Plug problem

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Personxyz
Posts: 18
Joined: Tue May 01, 2018 9:51 am

Plug problem

Post by Personxyz » Fri May 11, 2018 7:40 am

I’ve just gotten my machine and was going to give it a test run this afternoon before using it when I go to sleep tonight. Unfortunately that plan is in ruins because my plug adapter doesn’t work.

I purchased my resmed 10 airsense autoset from america (I live in the UK). This meant that the PAP machine came with a US plug so on amazon I bought a two pin plug adapter. I guess I can’t link directly too it but this is what it says on the adapter plug:

BS1363
1A FUSE
BS646
1A 250V
A.C.ONLY

The resmed plug fits in the adapter perfectly but when I plug the adapter into the socket and turn it on there is no power to the PAP machine. At first I thought the machine was broken but I have a shaving socket in my bathroom which is a 2 pin socket. I plugged the resmed plug into there and it instantly turned on so it must be the adapter.

I don’t know if I simply bought a faulty adapter or if I’ve overlooked something and ended up buying the wrong adapter (I know next to nothing about electronics). I thought I would ask on the forums before I buy another adapter because if I am overlooking something I’ll likely just buy another wrong adapter, delaying my treatment and wasting money.

Is there anything I should know before I buy another adapter or do you guys just think I was unlucky and got a broken product?

User avatar
Nick Danger
Posts: 621
Joined: Mon Jun 09, 2014 3:13 pm

Re: Plug problem

Post by Nick Danger » Fri May 11, 2018 7:53 am

According to Amazon, that is a plug for a toothbrush or shaver and has a 1 amp fuse. If the CPAP tries to draw more than 1 amp, it will blow the fuse and you won't get any power. There are several people on here who are more knowledgeable about electrical stuff than I - they will likely post later today.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: APAP mode, minimum pressure = 9. No ramp, EPR = 3, medium. Soft cervical collar. Sleepyhead software.

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: Plug problem

Post by prodigyplace » Fri May 11, 2018 8:02 am

Nick Danger wrote:
Fri May 11, 2018 7:53 am
According to Amazon, that is a plug for a toothbrush or shaver and has a 1 amp fuse. If the CPAP tries to draw more than 1 amp, it will blow the fuse and you won't get any power. There are several people on here who are more knowledgeable about electrical stuff than I - they will likely post later today.
I think you may be correct. Here is what the specification says on our sponsor's site.
AC Input Range - 90W Power Supply: 100 - 240V, 50 - 60Hz, 1.0 - 1.5A, Class II
from https://www.cpap.com/productpage/resmed ... #specs-tab

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

Woody
Posts: 591
Joined: Sat Mar 10, 2007 7:59 pm

Re: Plug problem

Post by Woody » Fri May 11, 2018 8:09 am

The cpap will run off either 110v 0r 220v so you don't need a converter
just a much cheaper adapter plug.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: SleapyHead software

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: Plug problem

Post by prodigyplace » Fri May 11, 2018 8:12 am

Woody wrote:
Fri May 11, 2018 8:09 am
The cpap will run off either 110v 0r 220v so you don't need a converter
just a much cheaper adapter plug.
Good catch! Either an adapter plug or a different power cord for the power brick.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

Personxyz
Posts: 18
Joined: Tue May 01, 2018 9:51 am

Re: Plug problem

Post by Personxyz » Fri May 11, 2018 9:28 am

Thank you to everyone for the replies. I've decided to get a different power cord. In case anyone else has this problem I found the new power cord by searching amazon using these keywords "uk 3 pin c7".

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: Plug problem

Post by Jas_williams » Fri May 11, 2018 9:30 am

Personxyz wrote:
Fri May 11, 2018 7:40 am
I’ve just gotten my machine and was going to give it a test run this afternoon before using it when I go to sleep tonight. Unfortunately that plan is in ruins because my plug adapter doesn’t work.

I purchased my resmed 10 airsense autoset from america (I live in the UK). This meant that the PAP machine came with a US plug so on amazon I bought a two pin plug adapter. I guess I can’t link directly too it but this is what it says on the adapter plug:

BS1363
1A FUSE
BS646
1A 250V
A.C.ONLY

The resmed plug fits in the adapter perfectly but when I plug the adapter into the socket and turn it on there is no power to the PAP machine. At first I thought the machine was broken but I have a shaving socket in my bathroom which is a 2 pin socket. I plugged the resmed plug into there and it instantly turned on so it must be the adapter.

I don’t know if I simply bought a faulty adapter or if I’ve overlooked something and ended up buying the wrong adapter (I know next to nothing about electronics). I thought I would ask on the forums before I buy another adapter because if I am overlooking something I’ll likely just buy another wrong adapter, delaying my treatment and wasting money.

Is there anything I should know before I buy another adapter or do you guys just think I was unlucky and got a broken product?


Just go To Tesco or somewhere and buy a normal three pin plug cut the US two pin off and wire the blue cable to the neutral and brown to live. Or get a figure 8 cable

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: Plug problem

Post by Jas_williams » Fri May 11, 2018 9:31 am

Ah you found the correct cable they are not expensive..

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Plug problem

Post by palerider » Fri May 11, 2018 11:05 am

Personxyz wrote:
Fri May 11, 2018 9:28 am
Thank you to everyone for the replies. I've decided to get a different power cord. In case anyone else has this problem I found the new power cord by searching amazon using these keywords "uk 3 pin c7".
This is the proper answer. The reason that the power bricks use a plug in cord from them to the wall is so that the manufacturers can make *one* power brick for the entire world, and just use the appropriate short cord for whatever country you're in.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Plug problem

Post by palerider » Fri May 11, 2018 11:07 am

Jas_williams wrote:
Fri May 11, 2018 9:30 am
cut the US two pin off
No cutting needed, just get the right cord.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: Plug problem

Post by prodigyplace » Fri May 11, 2018 11:11 am

palerider wrote:
Fri May 11, 2018 11:07 am
No cutting needed, just get the right cord.
No criticism needed, Just read where they corrected themselves.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Plug problem

Post by palerider » Fri May 11, 2018 11:13 am

. wrote:
Fri May 11, 2018 11:11 am

No criticism needed, Just read where they corrected themselves.
No intelligence exhibited, just making snarky posts.

Typical

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: Plug problem

Post by Jas_williams » Fri May 11, 2018 12:03 pm

palerider wrote:
Fri May 11, 2018 11:07 am
Jas_williams wrote:
Fri May 11, 2018 9:30 am
cut the US two pin off
No cutting needed, just get the right cord.
All depends if the user wants to use the machine today or in a few days time.... 3 Pin plugs and fuses are readily available Figure 8 cables are not and will take a 1-2 days to come.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask