Frustrated - 3 weeks in

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
triumph65
Posts: 5
Joined: Tue Nov 14, 2017 6:37 pm

Frustrated - 3 weeks in

Post by triumph65 » Thu Jan 04, 2018 12:15 pm

I have an Airsense 10 and have been trying the pillows and full mask (swift FX nasal pillows and airfit 10 mask).

Here is the history.

Started with Airfit 10 mask in medium, felt like it rode a little low on nose, causing pinching and made me want to breath out mouth.
Traded for Airfit 10 mask in large and swift FX nasal pillows.

Large mask will blow out above 11 psi, around cheeks. With either mask or nasal pillows I want to open my mouth and the resmed chinstrap does not keep that from happening.

With the pressure blow out on mask I feel the pillows would be my best bet if I can find a way to keep my mouth shut:(

lastlib
Posts: 136
Joined: Wed Aug 14, 2013 12:40 pm
Location: Columbus Ohio
Contact:

Re: Frustrated - 3 weeks in

Post by lastlib » Thu Jan 04, 2018 2:21 pm

If you take it off to make hot monkey love with someone who you think is hot then you will look forward to going to bed with it. You must wear it before and after otherwise the scheme is kaput.

User avatar
Hannibal 2
Posts: 505
Joined: Sun Nov 20, 2016 4:46 am
Location: Derbyshire UK

Re: Frustrated - 3 weeks in

Post by Hannibal 2 » Thu Jan 04, 2018 2:59 pm

triumph65 wrote:I have an Airsense 10 and have been trying the pillows and full mask (swift FX nasal pillows and airfit 10 mask).

Here is the history.

Started with Airfit 10 mask in medium, felt like it rode a little low on nose, causing pinching and made me want to breath out mouth.
Traded for Airfit 10 mask in large and swift FX nasal pillows.

Large mask will blow out above 11 psi, around cheeks. With either mask or nasal pillows I want to open my mouth and the resmed chinstrap does not keep that from happening.

With the pressure blow out on mask I feel the pillows would be my best bet if I can find a way to keep my mouth shut:(
You may want to try mouth taping, there are a few members that do this, but make sure you use the correct tape, micropore or similar medical tape. Search the forum as there are many threads regarding taping.

_________________
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear
Additional Comments: Air 10 ClimateLineAir,
"Welcome my son, welcome to the machine!
Where have you been? It's alright we know where you've been!"
(You've been in the pipeline...)

robz
Posts: 26
Joined: Thu Oct 27, 2016 5:06 am

Re: Frustrated - 3 weeks in

Post by robz » Thu Jan 04, 2018 3:26 pm

It can be challenging with the whole keeping the mouth/lips from burbling and spewing air.......take a look at pap cap, you can find it on amazon....its probably the best system you can find, in my opinion.....I went thru 4 or 5 chin strips that really didn't work until I found this one.....OH, and I have to tape as well......But once you find out the right taping method and the right strap, all will work fairly well.....

_________________
Machine

User avatar
AMK
Posts: 398
Joined: Mon Feb 23, 2015 8:34 am

Re: Frustrated - 3 weeks in

Post by AMK » Thu Jan 04, 2018 4:59 pm

I use the Swift FX and Airsense 10. I have to tape my mouth and use a chin strap. I use Bandaid brand paper tape. It's got just enough hold while not taking off a layer of skin in the morning. Is this an ideal setup? No. But it gets me through the night.

User avatar
zonker
Posts: 11435
Joined: Fri Jun 19, 2015 4:36 pm

Re: Frustrated - 3 weeks in

Post by zonker » Thu Jan 04, 2018 5:50 pm

keep going, you can do this! practice makes perfect and all that.

here's something that works for me in keeping my big mouth shut while i sleep-

viewtopic.php?f=1&t=112758&st=0&sk=t&sd ... i#p1086296

cheap and it works. at least, it works for me.

another tool i use to keep my jaw from opening is a foam cervical collar. that too is relatively cheap!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
jagzoo
Posts: 340
Joined: Tue Jan 15, 2013 7:14 pm
Location: Nh

Re: Frustrated - 3 weeks in

Post by jagzoo » Thu Jan 04, 2018 9:05 pm

Ditto on using a firm cervical collar. Cheap, easy to use and it’s all I needed to stop mouth leaks. And easy to try.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Heated Tubing; mask sometimes is TAP PAP mask; now P10 nasal pillows; use Sleepyhead software
APAP 9-12

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: Frustrated - 3 weeks in

Post by TedVPAP » Thu Jan 04, 2018 10:29 pm

triumph65 wrote:I have an Airsense 10 and have been trying the pillows and full mask (swift FX nasal pillows and airfit 10 mask).

