Leaks

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
RedRooster
Posts: 17
Joined: Sat Jun 03, 2017 10:50 am
Location: Omond Beach, FL

Leaks

Post by RedRooster » Tue Oct 16, 2018 10:00 am

How much leak is really acceptable? Mine are usually under 25 Ml but the Res Med report says need to adjust mask. Most leaks appear to be when I turn over & are short in duration. Sleepyhead reports show around 5-6% LL.

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Leaks

Post by Okie bipap » Tue Oct 16, 2018 2:53 pm

If Sleepyhead shows excess leaks 5 to 6 percent of the night, I wouldn't worry about it unless they disturb your sleep. The machine can probably handle leaks up to around 35 lpm, but the large leak flag is set for 24 lpm. I changed our Sleepyhead program to have the large leak flag set to 34 lpm.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

RedRooster
Posts: 17
Joined: Sat Jun 03, 2017 10:50 am
Location: Omond Beach, FL

Re: Leaks

Post by RedRooster » Tue Oct 16, 2018 8:48 pm

Thanks Okie....surprised that you are the only reply.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Leaks

Post by palerider » Tue Oct 16, 2018 8:54 pm

RedRooster wrote:
Tue Oct 16, 2018 8:48 pm
Thanks Okie....surprised that you are the only reply.
Leaks over 24lpm are Resmed's "we're starting to be uncomfortable with this, and at some point we won't be able to differentiate the type of apnea' level. It's a conservative level.

if they're not bothering you, and not much above 24lpm, then ... meh..

In other words, basically what Okie said.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65399
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Leaks

Post by Pugsy » Tue Oct 16, 2018 8:55 pm

I don't even bother looking very closely at the leak graphs unless I spend more than 10% of the night in large leak territory as long as I sleep through any leaks.
And then when I do look closely I check to see just how deep into large leak territory I went and how long was I deep. Unless I spend quite a bit of time above 35 L/min I don't look any further. The machine is fairly accurate up to at least 35 L/min leak and I think sometimes even further than that. If it has problems you will see a lot of unknown apnea flags meaning something happened but due to the leak it couldn't put a name to whatever happened.

So I look at the big picture and not just one number and then I decide just what I might want to do about it...if anything. And 99% of the time it's "do nothing" and the other time it usually means I need to either shrink my headgear or get new headgear. My big leaks are mask movement leaks usually from stretched out headgear.

All this is assuming the leaks aren't disturbing my sleep. Any leak that wakes a person up though...needs to be worked on no matter how big or small because anything that wakes us up is unwanted.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Leaks

Post by palerider » Tue Oct 16, 2018 10:49 pm

RedRooster wrote:
Tue Oct 16, 2018 8:48 pm
Thanks Okie....surprised that you are the only reply.
Sometimes, there's only one answer, because it's the right answer.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.