90% pressure level

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
jbgilm
Posts: 28
Joined: Mon Jan 25, 2016 9:40 am

90% pressure level

Post by jbgilm » Fri Mar 23, 2018 4:10 pm

If my cpap tells me that the 90% pressure level is the same as the maximum pressure, it seems that the maximum pressure should be higher. Is that true? Is there any guideline for the proper relationship between the 90% pressure reading and the max? Thanks. This website is very helpful!

User avatar
MamaGord
Posts: 38
Joined: Thu Mar 15, 2018 10:05 pm

Re: 90% pressure level

Post by MamaGord » Fri Mar 23, 2018 4:14 pm

I am also curious about this. My max pressure is 12 and my 90% the last two nights had been right under 12.
Emily ~ happily married mom to 4
Using CPAP since 3.21.18 pressure 12.5-18

jbgilm
Posts: 28
Joined: Mon Jan 25, 2016 9:40 am

Re: 90% pressure level

Post by jbgilm » Fri Mar 23, 2018 4:16 pm

Following up on this I read elsewhere on the interweb the following:

"Most people strive to set their "APAP" (not CPAP) pressures to 2-4 cm's below the 90% and 2 or 3 above the 90% in order to accommodate the unknowns."

I don't understand that. Can someone explain?

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: 90% pressure level

Post by Okie bipap » Fri Mar 23, 2018 4:37 pm

The 90% value is the point where your pressure was that value or less for 90% of the time. You need to look at sleepyhead to see how much of your time you were bumping up against the upper limit. If it is only a few minutes a night, then leave it where it is. If you are bumping up against the upper limit a lot of the night, then you probably need to raise the upper limit. You can leave the upper limit set to the highest possible value, and the machine will not go any higher than it need to. My upper limit is set to 25, but my pressure seldom exceeds 16 cm pressure.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: 90% pressure level

Post by palerider » Fri Mar 23, 2018 5:44 pm

jbgilm wrote:
Fri Mar 23, 2018 4:16 pm
Following up on this I read elsewhere on the interweb the following:

"Most people strive to set their "APAP" (not CPAP) pressures to 2-4 cm's below the 90% and 2 or 3 above the 90% in order to accommodate the unknowns."

I don't understand that. Can someone explain?
Only the clueless ones.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: 90% pressure level

Post by palerider » Fri Mar 23, 2018 5:46 pm

jbgilm wrote:
Fri Mar 23, 2018 4:10 pm
If my cpap tells me that the 90% pressure level is the same as the maximum pressure, it seems that the maximum pressure should be higher. Is that true?
No, that's not true.
jbgilm wrote:
Fri Mar 23, 2018 4:10 pm
Is there any guideline for the proper relationship between the 90% pressure reading and the max?
Read this:
http://adventures-in-hosehead-land.blog ... de-to.html

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

D.H.
Posts: 3533
Joined: Sun Sep 04, 2016 7:07 pm

Re: 90% pressure level

Post by D.H. » Fri Mar 23, 2018 6:19 pm

To the OP,

Is the 90% the same as the max setting (or "ceiling")? If that's the case, it can't go any higher! That would mean that you're spending at least 10% of the time at max. If that's correct, try raising the ceiling and see what happens.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: 90% pressure level

Post by Julie » Fri Mar 23, 2018 6:34 pm

.
Last edited by Julie on Fri Mar 23, 2018 10:16 pm, edited 1 time in total.

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: 90% pressure level

Post by TedVPAP » Fri Mar 23, 2018 9:10 pm

jbgilm wrote:
Fri Mar 23, 2018 4:16 pm
Following up on this I read elsewhere on the interweb the following:

"Most people strive to set their "APAP" (not CPAP) pressures to 2-4 cm's below the 90% and 2 or 3 above the 90% in order to accommodate the unknowns."

