SleepyHead Bug Reporting?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
iisfaq
Posts: 8
Joined: Wed Sep 05, 2018 2:36 am
Location: Christchurch, New Zealand

SleepyHead Bug Reporting?

Post by iisfaq » Wed Sep 05, 2018 5:02 pm

Hi

I have a major issue with SleepyHead in that it crashes at 96% when importing data. I only just installed it and first time trying to use it.

I tried to file a bug report but the bug reporting site is Under Maintenance - http://bugs.jedimark.net/

I cant find a contact for Jedimark that works (https://jedimark.net/contact - page not found) and I am wondering what is the proper way to submit a bug report in the current environment?

I am using the latest build SleepyHead-1.0.0-beta-2-Snapshot-Win32-OpenGL-20160422.exe

I am running on Windows 10 - Build 1709

The software says that I have a Remstar auto with A Flex.

The software imports the data up to 96% and then stops and does nothing for a few seconds and then crashes with a Access Violation 0xc0000005 error code (Found in Windows Event Viewer) - the application crashes silently - ie no messages are shown.

Faulting application name: SleepyHead.exe, version: 0.0.0.0, time stamp: 0x064d0b28
Faulting module name: SleepyHead.exe, version: 0.0.0.0, time stamp: 0x064d0b28
Exception code: 0xc0000005
Fault offset: 0x0010e1fb
Faulting process ID: 0x19f0
Faulting application start time: 0x01d4454e477f87ae
Faulting application path: C:\Program Files (x86)\SleepyHead\SleepyHead.exe
Faulting module path: C:\Program Files (x86)\SleepyHead\SleepyHead.exe
Report ID: 73c491b7-d4b3-44ce-a498-2b3e504b212e
Faulting package full name:
Faulting package-relative application ID:

Fault bucket 108114012831, type 1
Event Name: APPCRASH
Response: Not available
Cab Id: 0

Problem signature:
P1: SleepyHead.exe
P2: 0.0.0.0
P3: 064d0b28
P4: SleepyHead.exe
P5: 0.0.0.0
P6: 064d0b28
P7: c0000005
P8: 0010e1fb
P9:
P10:

I can provide SD card contents if it would help.

User avatar
Pugsy
Posts: 65196
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: SleepyHead Bug Reporting?

Post by Pugsy » Wed Sep 05, 2018 6:07 pm

Not much you can do and to be honest I wouldn't bother trying to submit a bug report on the SH version you have.
Mark (who came up with SleepyHead) was working on a newer version but seems to have disappeared and not responding to anyone about anything right now.

He won't be doing any bug fixing on the version you have and it remains to be seen when/if this latest version gets released.

Are you wanting just to see your results because you just did a short trial with a machine or are you still using the machine?
The model 561 PR S1 Auto cpap machine was discontinued some years ago and replaced by the DreamStation model line.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

iisfaq
Posts: 8
Joined: Wed Sep 05, 2018 2:36 am
Location: Christchurch, New Zealand

Re: SleepyHead Bug Reporting?

Post by iisfaq » Wed Sep 05, 2018 7:48 pm

I am just doing a trial of the machine at the moment (hired from the Sleep Specialist) and I like the idea of knowing how I do other than a single value I can see on the machine in the morning.

For some reason in New Zealand having private medical insurance does not help with a CPAP machine as they do not cover them at all.

So I am trying to get one though the public hospital system otherwise I will be buying a unit.

I have only been using the machine for 7 days.

Chris

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SleepyHead Bug Reporting?

Post by palerider » Wed Sep 05, 2018 8:22 pm

iisfaq wrote:
Wed Sep 05, 2018 5:02 pm
Hi

I have a major issue with SleepyHead in that it crashes at 96% when importing data. I only just installed it and first time trying to use it.
I can provide SD card contents if it would help.
Frankly, nobody cares about the version you have, anymore, the new version is supposed to be a full rewrite... (though it probably isn't)

If you want to, you can zip up the *entire* contents of your sd card and upload it to https://www.dropbox.com/request/0LalWvEsijRIgX8sdM7H
and I'll see if it'll import here.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
A KLERK
Posts: 53
Joined: Wed Jan 23, 2013 12:42 pm
Location: Raamsdonksveer, The Netherlands
Contact:

Re: SleepyHead Bug Reporting?

Post by A KLERK » Thu Sep 06, 2018 1:22 am

Mark Watkins had not disappeared to everybody. A week ago he told me he needs complete concentration in his work on version 2.0. Het was very disappointed after been carried away in discussions on Discord, while he hoped and expected assistance from friends helping him coding. This way he prevents being frustrated again, knowing it was his own mistake in the first place... Patience please!

