still getting high leak readings

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
debdix
Posts: 12
Joined: Fri Dec 08, 2017 10:06 pm

still getting high leak readings

Post by debdix » Thu Dec 14, 2017 6:27 pm

Using resmed 10 cpap. Things going ok. Big improvements to ahi by using a chin strap. However, on MyAirWeb I am still getting really high mask leakage readings. I have tried to make sure it is fitting well when I am using it and don't know what else to do. Any suggestions?

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: still getting high leak readings

Post by Wulfman... » Thu Dec 14, 2017 6:35 pm

debdix wrote:Using resmed 10 cpap. Things going ok. Big improvements to ahi by using a chin strap. However, on MyAirWeb I am still getting really high mask leakage readings. I have tried to make sure it is fitting well when I am using it and don't know what else to do. Any suggestions?
Use Sleepyhead software and see if you can see what the problem is.
From the lack of information in your post, we have nothing to go on.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

debdix
Posts: 12
Joined: Fri Dec 08, 2017 10:06 pm

Re: still getting high leak readings

Post by debdix » Thu Dec 14, 2017 7:07 pm

I do use sleepyhead but I don't know how to transfer the results into this post so you can see them

User avatar
Pugsy
Posts: 65284
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: still getting high leak readings

Post by Pugsy » Thu Dec 14, 2017 7:19 pm

How to post an image of your report and what we like to see (don't want all those graphs). Examples in this thread
viewtopic/t103468/Need-help-with-screen-shots.html

How to host and image using imgur
https://sleep.tnet.com/reference/tips/imgur

How to organize the graphs..3 pages..read all 3
https://sleep.tnet.com/resources/sleepyhead/shorganize

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

debdix
Posts: 12
Joined: Fri Dec 08, 2017 10:06 pm

Re: still getting high leak readings

Post by debdix » Thu Dec 14, 2017 9:30 pm

Image

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: still getting high leak readings

Post by Wulfman... » Thu Dec 14, 2017 9:44 pm

Since you mentioned using a chin strap, I'm guessing you're using a nasal mask.

Possibilities:
Mouth opening.
Bed pillow pushing your mask out of alignment. (less likely)


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
greatunclebill
Posts: 1503
Joined: Mon Feb 20, 2012 7:48 pm
Location: L.A. (lower alabama)

Re: still getting high leak readings

Post by greatunclebill » Thu Dec 14, 2017 10:03 pm

I'll second the mouth opening. the chin strap keeps the jaw tight, but does nothing to keep the mouth (lips) closed. some people tape the lips closed.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: myAir, OSCAR. cms-50D+. airsense 10 auto & (2009) remstar plus m series backups
First diagnosed 1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990

debdix
Posts: 12
Joined: Fri Dec 08, 2017 10:06 pm

Re: still getting high leak readings

Post by debdix » Thu Dec 14, 2017 11:50 pm

thanks for that. Dont know about taping my mouth shut (although I know a few people who would agree with you! LOL
Anyway, might see how it goes until after xmas, and make an appt with my sleep tech to maybe try full face mask again..although I didnt like it the first time

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: still getting high leak readings

Post by palerider » Fri Dec 15, 2017 1:39 am

debdix wrote:Dont know about taping my mouth shut (although I know a few people who would agree with you! LOL
Yes, they would, because it's not a joke.

Google it.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Muse-Inc
Posts: 4382
Joined: Fri Jun 12, 2009 8:44 pm
Location: Atlanta, GA

Re: still getting high leak readings

Post by Muse-Inc » Sat Dec 16, 2017 3:56 am

debdix wrote:make an appt with my sleep tech to maybe try full face mask again..although I didnt like it the first time
Try my mask (click on it in my signature). It's usually priced about $119, sometimes less or on sale. Has pillows and covers mouth so can breathe through nose of mouth and not lose therapy air.
ResMed S9 range 9.8-17, RespCare Hybrid FFM
Never, never, never, never say never.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: still getting high leak readings

Post by Goofproof » Sat Dec 16, 2017 9:07 am

debdix wrote:
make an appt with my sleep tech to maybe try full face mask again..although I didnt like it the first time



You may find in life there are some things we don't like, but need to do to make our lives better. Sometimes we need to adapt, for our own good. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: still getting high leak readings

Post by Okie bipap » Sat Dec 16, 2017 10:16 am

Goofproof wrote:You may find in life there are some things we don't like, but need to do to make our lives better. Sometimes we need to adapt, for our own good. Jim
I agree. I would love to be able to use nasal pillows. I have tried several times, but I cannot keep my lips sealed well enough to make it work. I am not a mouth breather when I am awake, but I am when sleeping. I had UPPP surgery several years ago which alleviated many of my apnea symptoms for several years, but I am now using a bi-level machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: still getting high leak readings

Post by TedVPAP » Sat Dec 16, 2017 10:46 am

Okie bipap wrote:
Goofproof wrote:You may find in life there are some things we don't like, but need to do to make our lives better. Sometimes we need to adapt, for our own good. Jim
I agree. I would love to be able to use nasal pillows. I have tried several times, but I cannot keep my lips sealed well enough to make it work. I am not a mouth breather when I am awake, but I am when sleeping. I had UPPP surgery several years ago which alleviated many of my apnea symptoms for several years, but I am now using a bi-level machine.
I use the Ultimate Chipstrap in order to control mouth leaking. Works great.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

BarclayVision
Posts: 1
Joined: Tue Sep 04, 2018 3:53 pm

Re: still getting high leak readings

Post by BarclayVision » Thu Sep 06, 2018 3:40 am

I've been fighting high mask leak since i switched to nasel pillows, have added chin strap and that has improved readings but still seem extremely high??? My sleep seems fine fr the most part, and I acceptable fit reading in the morning on Resmed Air 10 (green smiley face) - but Sleepy Head reading say otherwise???
Image
Image

User avatar
Pugsy
Posts: 65284
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: still getting high leak readings

Post by Pugsy » Thu Sep 06, 2018 7:00 am

BarclayVision wrote:
Thu Sep 06, 2018 3:40 am
I've been fighting high mask leak since i switched to nasel pillows, have added chin strap and that has improved readings but still seem extremely high??? My sleep seems fine fr the most part, and I acceptable fit reading in the morning on Resmed Air 10 (green smiley face) - but Sleepy Head reading say otherwise???
Welcome to the forum.

Please read the Sticky thread at the top of the main forum topic page.
Please start your own thread for help with your problem. It's going to get lost here in this old thread.
Mr Smiley for leak management doesn't go away until you spend 30 % or more of the night over 24 L/min. SleepyHead will tell you that your leak management sucks long before Mr Frowny shows up.

Please review this thread for how to post and in what format to show the graphs. You have omitted some useful stuff and included stuff that isn't particularly useful.
viewtopic/t158560/How-to-post-images-for-review.html

Have you tried taping your mouth one night just to rule in or out that the big leak you are seeing is from mouth breathing or mask movement?
See the large leaks here...people might think mouth breathing but they are all mask movement issues because I taped my mouth this night.
Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.