Too much humidity causing mask leak

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
rickhowe
Posts: 4
Joined: Sun Jan 14, 2018 1:11 pm

Too much humidity causing mask leak

Post by rickhowe » Sat Jan 27, 2018 7:33 am

During this cold dry weather, I had the humidity set pretty high (6 on a scale of 1-8) for my machine. Then I noticed that I was getting a high-pitched air leak in my full-face mask; it would come and go. I could adjust the mask, and it would stop temporarily. Finally figured out that I was getting water droplets in the mask, that we causing leaks. And if the water moved or evaporated, the leak would go away. Finally I turned down the humidity until it stopped. Now at about 2 (on a scale of 1-8). Mouth a little dryer in the morning (I'm a mouth breather), but no noise.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Quattro™ Air Full Face Mask with Headgear
Additional Comments: Also use a Dreamstation Go for travel with the Quattro Air Full Face Mask with Headgear

User avatar
Sheriff Buford
Posts: 4111
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: Too much humidity causing mask leak

Post by Sheriff Buford » Sat Jan 27, 2018 9:20 am

Please place you equipment in your profile so folks can see what kinda equipment you are using. Do you use a heated air hose?

Sheriff

User avatar
Pugsy
Posts: 65374
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Too much humidity causing mask leak

Post by Pugsy » Sat Jan 27, 2018 9:46 am

You mentioned in another of your posts that you were using the ResMed AirSense 10 machine..
So do you use the heated hose or the non heated hose?
If you are using the heated hose you could probably just increase the hose air temp and avoid the condensation in the mask without reducing the humidity.
If you are using the non heated hose...get a heated hose.
It will allow for condensation prevention without your having to sacrifice moisture in the air.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.