In the meantime @iisfaq should realize that AFAIK the CPAP he uses is not (yet?) compatible with SH. I can be wrong though.

Please do realize that a genius like Jedimark needs a different approach than normal slurfers like ourselves, pls be considerate 😍

_________________
MachineMask
Additional Comments: Software: OSCAR of course!
Arie Klerk: Member of the Dutch Apnea Patients Association staff (https://www.apneuvereniging.nl) and proud to be Translations Team Coordinator for OSCAR on ApneaBoard.com
Looking for MORE translators!

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SleepyHead Bug Reporting?

Post by palerider » Thu Sep 06, 2018 1:37 am

A KLERK wrote:
Thu Sep 06, 2018 1:22 am
In the meantime @iisfaq should realize that AFAIK the CPAP he uses is not (yet?) compatible with SH. I can be wrong though.
OP has a system one 60 series auto machine, the ones that came out around 2012, SH has been reading that machine for the last couple of versions. There's something corrupted in his data, and SH isn't handling the error gracefully.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
A KLERK
Posts: 53
Joined: Wed Jan 23, 2013 12:42 pm
Location: Raamsdonksveer, The Netherlands
Contact:

Re: SleepyHead Bug Reporting?

Post by A KLERK » Thu Sep 06, 2018 1:48 am

@iisfaq it looks like the data on your SD card is cortupted. Can you try to use another card in your cpap?

_________________
MachineMask
Additional Comments: Software: OSCAR of course!
Arie Klerk: Member of the Dutch Apnea Patients Association staff (https://www.apneuvereniging.nl) and proud to be Translations Team Coordinator for OSCAR on ApneaBoard.com
Looking for MORE translators!

iisfaq
Posts: 8
Joined: Wed Sep 05, 2018 2:36 am
Location: Christchurch, New Zealand

Re: SleepyHead Bug Reporting?

Post by iisfaq » Thu Sep 06, 2018 1:52 am

Encore can read the same sd card without issue

I have the same problem with a new sd card and just one. 005 file.

Chris

User avatar
Pugsy
Posts: 65196
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: SleepyHead Bug Reporting?

Post by Pugsy » Thu Sep 06, 2018 6:47 am

That latest version of SleepyHead will indeed work just fine with the Respironics model 561.

All I know left to try is what I always try.....get a fresh download and install after removing the old download and installed product.
Just on the off chance that the downloaded files might have been corrupted and thus not installed properly.

It will work on Windows 10.....that is what I am using on this laptop.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: SleepyHead Bug Reporting?

Post by prodigyplace » Thu Sep 06, 2018 7:02 am

Although not bug-free, Mark still has the 1.1.0 unstable test builds on his site at https://sleepyhead.jedimark.net/testing/1.1.0/

I know there were some import bugs fixed, among many other things. It is not ready for normal use, but then you are broken sometimes something is better than nothing.
1.1.0 uses a much newer version of some needed libraries.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

User avatar
Cpapian
Posts: 417
Joined: Sat Jun 24, 2017 2:00 pm
Location: Toronto, Ontario

Re: SleepyHead Bug Reporting?

Post by Cpapian » Thu Sep 06, 2018 9:49 am

Do you have a virus software installed on your computer?

I used to have that problem if my virus software was scanning the SD card at the same time I was trying to upload. Learned to be patient, let virus software do it's thing first them uploaded.

Worth a try.

_________________
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack
Additional Comments: SleepyHead , P10 and Dreamwear FFM, Airfit N20

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SleepyHead Bug Reporting?

Post by palerider » Thu Sep 06, 2018 11:39 am

Cpapian wrote:
Thu Sep 06, 2018 9:49 am
Do you have a virus software installed on your computer?

I used to have that problem if my virus software was scanning the SD card at the same time I was trying to upload. Learned to be patient, let virus software do it's thing first them uploaded.

Worth a try.
It's a SH bug, not a problem with the OPs computer.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

iisfaq
Posts: 8
Joined: Wed Sep 05, 2018 2:36 am
Location: Christchurch, New Zealand

Re: SleepyHead Bug Reporting?

Post by iisfaq » Fri Sep 07, 2018 1:45 am

The 1.1 version loads my data without issue - thanks for the link

Chris

User avatar
zonker
Posts: 11398
Joined: Fri Jun 19, 2015 4:36 pm

Re: SleepyHead Bug Reporting?

Post by zonker » Fri Sep 07, 2018 11:11 am

iisfaq wrote:
Fri Sep 07, 2018 1:45 am
The 1.1 version loads my data without issue - thanks for the link