Cracked resivoir
-
Audry.Camille
- Posts: 3
- Joined: Mon Nov 26, 2018 5:16 am
Cracked resivoir
After reading through several posts I didn't see what I've experienced. This is my 2nd resivoir and both have cracked along the bottom and into the seals. I only use distilled water, i don't put anything into the resivoir, and i daily wash/rinse it. I'm wondering if it has to do with ambient air temperature since my bedroom is always in the mid/high 60s (how the ventilation is) and the resivoir is heated.
No, i don't disassemble it
No, i haven't stepped on it or dropped it
I'm allergic to the majority of essential oils so they're banned from my home
Some insight is appreciated
Thanks
No, i don't disassemble it
No, i haven't stepped on it or dropped it
I'm allergic to the majority of essential oils so they're banned from my home
Some insight is appreciated
Thanks
Re: Cracked resivoir
Welcome to the forum.
Second water chamber to break over how long of a period of time?
What is your daily washing routine?
Do you use a dishwasher or hand wash it?
How hot does the water get that you wash and rinse it in?
Second water chamber to break over how long of a period of time?
What is your daily washing routine?
Do you use a dishwasher or hand wash it?
How hot does the water get that you wash and rinse it in?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Cracked reservoir
Is it possbile that - for whatever reason - the reservoir does not fit just right and you have push too hard to get it in or pull too hard to get it out?
- Wulfman...
- Posts: 6695
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: Cracked resivoir
Why do you feel the need to wash/rinse it daily if you're only using distilled water?Audry.Camille wrote: ↑Mon Nov 26, 2018 5:31 amAfter reading through several posts I didn't see what I've experienced. This is my 2nd resivoir and both have cracked along the bottom and into the seals. I only use distilled water, i don't put anything into the resivoir, and i daily wash/rinse it. I'm wondering if it has to do with ambient air temperature since my bedroom is always in the mid/high 60s (how the ventilation is) and the resivoir is heated.
No, i don't disassemble it
No, i haven't stepped on it or dropped it
I'm allergic to the majority of essential oils so they're banned from my home
Some insight is appreciated
Thanks
Are you sure it's actually cracked? Is it leaking?
Might the "crack" actually be a line from the mold it was formed in? Seen people mistake that before, too.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
-
Audry.Camille
- Posts: 3
- Joined: Mon Nov 26, 2018 5:16 am
Re: Cracked resivoir
The sleep therapist said to wash/rinse daily... so i am. And yes it's cracked- stepping in a pool of water first thing in the morning is not my normal routine. And the resivoir was bone dry. So there's that...
- Wulfman...
- Posts: 6695
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: Cracked resivoir
Well, maybe you should demand that your sleep therapist pay for/replace your HH tanks at NOT COST to you or your insurance because they should NOT crack or break. There is also no logical reason to wash/rinse them daily either. Distilled water is "pure" and the intake air filters in your machine should remove any other impurities in the air before it enters your machine.Audry.Camille wrote: ↑Mon Nov 26, 2018 1:51 pmThe sleep therapist said to wash/rinse daily... so i am. And yes it's cracked- stepping in a pool of water first thing in the morning is not my normal routine. And the resivoir was bone dry. So there's that...
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
- greatunclebill
- Posts: 1503
- Joined: Mon Feb 20, 2012 7:48 pm
- Location: L.A. (lower alabama)
Re: Cracked resivoir
your best answer will be found by bringing your machine with the broken reservoir to the business that sold it to you, where you get your supplies, and ask them what the problem is. they need to physically look at it. No one here can see and touch the machine or reservoir and tell you what the problem is. the temperature of your room is not going to be the answer.
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Additional Comments: myAir, OSCAR. cms-50D+. airsense 10 auto & (2009) remstar plus m series backups |
First diagnosed 1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990
- Okie bipap
- Posts: 3567
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: Cracked resivoir
The DME will always tell you to wash everything every day. The faster you wear things out or break them, the more they can sell you and make money. They are not in business to support you, they are there to maximize their profits. When I first started, the RT I was working with said she washed hers and her husband's equipment once a week. I have since gone to a schedule where I clean the mask cushions once a week, and everything once a month, with the exception being the hose. I never wash the hose unless I am recovering from one of my bronchitis attacks.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Evora Full Face Mask - Fitpack |
| Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
- Comfortably Numb
- Posts: 137
- Joined: Fri Mar 31, 2017 8:49 am
Re: Cracked resivoir
Sounds like bad luck to me. My reservoir began leaking and eventually ruined the top finish of my oak barn wood night stand. It wasn't a major blow out. Just small enough to make me run out of water every night for about a week before I discovered the problem and the underlying warped wood. Had to take it to the kitchen sink to confirm the leak along the seam. My tank was a little over one year old.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
-
Audry.Camille
- Posts: 3
- Joined: Mon Nov 26, 2018 5:16 am
Re: Cracked resivoir
Big help everyone! I have time to go by the supplies place later this week, so hopefully I'll get some answers.
Thanks!
-
prodigyplace
- Posts: 1795
- Joined: Mon Mar 27, 2017 5:45 pm
- Location: Central Virginia
Re: Cracked resivoir
The reservoirs usually supplied are not the dishwasher-safe ones. I rinse & dry mine out about every 3 months, as needed.
_________________
| Machine: AirSense 11 Autoset |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software |
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.
- babydinosnoreless
- Posts: 2375
- Joined: Fri Nov 02, 2018 2:53 pm
Re: Cracked resivoir
Washing the water tanks/ pillows wears them out ? How? I wash my pillow daily because gross, morning breath! Why wouldn't I wash it out ?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ N20 Nasal CPAP Mask with Headgear |
-
prodigyplace
- Posts: 1795
- Joined: Mon Mar 27, 2017 5:45 pm
- Location: Central Virginia
Re: Cracked resivoir
Washing the pillows by hand with a mild soap makes sense. Only distilled water & filtered air is in the water tank. Only occasional washing is needed.Cynmatthes wrote: ↑Tue Nov 27, 2018 9:21 amWashing the water tanks/ pillows wears them out ? How? I wash my pillow daily because gross, morning breath! Why wouldn't I wash it out ?
_________________
| Machine: AirSense 11 Autoset |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software |
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Cracked resivoir
SEE A DENTIST!Cynmatthes wrote: ↑Tue Nov 27, 2018 9:21 amWashing the water tanks/ pillows wears them out ? How? I wash my pillow daily because gross, morning breath! Why wouldn't I wash it out ?
Halitosis is a common symptom of dental problems.
If it is so bad it bothers YOU, it is past time for cleaning and repair.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
-
prodigyplace
- Posts: 1795
- Joined: Mon Mar 27, 2017 5:45 pm
- Location: Central Virginia
Re: Cracked resivoir
I assume you mean the teeth & not the CPAP machinechunkyfrog wrote: ↑Tue Nov 27, 2018 9:36 amSEE A DENTIST!Cynmatthes wrote: ↑Tue Nov 27, 2018 9:21 amWashing the water tanks/ pillows wears them out ? How? I wash my pillow daily because gross, morning breath! Why wouldn't I wash it out ?
Halitosis is a common symptom of dental problems.
If it is so bad it bothers YOU, it is past time for cleaning and repair.
_________________
| Machine: AirSense 11 Autoset |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software |