Used my new AirSense10 with humidifier and heating tube for 1st time last night. After about 4-5 hours of comfortable use I began to get loud annoying gurgling sound apparently coming from tubing.
Rainout?
Any thoughts on how to remedy?
Thanks

User avatar
Pugsy
Posts: 65119
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AirSense10 w/humidifier

Post by Pugsy » Fri May 11, 2018 3:33 pm

Probably should increase the hose air temp in the heated hose.
It's probably set to "auto" and most of the time auto works fine but sometimes when the ambient bedroom temps are on the cooler side of things the auto temp control doesn't quite get the job done.

Though there are several options available. See posts #2 and #3 in this thread for the various options available for dealing with condensation/rain out in the mask or hose.
viewtopic/t94035/Pugsys-Pointers-3Deali ... -road.html

Easiest fix IMHO...just increase the hose air temp a bit. Sometimes all you need is a couple of degrees.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
greatunclebill
Posts: 1503
Joined: Mon Feb 20, 2012 7:48 pm
Location: L.A. (lower alabama)

Re: AirSense10 w/humidifier

Post by greatunclebill » Fri May 11, 2018 7:52 pm

also, keep the water level at or below the full line. if the reservoir is overfilled even a little bit it will send water in the hose. the way around that is to have the hose come from the machine and go up and come down to you. normally my hose comes off the machine and goes down to the floor and up to the bed. the one time i overfilled it. i hung the hose above the headboard and it was fine.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: myAir, OSCAR. cms-50D+. airsense 10 auto & (2009) remstar plus m series backups
First diagnosed 1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990

User avatar
chartle
Posts: 152
Joined: Mon Mar 12, 2018 7:56 am

Re: AirSense10 w/humidifier

Post by chartle » Sun May 13, 2018 7:48 pm

I over filled my tank once which I thought was impossible because it has sort of an overflow gutter. Well I somehow filled my hose with water but it was intimidate.

Had to drain it.