how do I turn off AirCurve ASV?
- beautifuldreamer
- Posts: 398
- Joined: Fri Feb 01, 2013 7:47 pm
- Location: Tennessee
how do I turn off AirCurve ASV?
I can’t seem to turn off the AirCurve. I try pulling off the mask and pressing the start stop button. Nothing works except to pull the plug. And if I put the plug back in it starts blowing the air again.
Thanks for any advice
Update. I figured it out. I was trying to turn it off the way I used to with the old Respironics ASV and that wasn’t working. Just had to press and not do a prolonged hold. hold
Thanks for any advice
Update. I figured it out. I was trying to turn it off the way I used to with the old Respironics ASV and that wasn’t working. Just had to press and not do a prolonged hold. hold
Last edited by beautifuldreamer on Tue Jan 22, 2019 9:21 am, edited 1 time in total.
Machine: ResMed AirSense 11 Autoset
Mask: Bleep
Mask: Bleep
Re: Help how do I turn off AirCurve ASV?
It shouldn't be doing that.
Pushing the button to stop it...should work.
And when you pull the plug and then plug it back in...it shouldn't immediately go to blowing.
Call up the supplier tell them it is possessed by little devils.
Pushing the button to stop it...should work.
And when you pull the plug and then plug it back in...it shouldn't immediately go to blowing.
Call up the supplier tell them it is possessed by little devils.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- beautifuldreamer
- Posts: 398
- Joined: Fri Feb 01, 2013 7:47 pm
- Location: Tennessee
Re: Help how do I turn off AirCurve ASV?
LOL - Thanks I will call the local branch office, I was really confused when it kept blowing and my first wake up when I had to go to the restroom, I decided to simply unplug the thing -
but on a good note, last night was my first ever 0 AHI since starting PAP therapy!
Machine: ResMed AirSense 11 Autoset
Mask: Bleep
Mask: Bleep
Re: how do I turn off AirCurve ASV?
So the answer was to press the button and not hold it too long, right?
Reminds me of the radio in my last car. If you didn't hit the on/off button long enough, nothing happened, whereas if you held it too long, it was like hitting it twice (on and off again, or off an on again). The time tolerance was very narrow.
Reminds me of the radio in my last car. If you didn't hit the on/off button long enough, nothing happened, whereas if you held it too long, it was like hitting it twice (on and off again, or off an on again). The time tolerance was very narrow.
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: how do I turn off AirCurve ASV?
Changing brands can mess you up.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: how do I turn off AirCurve ASV?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: how do I turn off AirCurve ASV?
Turn on 'smartstart', it should turn itself off after a few seconds.beautifuldreamer wrote: ↑Tue Jan 22, 2019 9:18 amI can’t seem to turn off the AirCurve. I try pulling off the mask and pressing the start stop button. Nothing works except to pull the plug. And if I put the plug back in it starts blowing the air again.
Thanks for any advice
Update. I figured it out. I was trying to turn it off the way I used to with the old Respironics ASV and that wasn’t working. Just had to press and not do a prolonged hold. hold
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- beautifuldreamer
- Posts: 398
- Joined: Fri Feb 01, 2013 7:47 pm
- Location: Tennessee
Re: how do I turn off AirCurve ASV?
Thanks, I just checked, Smartstart is turned on.palerider wrote: ↑Tue Jan 22, 2019 5:20 pmTurn on 'smartstart', it should turn itself off after a few seconds.beautifuldreamer wrote: ↑Tue Jan 22, 2019 9:18 amI can’t seem to turn off the AirCurve. I try pulling off the mask and pressing the start stop button. Nothing works except to pull the plug. And if I put the plug back in it starts blowing the air again.
Thanks for any advice
Update. I figured it out. I was trying to turn it off the way I used to with the old Respironics ASV and that wasn’t working. Just had to press and not do a prolonged hold. hold
And yes, I believe that was the main problem, holding down on the on/off button for too long-D.H. wrote: ↑Tue Jan 22, 2019 10:09 amSo the answer was to press the button and not hold it too long, right?
Reminds me of the radio in my last car. If you didn't hit the on/off button long enough, nothing happened, whereas if you held it too long, it was like hitting it twice (on and off again, or off an on again). The time tolerance was very narrow.
on my other device, I used to have to hold at least for a couple of seconds before the respironics asv would turn off.
YES, I am trainable, and tonight I will hold down just for a brief touch, rather than the extended hold . . .
Machine: ResMed AirSense 11 Autoset
Mask: Bleep
Mask: Bleep
Re: how do I turn off AirCurve ASV?
