Are these oxygen concentrator generators any good?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Peter C
Posts: 15
Joined: Tue Mar 13, 2018 3:55 pm

Are these oxygen concentrator generators any good?

Post by Peter C » Tue Mar 20, 2018 11:49 am

I don't suffer from OSA, but Autonomic Instability giving me a low blood oxygen content whilst asleep.
Whilst I wait for the NHS to do something, (only been a year so far), is it worth getting one of these oxygen generators?, how do they work?, do they take air from your room and take the oxygen from it? Is there any danger from upsetting the balance between oxygen and carbon dioxide that the Autonomic Nervous System detects and reacts to in the blood?
Anybody else use them?
The symptoms I am trying to eliviate are a constant fatigue, feeling I am drunk and dizziness.
Thanks.Image

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Are these oxygen concentrator generators any good?

Post by palerider » Tue Mar 20, 2018 1:38 pm

Peter C wrote:
Tue Mar 20, 2018 11:49 am
I don't suffer from OSA, but Autonomic Instability giving me a low blood oxygen content whilst asleep.
Whilst I wait for the NHS to do something, (only been a year so far), is it worth getting one of these oxygen generators?, how do they work?, do they take air from your room and take the oxygen from it?
They're oxygen *concentrators* not *generators*.
What they do is filter the nitrogen out of the air, nitrogen is around 78% of normal air. They suck up air through a compressor, pump it into a sieve bed that absorbs the nitrogen, then they vent the oxygen into a holding tank and flush out the nitrogen.
Peter C wrote:
Tue Mar 20, 2018 11:49 am
Is there any danger from upsetting the balance between oxygen and carbon dioxide that the Autonomic Nervous System detects and reacts to in the blood?
That's not how it works, your respiration cycle is driven by CO2 in the blood, not oxygen, which is why you don't have a lot of effects from low oxygen... or oxygen supplementation.
Peter C wrote:
Tue Mar 20, 2018 11:49 am
Anybody else use them?
The symptoms I am trying to eliviate are a constant fatigue, feeling I am drunk and dizziness.
Lots of people use them.
Maybe you DO suffer from OSA.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Janknitz
Posts: 8499
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Are these oxygen concentrator generators any good?

Post by Janknitz » Tue Mar 20, 2018 2:37 pm

Good answer from Palerider, just to add that if you do have sleep apnea, the oxygen supplied by the concentrator via nasal canula will NOT keep your airway open. Have you been tested for sleep apnea?
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

Peter C
Posts: 15
Joined: Tue Mar 13, 2018 3:55 pm

Re: Are these oxygen concentrator generators any good?

Post by Peter C » Tue Mar 20, 2018 4:19 pm

I get low oxygen when I am awake as well, the hospital machine was constantly bleeping at 87% whilst I was awake, so I dont think its OSA, I literally forget to breathe sometimes, especially when talking of relaxing. I have seen my oxygen at 99% and lung function tests are good.
Not every night is as low as 85%, more often it is averaging 90%, maybe a quarter of the time its 85. So of course when I had a sleep apnea clinic appraisal, it was on pne of the 90% average days, when I showed the doctor it was sometimes in the area he said they would have been concerned about, he said I probably had a sweaty finger and so dismissed me. This is the NHS these days, this is why I am looking at treating myself, the NHS couldnt find I had Ventricular Tachycardia for two years until I bought my own ECG machine and showed them.

User avatar
Julie
Posts: 20032
Joined: Tue Feb 28, 2006 12:58 pm

Re: Are these oxygen concentrator generators any good?

Post by Julie » Tue Mar 20, 2018 4:43 pm

Well, the desats are actually just below the cutoff for low 02 line - 88% - though of course that''s not ideal, but what you experience is not horrendous either compared to some who desat to low 60's at times without Cpap. But do try and get tested for apnea anyway - are there no 'home' tests available there (if she won't schedule you for a lab one)?

And you said you have times when you just don't breathe - it would be important to know if that happens when you sleep because it could indicate 'clear airway' or central events (centrals happen when your brain does not signal you to breathe) and there are special machines that treat that.

Peter C
Posts: 15
Joined: Tue Mar 13, 2018 3:55 pm

Re: Are these oxygen concentrator generators any good?

Post by Peter C » Tue Mar 20, 2018 5:18 pm

Just lying in bed on the computer now and my sats are 90%.
This all started the day the put me on beta blockers for my Ventricular Tachycardia, I described feeling as if my lungs no longer worked. Turned out I was allergic to them so they took me off a year ago, but the symptoms remain. I dont feel hunger much now either. So I think its a Autonomic Nervous System, brain not functioning thing rather than an airways blocking thing.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Are these oxygen concentrator generators any good?

Post by palerider » Tue Mar 20, 2018 5:42 pm

Peter C wrote:
Tue Mar 20, 2018 5:18 pm
So I think its a Autonomic Nervous System, brain not functioning thing rather than an airways blocking thing.
Which doesn't mean that you couldn't benefit from a little help breathing at night.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Peter C
Posts: 15
Joined: Tue Mar 13, 2018 3:55 pm

Re: Are these oxygen concentrator generators any good?

