Resmed S9 autoset problem

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Resmed S9 autoset problem

Post by zoocrewphoto » Fri Oct 19, 2018 2:09 am

Fortunately, I have a spare/travel machine. So, we are functional for now.

My mom and I both use a Resmed S9 autoset. Tonight, hers was "blowing too much air". Not sure exactly what that means, but it sounds like it is starting and stopping once it gets to the starting pressure of 10 (no ramp). She has been using a range of 10-15 for the past 6 years or so (straight 10 before that), so she is used to 10. She said it feels like the mask is being blown off her face.

I was checking out her machine and noticed that the filter and cover were missing. We couldn't find it anywhere in the area, so I grabbed the one off mine to see if that corrected the problem. Nope. So, I grabbed my blower, hooked it up to her humidifier/power cord/etc and changed the range to hers. Started up, and it feels normal. So, something is up with her blower.

It doesn't appear to have any settings changed. She is on medicare, and it has been about 6 years since she got this machine, but she would probably have to do a couple doctor appointments and pay a copay. So, if I can't fix it, I will probably look for another used machine. I am grateful that I have a spare already, though I"m going out of town this weekend (friday morning) and was planning to use my spare/travel machine. I will use my travel machine while she uses my main machine. I may take her blower with me and see if I can experiment with it. I really don't have time tonight. I just wanted to get her set up as her oxygen goes down pretty low without cpap. She had her gallbladder removed last week, and they used a breathing tube because she went down so bad.

Does this issue sound familiar? Anything I can check on my own?

For anybody who doesn't have a spare - get one. These things happen at midnight!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?

User avatar
Pugsy
Posts: 65144
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed S9 autoset problem

Post by Pugsy » Fri Oct 19, 2018 6:50 am

About all I would know to check is the pressure with some sort of homemade manometer.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Resmed S9 autoset problem

Post by Okie bipap » Fri Oct 19, 2018 7:47 am

If she gets a new machine under Medicare, her payout would be around $20 per month for 13 months.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: Resmed S9 autoset problem

Post by zoocrewphoto » Sat Oct 20, 2018 1:07 pm

Update - problem solved!

My mom sleeps in the living room in a recliner with the machine on the floor. With the filter and cover missing, it sucked insome cat and/or dog hair. The cover propbably fell off in one of the hotels during our 3 week trip. I put the cover from my main machine on hers. We can tape it on one side to keep it from falling off again. When I get home, I will take the cover from my travel machine and move it to my main machine. I'm sure I can get another cover, so a fairly cheap repair. The main part was that we had a spare machine with available parts handy to troubleshoot the problem and/or use while figuring it out. So, no loss of use.

The timing did help too. During the trip, I did not have the spare machine with us. I took an extra mask and hose, but not the full machine. And I am out of town this weekend. So, it was good that she asked me Thursday night when I was still home and could change it out. She can't do the stairs to my room. And neither her or my dad are technically inclined. I am my parents' tech support for computer, phone, etc.

I did have trouble with rescscan importing data. I was going to see what it showed before I thought about the air intake being blocked. This laptop hasn't had either software on it, so I had to install. I will try it again at home on thedesktop. I am curious what the data shows. My mom mentioned Friday evening that it had been doing it off and on, though not as badly, for a few days.

So glad there is nothing actually wrong with the machine. Oh, and my mom may want to change her settings a little. I changed mine from 11-17 to her 10-15, but I didn't change anything else such as EPR, etc. I will have to see what hers is set at. She said she slept really well with my machine, and liked it even better.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?

User avatar
Pugsy
Posts: 65144
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed S9 autoset problem

Post by Pugsy » Sat Oct 20, 2018 9:08 pm

Are you by chance an Amazon Prime member?
It's an add on item if you are...$3.09
https://www.amazon.com/Resmed-36859-S9- ... B071GTZQS2

If you aren't a Prime member and can't find it any cheaper anywhere...get with me privately. I am always ordering something. :lol:

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: Resmed S9 autoset problem

Post by zoocrewphoto » Sat Oct 20, 2018 10:46 pm

Thanks!

My sister is a prime member.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?