Is your mask type (and hose type if not using a heated hose) set correctly?beautifuldreamer wrote: ↑Tue Jan 22, 2019 6:34 pmThanks, I just checked, Smartstart is turned on.palerider wrote: ↑Tue Jan 22, 2019 5:20 pmTurn on 'smartstart', it should turn itself off after a few seconds.beautifuldreamer wrote: ↑Tue Jan 22, 2019 9:18 amI can’t seem to turn off the AirCurve. I try pulling off the mask and pressing the start stop button. Nothing works except to pull the plug. And if I put the plug back in it starts blowing the air again.
Thanks for any advice
Update. I figured it out. I was trying to turn it off the way I used to with the old Respironics ASV and that wasn’t working. Just had to press and not do a prolonged hold. hold
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- beautifuldreamer
- Posts: 398
- Joined: Fri Feb 01, 2013 7:47 pm
- Location: Tennessee
Re: how do I turn off AirCurve ASV?
Yes currently the mask and hose type correct - i believe it was more because i was holding/pressing the on/off button too long that I was having the problems . . . Sometimes I had to hold the on/off button for 2 to 3 seconds for my Respironics ASV to turn off . . .palerider wrote: ↑Tue Jan 22, 2019 7:42 pmIs your mask type (and hose type if not using a heated hose) set correctly?beautifuldreamer wrote: ↑Tue Jan 22, 2019 6:34 pmThanks, I just checked, Smartstart is turned on.palerider wrote: ↑Tue Jan 22, 2019 5:20 pmTurn on 'smartstart', it should turn itself off after a few seconds.beautifuldreamer wrote: ↑Tue Jan 22, 2019 9:18 amI can’t seem to turn off the AirCurve. I try pulling off the mask and pressing the start stop button. Nothing works except to pull the plug. And if I put the plug back in it starts blowing the air again.
Thanks for any advice
Update. I figured it out. I was trying to turn it off the way I used to with the old Respironics ASV and that wasn’t working. Just had to press and not do a prolonged hold. hold
*on a side note, I am waiting to receive a heated hose to be shipped from my DME, I remembered to ask my Sleep Doc for the heated hose After he already placed the prescription/order for the new ASV. The local dme branch didn't have any heated hoses in stock or they would have given me one . . .
Machine: ResMed AirSense 11 Autoset
Mask: Bleep
Mask: Bleep
- beautifuldreamer
- Posts: 398
- Joined: Fri Feb 01, 2013 7:47 pm
- Location: Tennessee
Re: how do I turn off AirCurve ASV?
hmmm - after pushing stop, the air blowing appeared to turn off, but then after I returned from the bathroom, it seemed that there was still low level air blowing through the tube. And the only way to stop the low level air blowing was to un-plug from power. Then it stopped, totally.
Is that normal, to have the low level of air blowing?
Is that normal, to have the low level of air blowing?
Machine: ResMed AirSense 11 Autoset
Mask: Bleep
Mask: Bleep
Re: how do I turn off AirCurve ASV?
The low level air flow is normal.
Called the cool down period. Mainly it's air moving through the hose to dry things out a bit.
It will end 30 to 45 minutes on its own.
Only way to stop it is to pull the main plug.
Called the cool down period. Mainly it's air moving through the hose to dry things out a bit.
It will end 30 to 45 minutes on its own.
Only way to stop it is to pull the main plug.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- beautifuldreamer
- Posts: 398
- Joined: Fri Feb 01, 2013 7:47 pm
- Location: Tennessee
Re: how do I turn off AirCurve ASV?
Thank you very much! Yay, my device is working fine
As well as I am breathing through the night wonderfully with my new ASV - 2 nights in a row, 0 AHI
Machine: ResMed AirSense 11 Autoset
Mask: Bleep
Mask: Bleep
- beautifuldreamer
- Posts: 398
- Joined: Fri Feb 01, 2013 7:47 pm
- Location: Tennessee
Re: how do I turn off AirCurve ASV?
LOL this is too funny -
I realized this morning that I was not pressing on the on/off button !!
I have been pressing on the round dial that is to change settings/etc.
Ok, now I understand what the problem was- Just this morning after waking up and looking at my new ResMed ASV, I saw the on/off button with the power symbol on the top of the device!
I realized this morning that I was not pressing on the on/off button !!
I have been pressing on the round dial that is to change settings/etc.
Ok, now I understand what the problem was- Just this morning after waking up and looking at my new ResMed ASV, I saw the on/off button with the power symbol on the top of the device!
Machine: ResMed AirSense 11 Autoset
Mask: Bleep
Mask: Bleep
Re: how do I turn off AirCurve ASV?
those dang tricksy resmed machines!beautifuldreamer wrote: ↑Thu Jan 24, 2019 10:41 amLOL this is too funny -
I realized this morning that I was not pressing on the on/off button !!
I have been pressing on the round dial that is to change settings/etc.
Ok, now I understand what the problem was- Just this morning after waking up and looking at my new ResMed ASV, I saw the on/off button with the power symbol on the top of the device!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg