Resperonics Dreamstation battery backup project
-
- Posts: 1795
- Joined: Mon Mar 27, 2017 5:45 pm
- Location: Central Virginia
Re: Resperonics Dreamstation battery backup project
Before you can discuss "battery backup" you end to define your requirements & expectations.
For my Resmed system, I am assembling, but have not tested, a system to be used for extended power outage situations in the country.
I bought the 12V cord from our sponsor I purchased a 100W solar panel set & 45AH AGM battery from Harbor Freight. I have not yet assembled & tested.
My calculations estimate about 4 hours of good sun to charge and 5 days capacity if I do not use the humidifier or heated hose.
For my Resmed system, I am assembling, but have not tested, a system to be used for extended power outage situations in the country.
I bought the 12V cord from our sponsor I purchased a 100W solar panel set & 45AH AGM battery from Harbor Freight. I have not yet assembled & tested.
My calculations estimate about 4 hours of good sun to charge and 5 days capacity if I do not use the humidifier or heated hose.
_________________
Machine: AirSense 11 Autoset |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software |
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.
Re: Resperonics Dreamstation battery backup project
A rechargeable that could hold enough juice for 24 hours of operation and be able to re-up using variable voltages.
mc
mc
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: Simplus Full Face CPAP Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Just an average joe on his way from here to there...
Re: Resperonics Dreamstation battery backup project
This is exactly right. Too often I've struggled with design parameters of a project only to realize that I've not properly defined my objectives in the first place.prodigyplace wrote:Before you can discuss "battery backup" you end to define your requirements & expectations.
You need to decide:
- DC or AC, or both?
- Portable, seldom moved, or will it rarely be moved?
- If portable, then by what means and what is your design weight limit?
- How many nights does it need to last?
- Does it also need to run the humidifier, heated hose, etc?
- what pressure do you use?
- How much do you want to spend?
- Is there anything else you need to power off this battery system (e.g. cell phone charger, etc)?
I'm workin' on it.
Re: Resperonics Dreamstation battery backup project
Not even doing the math I know there is no way 4 hours of sun is charging a 45 AH battery. The down side to traditional Lead acid batteries besides weight is most you kill the life of battery by drawing down in power to much.prodigyplace wrote:Before you can discuss "battery backup" you end to define your requirements & expectations.
For my Resmed system, I am assembling, but have not tested, a system to be used for extended power outage situations in the country.
I bought the 12V cord from our sponsor I purchased a 100W solar panel set & 45AH AGM battery from Harbor Freight. I have not yet assembled & tested.
My calculations estimate about 4 hours of good sun to charge and 5 days capacity if I do not use the humidifier or heated hose.
10 to 15 cm h20
Re: Resperonics Dreamstation battery backup project
Thank you! This is the type of info I was asking for.mrcrayola wrote:A rechargeable that could hold enough juice for 24 hours of operation and be able to re-up using variable voltages.
mc
10 to 15 cm h20
Re: Resperonics Dreamstation battery backup project
Not for a 45 AH battery at 0% or 20% charged state it would not, but to replenish one night's CPAP usage (no heat or humidifier) it should be more that adequate. I've done it with a 27 Watt panel no problem.USMCVet wrote: Not even doing the math I know there is no way 4 hours of sun is charging a 45 AH battery. The down side to traditional Lead acid batteries besides weight is most you kill the life of battery by drawing down in power to much.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Oracle HC452 Oral CPAP Mask |
Re: Resperonics Dreamstation battery backup project
The resperonics website says that without heated tube or humidification you can get 14 hours of run time at 10cm on 90 wh. So even only getting 50 watts from a 100 watt panel one night of use should only take an hour. I know that is over simplification but shows that poster is not talking about charging after only one night.amenite wrote:Not for a 45 AH battery at 0% or 20% charged state it would not, but to replenish one night's CPAP usage (no heat or humidifier) it should be more that adequate. I've done it with a 27 Watt panel no problem.USMCVet wrote: Not even doing the math I know there is no way 4 hours of sun is charging a 45 AH battery. The down side to traditional Lead acid batteries besides weight is most you kill the life of battery by drawing down in power to much.