Here is the history.

Started with Airfit 10 mask in medium, felt like it rode a little low on nose, causing pinching and made me want to breath out mouth.
Traded for Airfit 10 mask in large and swift FX nasal pillows.

Large mask will blow out above 11 psi, around cheeks. With either mask or nasal pillows I want to open my mouth and the resmed chinstrap does not keep that from happening.

With the pressure blow out on mask I feel the pillows would be my best bet if I can find a way to keep my mouth shut:(
I never have problems with mask leaking using the P10. I use the Ultimate Chinstrap to control mouth leaking. Posting last night's data which is typical. Note that it is not zero, but it is well controlled. I line it with a folded sock to keep the strap clean.
Image[/quote]

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

User avatar
Hannibal 2
Posts: 505
Joined: Sun Nov 20, 2016 4:46 am
Location: Derbyshire UK

Re: Frustrated - 3 weeks in

Post by Hannibal 2 » Fri Jan 05, 2018 4:22 am

TBH ....I tried mouth taping for a while but it seemed such a wrong thing to do. For me, having to use a mask is enough, I want to keep additional aids to a minimum. I persevered with finding the right FF Mask for me, that turned out to be the Amara View (after much tweaking). Yes I do have a dry mouth in the morning but it's not a big deal because I'm sleeping well, getting low AHI, minimal leaks and feeling great. Nothing is going to be perfect, there will always be compromises, finding the right mask is the hardest part.

_________________
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear
Additional Comments: Air 10 ClimateLineAir,
"Welcome my son, welcome to the machine!
Where have you been? It's alright we know where you've been!"
(You've been in the pipeline...)

Jgirl
Posts: 2
Joined: Fri Jan 05, 2018 12:37 pm

Re: Frustrated - 3 weeks in

Post by Jgirl » Fri Jan 05, 2018 12:55 pm

I’ve been on CPAP therapy for 2 mon5hs. I sympathize! I started with Philips Responics Nuance Pro nasal pillow, had message from Resmed that I was having too many leaks. Went to a Fisher and Paykel Simplus, then got a message from Resmed I was having too many events. Went to doctor and he changed my pressure from 9 to auto, so it varies from 5 to 15, still had too many events. The last two nights I used the full face mask ( Fisher & Paykel) I had 11 ahi and 12ahi. I only had 13 ahi during the sleep study. I went back to the nasal pillow and added a chin strap. Now I have good ahi scores and no leaks .Ahi’s are usually under 1per hour with a couple that were over 2 and over 3, still acceptable. I sleep with CPAP from the time I go to bed until I get up. Saw dr. Yesterday and he said I have perfect compliance. Am feeling better and hopefully over the frustration. I do tend to have a chapped nose from the mask and am work8ng on resolving that. Good luck.

User avatar
Hannibal 2
Posts: 505
Joined: Sun Nov 20, 2016 4:46 am
Location: Derbyshire UK

Re: Frustrated - 3 weeks in

Post by Hannibal 2 » Fri Jan 05, 2018 1:16 pm

Jgirl wrote:I’ve been on CPAP therapy for 2 mon5hs. I sympathize! I started with Philips Responics Nuance Pro nasal pillow, had message from Resmed that I was having too many leaks. Went to a Fisher and Paykel Simplus, then got a message from Resmed I was having too many events. Went to doctor and he changed my pressure from 9 to auto, so it varies from 5 to 15, still had too many events. The last two nights I used the full face mask ( Fisher & Paykel) I had 11 ahi and 12ahi. I only had 13 ahi during the sleep study. I went back to the nasal pillow and added a chin strap. Now I have good ahi scores and no leaks .Ahi’s are usually under 1per hour with a couple that were over 2 and over 3, still acceptable. I sleep with CPAP from the time I go to bed until I get up. Saw dr. Yesterday and he said I have perfect compliance. Am feeling better and hopefully over the frustration. I do tend to have a chapped nose from the mask and am work8ng on resolving that. Good luck.
Well done! You're very fortunate to achieve 'perfect compliance' in such a short space of time, some do but for most it can take a lot longer. Hope it continues to work well for you.

_________________
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear
Additional Comments: Air 10 ClimateLineAir,
"Welcome my son, welcome to the machine!
Where have you been? It's alright we know where you've been!"
(You've been in the pipeline...)