWJ".
They are meaningless, but the machine may not be able to distinguish them.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

ajack
Posts: 977
Joined: Thu Mar 02, 2017 2:54 am
Location: australia

Re: Started CPAP with AirSense 10 AutoSet. SleepyHead showing a lot of clear airway apneas. Is this normal?

Post by ajack » Sat Jan 26, 2019 5:22 am

The CA on the machine actually means clear airway and not central apnea. It can't tell the difference. As has been said, clear airway can be caused by a few things, including new users getting their CO2 and O2 stable. The body waits till the co2 builds up and trigger a breath. sometimes they can be a true central apnea with an O2 drop. You have a relatively low number that are clustered. I wouldn't worry and they will probably settle down in a month or so.

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 ST-A iVAPS and adapt ASV

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Started CPAP with AirSense 10 AutoSet. SleepyHead showing a lot of clear airway apneas. Is this normal?

Post by palerider » Sat Jan 26, 2019 1:07 pm

ajack wrote:
Sat Jan 26, 2019 5:22 am
The CA on the machine actually means clear airway and not central apnea. It can't tell the difference. As has been said, clear airway can be caused by a few things, including new users getting their CO2 and O2 stable. The body waits till the co2 builds up and trigger a breath. sometimes they can be a true central apnea with an O2 drop. You have a relatively low number that are clustered. I wouldn't worry and they will probably settle down in a month or so.
Best to just ignore ajack, he reads a lot, but understands little, and thus tends to just confuse any issue he chooses to wade into.

"clear airway' is a term made up by Respironics.... sleepyhead was originally written for Respironics machines... the 'clear airway' term is a holdover from that.

Resmed does not use the term "clear airway", anywhere, nor does any other manufacturer in the sleep disordered breathing treatment industry.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Sat Jan 26, 2019 5:04 pm, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Highlander
Posts: 15
Joined: Fri Jan 25, 2019 1:58 am

Re: Started CPAP with AirSense 10 AutoSet. SleepyHead showing a lot of clear airway apneas. Is this normal?

Post by Highlander » Sat Jan 26, 2019 4:53 pm

Thank you all for your advises on this. For what it is worth, when I load my SD card data in ResScan 6 from ResMed, those events are called Centrals, not Clear Airways.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Started CPAP with AirSense 10 AutoSet. SleepyHead showing a lot of clear airway apneas. Is this normal?

Post by palerider » Sat Jan 26, 2019 5:03 pm

Highlander wrote:
Sat Jan 26, 2019 4:53 pm
Thank you all for your advises on this. For what it is worth, when I load my SD card data in ResScan 6 from ResMed, those events are called Centrals, not Clear Airways.
Yeah, another 'confusing issue' (tm) from Sleepyhead.

Hopefully, the new sleepyhead team can address this (hint hint) and other things, like inappropriate flagging of leaks on Respironics machines.

What you're having there are central apneas... yes. What pugsy and others were getting at is that they may, or may not be central *sleep* apneas. If you wake up a little, cuz you're in an uncomfortable position, and you wake up a bit to shift position, it's quite natural and common to hold your breath for a bit while doing so. Not breathing is an apnea, not trying to breathe means it's central not obstructive... but the machine can't tell whether you're still asleep, so it just flags them... as 'central apnea'.

Even if they're real, honest to goodness central sleep apneas, there aren't enough of them on your charts to be concerned about.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

ajack
Posts: 977
Joined: Thu Mar 02, 2017 2:54 am
Location: australia

Re: Started CPAP with AirSense 10 AutoSet. SleepyHead showing a lot of clear airway apneas. Is this normal?

Post by ajack » Wed Jan 30, 2019 11:10 pm

Highlander wrote:
Sat Jan 26, 2019 4:53 pm
Thank you all for your advises on this. For what it is worth, when I load my SD card data in ResScan 6 from ResMed, those events are called Centrals, not Clear Airways.
Well the machine can't distinguish between a true central and simply not breathing without an obstruction and registering CA, AKA clear airway. There are causes other than a true central apnea. To get yourself into a state without knowing the facts isn't a good idea. Palerider should be aware of this.