Aircurve 10 ASV Setup for non-heated tubing

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
cpaok
Posts: 1
Joined: Fri Dec 21, 2018 9:57 pm

Aircurve 10 ASV Setup for non-heated tubing

Post by cpaok » Fri Dec 21, 2018 10:02 pm

In My Options do i select Tubing, and then Standard to select non-heated 8 foot tubing ? I was told Slimline is for Climate Control Heated 6 foot tubing. Thank you

User avatar
LSAT
Posts: 13347
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Aircurve 10 ASV Setup for non-heated tubing

Post by LSAT » Sat Dec 22, 2018 10:42 am

cpaok wrote:
Fri Dec 21, 2018 10:02 pm
In My Options do i select Tubing, and then Standard to select non-heated 8 foot tubing ? I was told Slimline is for Climate Control Heated 6 foot tubing. Thank you
YOU were told wrong. If you are using a non heated hose, you choose standard or slimline depending on the diameter of the hose. If you do get a heated hose the machine will adjust itself when the hose is plugged in. (Must have been one of the dummys at the DME)

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Aircurve 10 ASV Setup for non-heated tubing

Post by palerider » Sat Dec 22, 2018 2:39 pm

cpaok wrote:
Fri Dec 21, 2018 10:02 pm
In My Options do i select Tubing, and then Standard to select non-heated 8 foot tubing ? I was told Slimline is for Climate Control Heated 6 foot tubing. Thank you
If you've got a heated tube, then the type is set automatically, and you can't change it.

if you've got a non-heated tube, then select the right type, slim, standard, or 3m meter for a 3 meter (10 foot) tube.