AHI very high and BiPAP doesnt help

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
highapnea
Posts: 4
Joined: Fri Feb 02, 2018 1:01 pm

AHI very high and BiPAP doesnt help

Post by highapnea » Fri Feb 02, 2018 1:05 pm

Hi

My AHI was 80 while diagnosed for sleep apnea in 2015. Doctor advised BiPAP and brought down the AHI to 7-10 with 7 hrs use .. But I still struggle during the day with watery , burning eyes. Any suggestions ? Does this reduced AHi shown are really real ?

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: AHI very high and BiPAP doesnt help

Post by Julie » Fri Feb 02, 2018 1:36 pm

What mask do you use? It's probably allowing air to get into your eyes and there are ways to fix that, but we need more info to help. Do try artificial tears (drugstore) for now though.

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: AHI very high and BiPAP doesnt help

Post by Goofproof » Fri Feb 02, 2018 1:45 pm

Your treatment isn't as good as it should be, AHI should be lower under 5, but really under two. Get the software and get-er-done! Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: AHI very high and BiPAP doesnt help

Post by TedVPAP » Fri Feb 02, 2018 7:16 pm

Goofproof wrote:Your treatment isn't as good as it should be, AHI should be lower under 5, but really under two. Get the software and get-er-done! Jim
Agree.
OP - see the three links below

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

User avatar
Mikebear
Posts: 132
Joined: Sat Dec 16, 2017 3:58 pm
Location: Mid-Michigan

Re: AHI very high and BiPAP doesnt help

Post by Mikebear » Fri Feb 02, 2018 8:19 pm

highapnea wrote:Hi

My AHI was 80 while diagnosed for sleep apnea in 2015. Doctor advised BiPAP and brought down the AHI to 7-10 with 7 hrs use .. But I still struggle during the day with watery , burning eyes. Any suggestions ? Does this reduced AHi shown are really real ?
Buy one of these: https://www.walmart.com/ip/Spa-Bella-Sl ... /540361703

Wear it over your eyes when you put on your cpap mask. IF the mask (FFM? likely) is leaking air into your eyeballs due to a bad seal, this will stop it from drying out your eyeballs and you'll know for sure.

It's real cheap so can't hurt either way, and comes in handy.

Also as others have said, post your info and data to see if things can be tweaked even better on your cpap therapy

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Pressures = 10 CmH20 - 13 CmH20 EPR 2. COPD, Asthma, 30lbs overweight. Airsense 10 AutoSet

highapnea
Posts: 4
Joined: Fri Feb 02, 2018 1:01 pm

Re: AHI very high and BiPAP doesnt help

Post by highapnea » Fri Feb 16, 2018 11:20 am

Julie wrote:
Fri Feb 02, 2018 1:36 pm
What mask do you use? It's probably allowing air to get into your eyes and there are ways to fix that, but we need more info to help. Do try artificial tears (drugstore) for now though.

I use a full mask covering nose and mouth. I most of the time breath through my mouth.
I also don't think my mask is leaking .. its air tight always or most of the time .. and I notice it in my app too. I agree to another comment where if my AHI Is less than 4 , it might help me better ..

highapnea
Posts: 4
Joined: Fri Feb 02, 2018 1:01 pm

Re: AHI very high and BiPAP doesnt help

Post by highapnea » Fri Feb 16, 2018 11:23 am

TedVPAP wrote:
Fri Feb 02, 2018 7:16 pm
Goofproof wrote:Your treatment isn't as good as it should be, AHI should be lower under 5, but really under two. Get the software and get-er-done! Jim
Agree.
OP - see the three links below
I will meet the doctor and request for more pressure on my BiPAP .. At present it is 18/20 if I am correct.

highapnea
Posts: 4
Joined: Fri Feb 02, 2018 1:01 pm

Re: AHI very high and BiPAP doesnt help

Post by highapnea » Fri Feb 16, 2018 11:26 am

Goofproof wrote:
Fri Feb 02, 2018 1:45 pm
Your treatment isn't as good as it should be, AHI should be lower under 5, but really under two. Get the software and get-er-done! Jim
Any suggestions for drooling issue ? Whenever I sleep sideways, I drool and wet my face inside the mask. This wakes me up every time to remove mask and clean face and get back to sleep.. This definitely disturbs the sleep a lot more than the apnea itself ...