10 to 15 cm h20
Re: Resperonics Dreamstation battery backup project
The Dreamstation runs on 12 Volts, as does all modern Respironics units. You're thinking of the ResMed units that run on 24 volts and require a special signal in the plug. Although its possible to hack a plug that works, most people are content to use the DC/DC converter from ResMed. There are a few battery packs that come with the special plug and circuitry, but most ResMed users should use the converter if they desire battery power.Woody wrote:You need to consider that the Dreamstation runs on a 24v dc supply not 12v dc. So just using
a 12v battery isn't going to work. Resperonics does make a small adapter to let you use a 12v
battery for about $ 90.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: Quattro™ Air Full Face Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: Pressure 9-20, average ~9.5; often use battery power while off-grid |
Hark, how hard he fetches breath . . . Act II, Scene IV, King Henry IV Part I, William Shakespeare
Choosing a Battery thread: http://www.cpaptalk.com/viewtopic/t1140 ... ttery.html
Choosing a Battery thread: http://www.cpaptalk.com/viewtopic/t1140 ... ttery.html
Re: Resperonics Dreamstation battery backup project
A 100 watt panel can deliver about 8 amps in good sun. The "rule of thumb" is that you can expect on average 3 to 4 hours a day, depending on location and season. (Google "average sun hours" to find charts and tables.) This assumes you have a good view - no trees, etc - and reasonable luck with the weather. Since even a mild overcast can cut the power in half, and clouds can cut it 90%, you need a backup plan or large batteries. Also, the battery has issues that must be considered, such as a low "acceptance rate" when it get more than 80% charged, and you don't want want to drain too low. I try to keep my batteries healthy through several hundred cycles over 6 years, so I tend to run between 50% and 85% daily.USMCVet wrote:Not even doing the math I know there is no way 4 hours of sun is charging a 45 AH battery. The down side to traditional Lead acid batteries besides weight is most you kill the life of battery by drawing down in power to much.prodigyplace wrote:Before you can discuss "battery backup" you end to define your requirements & expectations.
For my Resmed system, I am assembling, but have not tested, a system to be used for extended power outage situations in the country.
I bought the 12V cord from our sponsor I purchased a 100W solar panel set & 45AH AGM battery from Harbor Freight. I have not yet assembled & tested.
My calculations estimate about 4 hours of good sun to charge and 5 days capacity if I do not use the humidifier or heated hose.
A 100 watt panel, well positioned (facing south with no trees) could generate 25 to 30 amp-hours a day on average. In practice I would expect less, maybe a lot less in a country area with limited horizons. On my boat I have 350 Watts of panels. There are no trees, but as the boat is turning on a mooring, only some of the panels are powered at a time. I average around 60 Amp-hours a day, enough to power the fridge and pumps, etc while I'm away, but If I'm on board (using more power) during a cloudy period I have to run an engine to keep up.
My hunch is that the 100 watt panel with a 45 AH battery would work fairly well to keep a cpap running, if that was the only load. If you were trying to run many devices I don't think you could count on fully charging the battery every day.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: Quattro™ Air Full Face Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: Pressure 9-20, average ~9.5; often use battery power while off-grid |
Hark, how hard he fetches breath . . . Act II, Scene IV, King Henry IV Part I, William Shakespeare
Choosing a Battery thread: http://www.cpaptalk.com/viewtopic/t1140 ... ttery.html
Choosing a Battery thread: http://www.cpaptalk.com/viewtopic/t1140 ... ttery.html
Re: Resperonics Dreamstation battery backup project
Exactly and the cord isn't $90 either https://www.amazon.com/gp/aw/d/B01FT8QU ... MK7J&psc=1 .CapnLoki wrote:The Dreamstation runs on 12 Volts, as does all modern Respironics units. You're thinking of the ResMed units that run on 24 volts and require a special signal in the plug. Although its possible to hack a plug that works, most people are content to use the DC/DC converter from ResMed. There are a few battery packs that come with the special plug and circuitry, but most ResMed users should use the converter if they desire battery power.Woody wrote:You need to consider that the Dreamstation runs on a 24v dc supply not 12v dc. So just using
a 12v battery isn't going to work. Resperonics does make a small adapter to let you use a 12v
battery for about $ 90.
