Help With Hypos
Help With Hypos
Hello:
Attached charts are typical. I won appreciate any help with hypos. Thanks.
Attached charts are typical. I won appreciate any help with hypos. Thanks.
- Attachments
-
- Screen Shot 2018-05-08 at 7.30.44 AM.png (232.03 KiB) Viewed 2068 times
-
- Screen Shot 2018-05-08 at 7.30.22 AM.png (202.23 KiB) Viewed 2068 times
-
- Screen Shot 2018-05-08 at 7.29.52 AM.png (230.35 KiB) Viewed 2068 times
Re: Help With Hypos
Why are you turning the machine off so many times? You could try turning off the ramp if you are using it and/or increase the min pressure to 10. Overall, it really doesn't look that bad.
Re: Help With Hypos
Thank you LSAT. Not using ramp. I appreciate the raise to 10 suggestion.
- raisedfist
- Posts: 1176
- Joined: Wed Jun 15, 2016 7:21 am
Re: Help With Hypos
A bit more min pressure could help - you also have very small amounts of flow limitation and snoring. I'd try 9.5 or 10 cm h2o as your min. Your sleep seems rather fragmented...are you having problems wearing the mask?
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12
Re: Help With Hypos
I suggest raising the high end too. I have my high end at 20, but it never actually goes there.
Re: Help With Hypos
There is no reason to increase the OP max, he is not ever approaching the set max. Increasing the minimum may help as some of the hypopneas occurs when the machine is at it's relatively lower pressure, due to restarting.
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very
Re: Help With Hypos
All this time on the forum making snarky cutting comments, and you still haven't read any of the MANY "how to post your data" threads of information. :shaking head:
See below.
See below.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Help With Hypos
Pale Rider. Thanks for your help I will check out your advice. Appreciate your concern about correct posting. It will help me get meaningful insight into my charts. Bless you. Have a good afternoon.
Re: Help With Hypos
Pale Rider. You were right. Your link was very helpful. It is good that you are willing to help educate less experienced forum users. Keep
up the good work.
up the good work.
Re: Help With Hypos
Here's a spot of advice, take a lesson from Gator5000e, who obviously spent a little time on the forum *READING* and *LEARNING*... perfect data post right out of the gate.
You? you were obviously too busy making cutting sarcastic troll posts to learn anything at all.
Try and be more like Gator5000e.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Help With Hypos
Thanks Pale Rider. Is this graph better?
Re: Help With Hypos
Does it look like the examples? Does it look like gator5000e's screenshot? Do you have redundant things on the page causing the charts to be scrunched up so they're too small to be meaningful?
(no, no, yes)
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