Post by Peter C » Fri Mar 23, 2018 3:55 pm

Well I bought one, an a Dedakj https://www.amazon.co.uk/Vogvigo-Portab ... op?ie=UTF8 I had seen it top of a review of the ten best machines. However, it states it produces 90% oxygen, I am using the tube that goes up each nostril, yet my blood oxygen only goes up between 1 and 3 percent at the most. How much would you expect one of these machines to improve your sats by please?

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Are these oxygen concentrator generators any good?

Post by raisedfist » Fri Mar 23, 2018 4:42 pm

The flow meter goes from 1 - 6 liters/min. What setting is yours at?

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

Peter C
Posts: 15
Joined: Tue Mar 13, 2018 3:55 pm

Re: Are these oxygen concentrator generators any good?

Post by Peter C » Fri Mar 23, 2018 4:54 pm

raisedfist wrote:
Fri Mar 23, 2018 4:42 pm
The flow meter goes from 1 - 6 liters/min. What setting is yours at?
Seems the machine has a fixed amount of oxygen, (even though the manual claims both flow rate and oxygen concentration are adjustable, but there is only the dial for flow rate). So, the 90% oxygen appears when the flow rate is at the minimum, increasing the flow rate dilutes the oxygen concentration. Its fine for me as I get the max concentration of oxygen without a gale blowing up my nose.

So I have it just after one, but I did try all the way around to six liters and found it made no difference at all to my blood oxygen.

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Are these oxygen concentrator generators any good?

Post by raisedfist » Fri Mar 23, 2018 4:59 pm

Peter C wrote:
Fri Mar 23, 2018 4:54 pm
raisedfist wrote:
Fri Mar 23, 2018 4:42 pm
The flow meter goes from 1 - 6 liters/min. What setting is yours at?
Seems the machine has a fixed amount of oxygen, (even though the manual claims both flow rate and oxygen concentration are adjustable, but there is only the dial for flow rate). So, the 90% oxygen appears when the flow rate is at the minimum, increasing the flow rate dilutes the oxygen concentration. Its fine for me as I get the max concentration of oxygen without a gale blowing up my nose.

So I have it just after one, but I did try all the way around to six liters and found it made no difference at all to my blood oxygen.
Sounds like the machine you got is a cheapo unfortunately...

I would look into getting a Respironics Everflo Q machine if you can find one. The amount of oxygen purity stays the same with all flow rates.

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Are these oxygen concentrator generators any good?

Post by palerider » Fri Mar 23, 2018 5:43 pm

Peter C wrote:
Fri Mar 23, 2018 4:54 pm
So, the 90% oxygen appears when the flow rate is at the minimum, increasing the flow rate dilutes the oxygen concentration.
Not unless it's broken.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Matt00926
Posts: 266
Joined: Sat Apr 25, 2015 9:15 am

Re: Are these oxygen concentrator generators any good?

Post by Matt00926 » Fri Mar 23, 2018 6:22 pm

palerider wrote:
Fri Mar 23, 2018 5:43 pm
Peter C wrote:
Fri Mar 23, 2018 4:54 pm
So, the 90% oxygen appears when the flow rate is at the minimum, increasing the flow rate dilutes the oxygen concentration.
Not unless it's broken.
Oddly, looking at the photos of the unit - for some reason the percentage does seem to actually change when you adjust the l/m flow. I've never heard of such a thing. But I imagine that's what you get for less than $500 USD? The product description is pretty misleading...
Machine: ResMed AirCurve 10 ST
Mask: Fisher & Paykel Simplus FFM

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Are these oxygen concentrator generators any good?

Post by palerider » Fri Mar 23, 2018 6:29 pm

Matt00926 wrote:
Fri Mar 23, 2018 6:22 pm
palerider wrote:
Fri Mar 23, 2018 5:43 pm
Peter C wrote:
Fri Mar 23, 2018 4:54 pm
So, the 90% oxygen appears when the flow rate is at the minimum, increasing the flow rate dilutes the oxygen concentration.
Not unless it's broken.
Oddly, looking at the photos of the unit - for some reason the percentage does seem to actually change when you adjust the l/m flow. I've never heard of such a thing. But I imagine that's what you get for less than $500 USD? The product description is pretty misleading...
Ok, not broken, just junk.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Peter C
Posts: 15
Joined: Tue Mar 13, 2018 3:55 pm

Re: Are these oxygen concentrator generators any good?

Post by Peter C » Sat Mar 24, 2018 5:15 am

So here is a oximeter report from my first might with it, deepest sleep I have had for a long time and as you can see, the basal average of 91.8% for sleeping is fine and the best I have ever seen it since I had an oximeter.

I was thinking I could send it back and pay 3 times as much for one that delivers 90% oxygen at faster rates than 1lt a minute, but on consideration, this 90% I get at 1lt will be as good as 90% at 1lt a minute from a better machine, and as this level is enough of an improvement, would I even use the faster flow as I dont like the feeling of a faster rate of air being blown up my nose anyway.

Would still like to hear figures from other people what their Blood oxygen % improvement is when using a flow of just 1lt a minute, if it is vastly better than mine, I still might be tempted to change. Image