10 to 15 cm h20
Re: Resperonics Dreamstation battery backup project
What do you mean by "variable voltages"? Do you mean international voltages, or do you have some other special need?mrcrayola wrote:A rechargeable that could hold enough juice for 24 hours of operation and be able to re-up using variable voltages.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: Quattro™ Air Full Face Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: Pressure 9-20, average ~9.5; often use battery power while off-grid |
Hark, how hard he fetches breath . . . Act II, Scene IV, King Henry IV Part I, William Shakespeare
Choosing a Battery thread: http://www.cpaptalk.com/viewtopic/t1140 ... ttery.html
Choosing a Battery thread: http://www.cpaptalk.com/viewtopic/t1140 ... ttery.html
Re: Resperonics Dreamstation battery backup project
Here is a what I'm thinking for components so for.
Battery:
http://www.batteryspace.com/lifepo4-pri ... d-dgr.aspx
A/C charger:
http://www.batteryspace.com/smartcharge ... 40vac.aspx
Solar Charger:
http://www.batteryspace.com/Charge-Cont ... r-LFP.aspx
Cell balancing:
http://www.batteryspace.com/Smart-LED-B ... rrent.aspx
Still plenty to figure out like how to package it. Maybe a a hard plastic case so CPAP itself can be inside for protection when being transported?
Battery:
http://www.batteryspace.com/lifepo4-pri ... d-dgr.aspx
A/C charger:
http://www.batteryspace.com/smartcharge ... 40vac.aspx
Solar Charger:
http://www.batteryspace.com/Charge-Cont ... r-LFP.aspx
Cell balancing:
http://www.batteryspace.com/Smart-LED-B ... rrent.aspx
Still plenty to figure out like how to package it. Maybe a a hard plastic case so CPAP itself can be inside for protection when being transported?
10 to 15 cm h20
Re: Resperonics Dreamstation battery backup project
If you are carrying your XPAP in a case with batteries, feel free to buy a extra cpap or to, very poor idea.. JimUSMCVet wrote:Here is a what I'm thinking for components so for.
Battery:
http://www.batteryspace.com/lifepo4-pri ... d-dgr.aspx
A/C charger:
http://www.batteryspace.com/smartcharge ... 40vac.aspx
Solar Charger:
http://www.batteryspace.com/Charge-Cont ... r-LFP.aspx
Cell balancing:
http://www.batteryspace.com/Smart-LED-B ... rrent.aspx
Still plenty to figure out like how to package it. Maybe a a hard plastic case so CPAP itself can be inside for protection when being transported?
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
Re: Resperonics Dreamstation battery backup project
Care to elaborate? Battery would be secured to case with separate section for CPAP with foam for it to rest in.Goofproof wrote:If you are carrying your XPAP in a case with batteries, feel free to buy a extra cpap or to, very poor idea.. JimUSMCVet wrote:Here is a what I'm thinking for components so for.
Battery:
http://www.batteryspace.com/lifepo4-pri ... d-dgr.aspx
A/C charger:
http://www.batteryspace.com/smartcharge ... 40vac.aspx
Solar Charger:
http://www.batteryspace.com/Charge-Cont ... r-LFP.aspx
Cell balancing:
http://www.batteryspace.com/Smart-LED-B ... rrent.aspx
Still plenty to figure out like how to package it. Maybe a a hard plastic case so CPAP itself can be inside for protection when being transported?
10 to 15 cm h20
Re: Resperonics Dreamstation battery backup project
Your back will love the strain, the machine will love the vibration and the fumes, just a poor idea in general. Jim
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire