AHI still not on track

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Danlo67
Posts: 53
Joined: Tue Dec 19, 2017 11:38 pm

AHI still not on track

Post by Danlo67 » Mon Jan 01, 2018 6:29 pm

Hi all. I'm posting my dreammapper results for the last 2 nights. I don't have SleepyHead data yet.
I'm set on minimum 11.5 cm H20 and still have 5 and 6 AHI. Do I need to minimum up again? ImageImage

Sent from my Kogan Agora 8 using Tapatalk
Dreamstation Auto Nasal Mask
Pressure Setting at 9-20 cm H20. currently. Still working on getting AHI below 5

_________________
MaskHumidifier
Using Dreamstation AutoCPAP
Setting at 13cm Fixed
Dreamweaver gel pillow mask

User avatar
Bill44133
Posts: 1087
Joined: Tue Jul 09, 2013 8:34 pm
Location: North Royalton, OH

Re: AHI still not on track

Post by Bill44133 » Mon Jan 01, 2018 6:42 pm

Looks like you need more pressure to me.

_________________
Machine: DreamStation BiPAP® Auto Machine
Humidifier: DreamStation Heated Humidifier
Additional Comments: Settings are IPap 23 EPap 19

User avatar
Danlo67
Posts: 53
Joined: Tue Dec 19, 2017 11:38 pm

Re: AHI still not on track

Post by Danlo67 » Mon Jan 01, 2018 7:14 pm

My 90% pressure is showing a range from 13.2 - 13.9 over the 6 nights. Should I try 13 minimum?

Sent from my Kogan Agora 8 using Tapatalk
Dreamstation Auto Nasal Mask
Pressure Setting at 9-20 cm H20. currently. Still working on getting AHI below 5

_________________
MaskHumidifier
Using Dreamstation AutoCPAP
Setting at 13cm Fixed
Dreamweaver gel pillow mask

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI still not on track

Post by Pugsy » Mon Jan 01, 2018 7:22 pm

Why do you think you need to try 13minimum?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: AHI still not on track

Post by TedVPAP » Mon Jan 01, 2018 7:39 pm

Danlo67 wrote:Hi all. I'm posting my dreammapper results for the last 2 nights. I don't have SleepyHead data yet.
I'm set on minimum 11.5 cm H20 and still have 5 and 6 AHI. Do I need to minimum up again? ImageImage

Sent from my Kogan Agora 8 using Tapatalk
Dreamstation Auto Nasal Mask
Pressure Setting at 9-20 cm H20. currently. Still working on getting AHI below 5
Without looking at sleepyhead data, guessing that more pressure is needed is a reasonable guess. But it is just a guess.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI still not on track

Post by Pugsy » Mon Jan 01, 2018 7:46 pm

Prior thread with one SleepyHead report image.

viewtopic.php?f=1&t=157946&p=1214220#p1214220

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Danlo67
Posts: 53
Joined: Tue Dec 19, 2017 11:38 pm

Re: AHI still not on track

Post by Danlo67 » Mon Jan 01, 2018 7:51 pm

Only asking because between 13 - 14 is the 90⅝ level. I figure that's about where there pressure has been sitting for last 6 nights?

Sent from my Kogan Agora 8 using Tapatalk
Dreamstation Auto Nasal Mask
Pressure Setting at 9-20 cm H20. currently. Still working on getting AHI below 5

_________________
MaskHumidifier
Using Dreamstation AutoCPAP
Setting at 13cm Fixed
Dreamweaver gel pillow mask

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI still not on track

Post by Pugsy » Mon Jan 01, 2018 8:17 pm

The definition of 90% pressures is NOT where the machine was at for 90% of the night.

The definition of 90% numbers is....at OR BELOW for 90% of the night. That "or below" part of the definition is kinda important.

What is the overall average or median average pressure?....you would need to open up SleepyHead to see that number.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Danlo67
Posts: 53
Joined: Tue Dec 19, 2017 11:38 pm

Re: AHI still not on track

Post by Danlo67 » Mon Jan 01, 2018 8:26 pm

Thanks. I got it wrong then
Will have to check SleepyHead when I get home

Sent from my Kogan Agora 8 using Tapatalk
Dreamstation Auto Nasal Mask
Pressure Setting at 9-20 cm H20. currently. Still working on getting AHI below 5

_________________
MaskHumidifier
Using Dreamstation AutoCPAP
Setting at 13cm Fixed
Dreamweaver gel pillow mask

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: AHI still not on track

Post by TedVPAP » Mon Jan 01, 2018 8:35 pm

Pugsy wrote:Prior thread with one SleepyHead report image.

viewtopic.php?f=1&t=157946&p=1214220#p1214220
Thanks.
I posted in the linked thread.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

User avatar
Danlo67
Posts: 53
Joined: Tue Dec 19, 2017 11:38 pm

Re: AHI still not on track

Post by Danlo67 » Tue Jan 02, 2018 12:41 am

I haven't checked SleepyHead yet but I've got a sneaking suspicion the median it will give me will be the same as my minimum is now at (11.5 cm) . What do I do then? Since AHI is still not on track and I'm still not feeling as I should?

Sent from my Kogan Agora 8 using Tapatalk
Dreamstation Auto Nasal Mask
Pressure Setting at 9-20 cm H20. currently. Still working on getting AHI below 5

_________________
MaskHumidifier
Using Dreamstation AutoCPAP
Setting at 13cm Fixed
Dreamweaver gel pillow mask

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI still not on track

Post by Pugsy » Tue Jan 02, 2018 7:45 am

Increase the minimum again.

It's a process and takes some time because we don't sleep the same each night.
You keep increasing until you get the AHI where you are want it or close to it.

If you want to make a bigger jump in minimum that is fine...just do it for the right reason and not the wrong reason.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.