User avatar
Muse-Inc
Posts: 4382
Joined: Fri Jun 12, 2009 8:44 pm
Location: Atlanta, GA

Re: Resmed S9 autoset problem

Post by Muse-Inc » Sun Oct 21, 2018 12:02 pm

zoocrewphoto wrote:
Sat Oct 20, 2018 1:07 pm
...machine on the floor. With the filter and cover missing, it sucked insome cat and/or dog hair...
I used to put mine on a step stool, the kind you buy to step up higher in the kitchen to reach upper cabinets. They typically have a textured surface so the S9 won't slide around and it raises it off the floor.
ResMed S9 range 9.8-17, RespCare Hybrid FFM
Never, never, never, never say never.

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: Resmed S9 autoset problem

Post by zoocrewphoto » Sun Oct 21, 2018 2:05 pm

Muse-Inc wrote:
Sun Oct 21, 2018 12:02 pm
zoocrewphoto wrote:
Sat Oct 20, 2018 1:07 pm
...machine on the floor. With the filter and cover missing, it sucked insome cat and/or dog hair...
I used to put mine on a step stool, the kind you buy to step up higher in the kitchen to reach upper cabinets. They typically have a textured surface so the S9 won't slide around and it raises it off the floor.
I am going to suggest something as it really is just sitting on the floor. There are dog beds behind the recliner, so the dogs go back and forth past the machine. Every time we travel together, I usually end up changing her filter as she never does. My machine is in the drawer of an end table next to my bed, so it really doesn't get exposed to much.

I saved the hair I pulled out and put it in a little baggie so that I can show her what was in there. And I will try to check her machine more often. She uses it faithfully, but she doesn't do much to maintain it. She doesn't use the humidifier either, so no need to add water or look at it even. She uses the smart start, so basically she just grabs the mask and puts it on. When she feels she needs a new cushion, she lets me know.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed S9 autoset problem

Post by palerider » Sun Oct 21, 2018 2:08 pm

zoocrewphoto wrote:
Sun Oct 21, 2018 2:05 pm
Muse-Inc wrote:
Sun Oct 21, 2018 12:02 pm
zoocrewphoto wrote:
Sat Oct 20, 2018 1:07 pm
...machine on the floor. With the filter and cover missing, it sucked insome cat and/or dog hair...
I used to put mine on a step stool, the kind you buy to step up higher in the kitchen to reach upper cabinets. They typically have a textured surface so the S9 won't slide around and it raises it off the floor.
I am going to suggest something as it really is just sitting on the floor. There are dog beds behind the recliner, so the dogs go back and forth past the machine. Every time we travel together, I usually end up changing her filter as she never does. My machine is in the drawer of an end table next to my bed, so it really doesn't get exposed to much.

I saved the hair I pulled out and put it in a little baggie so that I can show her what was in there. And I will try to check her machine more often. She uses it faithfully, but she doesn't do much to maintain it. She doesn't use the humidifier either, so no need to add water or look at it even. She uses the smart start, so basically she just grabs the mask and puts it on. When she feels she needs a new cushion, she lets me know.
It would be worth getting a Torx T10 screwdriver and taking it apart, and vacuuming out the inside, if it was run for any length of time there without a filter.

https://imgur.com/a/S0ojG

Let me know, and I'd be happy to walk you through disassembly... it's not difficult, though there are a couple tricky bits... but if you know how to get past those, it's pretty easy.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: Resmed S9 autoset problem

Post by zoocrewphoto » Sun Oct 21, 2018 2:15 pm

palerider wrote:
Sun Oct 21, 2018 2:08 pm
It would be worth getting a Torx T10 screwdriver and taking it apart, and vacuuming out the inside, if it was run for any length of time there without a filter.

https://imgur.com/a/S0ojG

Let me know, and I'd be happy to walk you through disassembly... it's not difficult, though there are a couple tricky bits... but if you know how to get past those, it's pretty easy.
[/quote]

Thanks. That's a really good idea. I was wondering if I got it all. I will check with my dad and see if he already has one of those screwdrivers. He has a magic garage. If it is tool related or can be done with tools, just tell him what you need, and he comes back with it shortly. She's using my main machine currently and happy, so I can want another day to put it back in service. I get home tonight and have tomorrow off.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?