AirCurve ASV Auto/Aerophagia

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Zzzzzzzzzzz...
Posts: 331
Joined: Sun Mar 04, 2012 9:19 am

AirCurve ASV Auto/Aerophagia

Post by Zzzzzzzzzzz... » Tue Mar 06, 2018 7:24 am

Hi all,

Been awhile. So... I've been having luck with my sleep therapy, however not so much with aerophagia... please look over my data and see if there's anything you might tweak. Open
to any and all suggestions. Settings, etc. Wonder if I could get by with positive results with less pressure (if that's the problem). Thanks so much.

-Z
Attachments
Screen Shot 2018-03-06 at 7.17.28 AM.png
Screen Shot 2018-03-06 at 7.17.28 AM.png (246.75 KiB) Viewed 5494 times

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AirCurve ASV Auto/Aerophagia

Post by Pugsy » Tue Mar 06, 2018 11:05 am

It’s tough when you need the higher bursts for treating the centrals but the higher burst causes the aerophagia issues. You really can’t lower your EPAP because it is already pretty much as low as it will go and the machine only goes up when it really thinks it needs to.

I would suggest talking with your doctor about maybe limiting that big burst when it is responding to the centrals. I am not sure if it would even help the aerophagia all that much and could mean sub optimal treatment of the centrals.

EPAP is the most likely offender but you settings are already pretty much at the minimum.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
ChicagoGranny
Posts: 15534
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: AirCurve ASV Auto/Aerophagia

Post by ChicagoGranny » Tue Mar 06, 2018 11:26 am

Zzzzzzzzzzz... wrote: ↑
Tue Mar 06, 2018 7:24 am
aerophagia
This can be tough to resolve. You might want to check out some of the tips in the wiki article - wiki/index.php/Aerophagia .

User avatar
Zzzzzzzzzzz...
Posts: 331
Joined: Sun Mar 04, 2012 9:19 am

Re: AirCurve ASV Auto/Aerophagia

Post by Zzzzzzzzzzz... » Tue Mar 06, 2018 1:31 pm

Thank you both! I'll do some more research... certainly has become a problem... and just when treatment has started to work.

User avatar
Mr Bill
Posts: 532
Joined: Fri Feb 18, 2011 11:56 pm
Location: Grand Junction, CO

Re: AirCurve ASV Auto/Aerophagia

Post by Mr Bill » Tue Mar 06, 2018 1:46 pm

My machine used to blast me pretty powerfully. But over time, it seems to have learned me because it hardly ever happens now. Come to think of it, I think we reduced the backup rate. My average is 18.2 BPM but we reduced the forced backup rate to 8 BPM.

I had a new sleep study done in 2017 and it reconfirms all the 2009 findings. The one thing the Dr thought was that the backup rate was low. But I remember what a struggle it was to have it blast me just as I was falling asleep. My local DME assured me that a forced backup rate 8 BPM would not croak me and its been mostly smooth sailing since.
EPAP min=6, EPAP max=15, PS min=3, PS max=12, Max Pressure=30, Backup Rate=8 bpm, Flex=0, Rise Time=1,
90% EPAP=7.0, Avg PS=4.0, Avg bpm 18.3, Avg Min vent 9.2 Lpm, Avg CA/OA/H/AHI = 0.1/0.1/2.1/2.3 ... updated 02/17/12

User avatar
Zzzzzzzzzzz...
Posts: 331
Joined: Sun Mar 04, 2012 9:19 am

Re: AirCurve ASV Auto/Aerophagia

Post by Zzzzzzzzzzz... » Tue Mar 06, 2018 1:50 pm

Just got a bit of great advice from my Sleep Tech/DME. She said that in most cases the number cause of aerophagia is dry mouth/nose. Both of which cause excess swallowing in order to alleviate the dryness. I know I do this often. It is NOT caused by leaks... in my case anyway. Changing humidifier and temp settings on the machine just might do the trick. I had mine set to Auto. Something she never recommends. SO... I'm starting at level 5 on the Humidifier and 82 degrees for the temp
and dialing them up +1 a day and see what happens. I'll post back the results. Fingers crossed.

Z

klv329
Posts: 500
Joined: Mon Aug 31, 2015 6:43 pm
Location: Bradenton, Florida

Re: AirCurve ASV Auto/Aerophagia

Post by klv329 » Tue Mar 06, 2018 4:15 pm

Best of luck. My DME tech mentioned dry mouth as a cause of mouth breathing about 15 years ago.

I have to keep my acid reflux under control to limit tummy air. Diet change and nexium and incline bed.

There were times when I'd wake up and it was like my stomach was a third lung!

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Additional Comments: Resmed Aircurve 10 ASVAUTO Min Epap 10.4, Max Epap 11.6, PS 1.6-12.0

User avatar
Zzzzzzzzzzz...
Posts: 331
Joined: Sun Mar 04, 2012 9:19 am

Re: AirCurve ASV Auto/Aerophagia

Post by Zzzzzzzzzzz... » Tue Mar 06, 2018 9:32 pm

klv329 wrote: ↑
Tue Mar 06, 2018 4:15 pm
Best of luck. My DME tech mentioned dry mouth as a cause of mouth breathing about 15 years ago.

I have to keep my acid reflux under control to limit tummy air. Diet change and nexium and incline bed.

There were times when I'd wake up and it was like my stomach was a third lung!
Thanks.

ajack
Posts: 977
Joined: Thu Mar 02, 2017 2:54 am
Location: australia

Re: AirCurve ASV Auto/Aerophagia

Post by ajack » Wed Mar 07, 2018 3:15 am

First you need to fix your leaks, you are wasting your time with leaks and ASV. The machine will max out trying to get treatment pressure. If you can't get the nasal to work, go full face
When you have done that, post another chart and if you want to, pm me if I don't notice
For now you can also try min epap 5.6 and max ps 11, If this chart is your average 95% pressures. That should help with the gas for now, it isn't for long term, but may get you over the hump.

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 ST-A iVAPS and adapt ASV

User avatar
Zzzzzzzzzzz...
Posts: 331
Joined: Sun Mar 04, 2012 9:19 am

Re: AirCurve ASV Auto/Aerophagia

Post by Zzzzzzzzzzz... » Fri Mar 09, 2018 7:06 pm

Just circling back on my results... changing humidifier and temp settings on the machine has shown a MARKED improvement, if not complete resolution.
I have little to no swallowed air. First night, it was nominal, then I raised each setting by 1 for the next two nights... and voila.
It's gone. Easy fix.

Thanks for all the feedback.

Z