D.H.
Posts: 3535
Joined: Sun Sep 04, 2016 7:07 pm

Re: AHI very high and BiPAP doesnt help

Post by D.H. » Fri Feb 16, 2018 11:34 am

The OP states that his(her) AHI is down from 80 to ten or less. While still too high, it's a big improvement!

It would help if (s)he fills out his(her) profile so we know which machine and mask is being used and also the settings.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
Pugsy
Posts: 65332
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI very high and BiPAP doesnt help

Post by Pugsy » Fri Feb 16, 2018 11:50 am

highapnea wrote:
Fri Feb 16, 2018 11:23 am
I will meet the doctor and request for more pressure on my BiPAP .. At present it is 18/20 if I am correct.
You might want to figure out what kind of apnea events you are having before going down the "let's increase the pressure" road.
More pressure isn't always the solution depending on what that AHI is composed of.
Use available software and see for sure what is happening and when.
https://sleep.tnet.com/equipment

If you are waking often fiddling with the mask you increase the odds of some awake breathing events getting flagged and if you are awake they don't count.
Consider a mask liner to help with the drool issue. It will absorb the slobber.
highapnea wrote:
Fri Feb 16, 2018 11:26 am
Whenever I sleep sideways, I drool and wet my face inside the mask. This wakes me up every time to remove mask and clean face and get back to sleep.. This definitely disturbs the sleep a lot more than the apnea itself ...

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
MrsRinPDX
Posts: 246
Joined: Tue Nov 21, 2017 9:53 am
Location: Portland, OR USA

Re: AHI very high and BiPAP doesnt help

Post by MrsRinPDX » Sat Feb 17, 2018 5:20 pm

Regarding the dry eyes, it may be exhaust air reflected off your blanket if mask leaking isn't the issue. For me it was very subtle air escaping near the bridge of my nose. Either way, the eye drops will help. Give it a try.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Pressures: IPAP 18/ EPAP 14; Secondary mask - Nuance Pro Nasal Pillow

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AHI very high and BiPAP doesnt help

Post by palerider » Sat Feb 17, 2018 5:32 pm

Mikebear wrote:
Fri Feb 02, 2018 8:19 pm
highapnea wrote:Hi

My AHI was 80 while diagnosed for sleep apnea in 2015. Doctor advised BiPAP and brought down the AHI to 7-10 with 7 hrs use .. But I still struggle during the day with watery , burning eyes. Any suggestions ? Does this reduced AHi shown are really real ?
Buy one of these: https://www.walmart.com/ip/Spa-Bella-Sl ... /540361703

Wear it over your eyes when you put on your cpap mask. IF the mask (FFM? likely) is leaking air into your eyeballs due to a bad seal, this will stop it from drying out your eyeballs and you'll know for sure.
It's *BETTER* to use mask liners if there are leaks, instead of leaving the excessive leaks and trying to work around them.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
ShinRyoku
Posts: 59
Joined: Wed Jan 10, 2018 10:05 am

Re: AHI very high and BiPAP doesnt help

Post by ShinRyoku » Sat Feb 17, 2018 5:33 pm

D.H. wrote:
Fri Feb 16, 2018 11:34 am
The OP states that his(her) AHI is down from 80 to ten or less. While still too high, it's a big improvement!
Agreed. Ultimately, goals should be individualized. Even with optimized settings and gear, under 2 is not a realistic goal for all of us.
-Amin
Owner/editor: https://www.sleepapneagroup.com/
Nothing I say on the forum should be taken as medical advice.

ajack
Posts: 977
Joined: Thu Mar 02, 2017 2:54 am
Location: australia

Re: AHI very high and BiPAP doesnt help

Post by ajack » Sat Feb 17, 2018 6:14 pm

highapnea wrote:
Fri Feb 16, 2018 11:26 am
Any suggestions for drooling issue ? Whenever I sleep sideways, I drool and wet my face inside the mask. This wakes me up every time to remove mask and clean face and get back to sleep.. This definitely disturbs the sleep a lot more than the apnea itself ...
Did you drool before? If not, it may just go away as you settle in.

Fixing mask leaks could help the watery eyes, eye guards can help too.

I would get some sleepyhead charts up and see if anyone has some suggestions on your settings.

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 ST-A iVAPS and adapt ASV