I don't understand that. Can someone explain?
Yes this is good general advice. Certainly not perfect advice for everyone. It is best to review your data and make changes.
As others have said, the 90% pressure is the pressure that you were at or less 90% of the time. That means that 10% of the time you were above. When setting an APAP, set the minimum below this value. Some are able to set the minimum limit to much less, but others do better when they set the minimum closer to where the need is. When I set my minimum limit too low, the pressure can settle too low at times and can not increase fast enough to prevent many of my events. My min setting is about 2 cm less than the 90% pressure.

Setting the maximum level above the 90% pressure enables the machine to go higher when needed. Unless you have a good reason to limit the maximum pressure, set it higher than needed.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

User avatar
RogerSC
Posts: 1912
Joined: Sat Jul 21, 2012 1:11 pm

Re: 90% pressure level

Post by RogerSC » Sat Mar 24, 2018 11:07 am

TedVPAP wrote:
Fri Mar 23, 2018 9:10 pm
jbgilm wrote:
Fri Mar 23, 2018 4:16 pm
Following up on this I read elsewhere on the interweb the following:

"Most people strive to set their "APAP" (not CPAP) pressures to 2-4 cm's below the 90% and 2 or 3 above the 90% in order to accommodate the unknowns."

I don't understand that. Can someone explain?
Yes this is good general advice. Certainly not perfect advice for everyone. It is best to review your data and make changes.
As others have said, the 90% pressure is the pressure that you were at or less 90% of the time. That means that 10% of the time you were above. When setting an APAP, set the minimum below this value. Some are able to set the minimum limit to much less, but others do better when they set the minimum closer to where the need is. When I set my minimum limit too low, the pressure can settle too low at times and can not increase fast enough to prevent many of my events. My min setting is about 2 cm less than the 90% pressure.

Setting the maximum level above the 90% pressure enables the machine to go higher when needed. Unless you have a good reason to limit the maximum pressure, set it higher than needed.
That's a starting point...ideally, as you go along you should be looking at your data, and tuning your pressure range to optimize things if you feel that your therapy could be better. Sometimes raising your minimum pressure can help to make your therapy work better, so that the machine doesn't take too much time to get up to the pressure needed to block an apnea episode. Or raising your top pressure if you see your pressure hanging out at the top limit too much. And so on. But doing this depends on your monitoring your sleep data, and asking questions as you develop a feel for what changing various parameters on your machine does.

Your needs can change over time, and you can accommodate that by tracking your sleep data. Rather than just setting an arbitrary range at the beginning and calling it a day *smile*.

ajack
Posts: 977
Joined: Thu Mar 02, 2017 2:54 am
Location: australia

Re: 90% pressure level

Post by ajack » Sun Mar 25, 2018 10:18 pm

jbgilm wrote:
Fri Mar 23, 2018 4:10 pm
If my cpap tells me that the 90% pressure level is the same as the maximum pressure, it seems that the maximum pressure should be higher. Is that true? Is there any guideline for the proper relationship between the 90% pressure reading and the max? Thanks. This website is very helpful!
if the 90% and max is the same, you need to raise the max pressure. You are needing more max pressure, to give the machine room to move.
If you go 1 or 2 below and 2 above the 90-95% number, you will be in the ball park. It works for most people, 90/95% is the number they use to set a fixed pressure when auto titrating. Put up some sleepyhead charts for fine tuning, if you wish to.

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 ST-A iVAPS and adapt ASV

D.H.
Posts: 3533
Joined: Sun Sep 04, 2016 7:07 pm

Re: 90% pressure level

Post by D.H. » Mon Mar 26, 2018 1:37 pm

Actually, I have my max at 20, which is the capacity. It has never gone above 17, so I guess 17 would behave the same as 20 in my case.

I would try setting the therapy max to the machine max. If you experience an increase in "clear airway" events, increased dryness of the sinuses, pain/discomfort from the increased pressure, or increased restlessness/awakenings, then you would want to back down.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max