Bathroom visits - pressure changes

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
snoozysue
Posts: 287
Joined: Sun Oct 17, 2010 2:32 am
Location: Australia

Bathroom visits - pressure changes

Post by snoozysue » Sat Aug 04, 2018 3:23 pm

Hi
Sometimes (not often) I need to go for a quick bathroom trip at night and I turn the machine (Dreamstation Auto) off.
I tend to find that some events occur when I go back to sleep as the machine can't get to the required pressure quickly enough.
If I was to leave the machine running would I find that the pressure has increased greatly when I return.
I don't want this to happen as I already average about 15. The machine is set from 8-18.

Thanks

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Bathroom visits - pressure changes

Post by palerider » Sat Aug 04, 2018 3:30 pm

snoozysue wrote:
Sat Aug 04, 2018 3:23 pm
Hi
Sometimes (not often) I need to go for a quick bathroom trip at night and I turn the machine (Dreamstation Auto) off.
I tend to find that some events occur when I go back to sleep as the machine can't get to the required pressure quickly enough.
If I was to leave the machine running would I find that the pressure has increased greatly when I return.
no.
snoozysue wrote:
Sat Aug 04, 2018 3:23 pm
I don't want this to happen as I already average about 15. The machine is set from 8-18.
8 is too low. Raise it.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

jim22
Posts: 237
Joined: Tue Aug 12, 2014 8:11 pm

Re: Bathroom visits - pressure changes

Post by jim22 » Sat Aug 04, 2018 7:28 pm

Is the ramp feature starting you at a lower pressure? If so, disable it.
Jim

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack
Additional Comments: Heated Humidifer, Non-Heated tube, Oscar
Resmed Airsense 10 Autoset, Swift FX nasal pillows mask

User avatar
snoozysue
Posts: 287
Joined: Sun Oct 17, 2010 2:32 am
Location: Australia

Re: Bathroom visits - pressure changes

Post by snoozysue » Sat Aug 04, 2018 11:12 pm

Thank you for the suggestions.
I don't use the ramp and my low setting is fine for almost all days except the once or twice a month when I wake and get up.
Just curious how long the pressure would remain constant if I disconnect from the machine.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Bathroom visits - pressure changes

Post by palerider » Sat Aug 04, 2018 11:16 pm

snoozysue wrote:
Sat Aug 04, 2018 11:12 pm
Thank you for the suggestions.
I don't use the ramp and my low setting is fine for almost all days except the once or twice a month when I wake and get up.
Just curious how long the pressure would remain constant if I disconnect from the machine.
It can tell you're not breathing, it'll either shut off after a minute (if you've got 'auto off' turned off, or it'll just record a black BND (Breathing Not Detected) spot on the data.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
snoozysue
Posts: 287
Joined: Sun Oct 17, 2010 2:32 am
Location: Australia

Re: Bathroom visits - pressure changes

Post by snoozysue » Sun Aug 05, 2018 12:36 am

Thank you

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: Bathroom visits - pressure changes

Post by zoocrewphoto » Sun Aug 05, 2018 2:23 am

snoozysue wrote:
Sat Aug 04, 2018 11:12 pm
Thank you for the suggestions.
I don't use the ramp and my low setting is fine for almost all days except the once or twice a month when I wake and get up.
Just curious how long the pressure would remain constant if I disconnect from the machine.
If you are averaging 15, then 8 is too low. It will take too much time to go from 8 to 15.

Perhaps if you post the graphs from a typical night, we can help you find a better minimum pressure that will work every night, including those with a bathroom break.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?