How many thousand hours is considered an old machine?
-
Suscamoer
How many thousand hours is considered an old machine?
Looking at used machines for back up. Wouldn't get a car with over 100,000 miles on it. What is the equivalent in a cpap?
Re: How many thousand hours is considered an old machine?
For a back up machine...depends on the price of said used machine and model age as much as usage hours age.
There are some really low hour old machine but they are REALLY old in terms of model age....dinosaurs that I wouldn't want if they were free.
Now a newer model with maybe 2 to 4 K hours on it wouldn't scare me off depending on the price and what I was going to use it for.
If the price was right and I was only going to use it for travel (once or twice a year) or as a back up...I wouldn't be afraid to get a machine with 5 K hours on it. I once gave away a machine that had 8500 hours on it ....it was only for emergency trips to the hospital.
There are some really low hour old machine but they are REALLY old in terms of model age....dinosaurs that I wouldn't want if they were free.
Now a newer model with maybe 2 to 4 K hours on it wouldn't scare me off depending on the price and what I was going to use it for.
If the price was right and I was only going to use it for travel (once or twice a year) or as a back up...I wouldn't be afraid to get a machine with 5 K hours on it. I once gave away a machine that had 8500 hours on it ....it was only for emergency trips to the hospital.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: How many thousand hours is considered an old machine?
The average user will put 2500-2600 hours a year on a machine. Average life of a machine is 13,000-15,000 hours. If you are looking for a used machine, you should be able to find really good machines with less than 5000 hours for under $250 on Craigslist.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
- DreamStalker
- Posts: 7509
- Joined: Mon Aug 07, 2006 9:58 am
- Location: Nowhere & Everywhere At Once
Re: How many thousand hours is considered an old machine?
I think I have over 23,000 hrs on mine ... but it's a "dinosaur" legacy model. I do have a couple of back ups 1) another "dinosaur" legacy model, and 2) a newer M-Series legacy model (not sure the hrs on that one cuz I only use it for travel and I don't travel much) -- would that be an "Ice Age" model?Suscamoer wrote:Looking at used machines for back up. Wouldn't get a car with over 100,000 miles on it. What is the equivalent in a cpap?
But any working machine is better than no backup machine ... IMO.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.
Re: How many thousand hours is considered an old machine?
+1 rather have one that'll last a couple nights till I can get a replacement than go one night without xPAP.DreamStalker wrote:I think I have over 23,000 hrs on mine ... but it's a "dinosaur" legacy model. I do have a couple of back ups 1) another "dinosaur" legacy model, and 2) a newer M-Series legacy model (not sure the hrs on that one cuz I only use it for travel and I don't travel much) -- would that be an "Ice Age" model?Suscamoer wrote:Looking at used machines for back up. Wouldn't get a car with over 100,000 miles on it. What is the equivalent in a cpap?
But any working machine is better than no backup machine ... IMO.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Backup/travel unit is an identical S9 AutoSet for Her w/Eson nasal mask |
- greatunclebill
- Posts: 1503
- Joined: Mon Feb 20, 2012 7:48 pm
- Location: L.A. (lower alabama)
Re: How many thousand hours is considered an old machine?
In about 6 years I have put over 15,000 hours on my PR550auto
. That's with 98.6% usage rate and average over 7 hrs usage per night. I have seen no signs of anything wrong.
From Second Wind that buy and sell used machines:
"The average life expectancy of a CPAP or BPAP machine is approximately 20,000 hours, or about seven to eight years full time use. That said, many times, if properly cared for, these devices will last much longer. Many properly maintained machines will last upwards of 50,000 hours."
. That's with 98.6% usage rate and average over 7 hrs usage per night. I have seen no signs of anything wrong.
From Second Wind that buy and sell used machines:
"The average life expectancy of a CPAP or BPAP machine is approximately 20,000 hours, or about seven to eight years full time use. That said, many times, if properly cared for, these devices will last much longer. Many properly maintained machines will last upwards of 50,000 hours."
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Additional Comments: myAir, OSCAR. cms-50D+. airsense 10 auto & (2009) remstar plus m series backups |
Last edited by greatunclebill on Wed Oct 25, 2017 11:23 am, edited 1 time in total.
First diagnosed 1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: How many thousand hours is considered an old machine?
" . . . Properly maintained" would in my opinion, mean keeping fresh filters in the intake.
---to avoid over-working the motor by continually pulling air through a clogged filter.
Twice a month is the Medicare guideline. Once a year is OUTRAGEOUS.
Manufacturers who even mention this destructive timeline need to be held responsible.
Determining if the machine has not been ABUSED by this practice is next to impossible,
but the life expectancy could be severely reduced.
---to avoid over-working the motor by continually pulling air through a clogged filter.
Twice a month is the Medicare guideline. Once a year is OUTRAGEOUS.
Manufacturers who even mention this destructive timeline need to be held responsible.
Determining if the machine has not been ABUSED by this practice is next to impossible,
but the life expectancy could be severely reduced.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: How many thousand hours is considered an old machine?
It all depends on the filter, basically, the bigger the filter, the less often it probably needs to be changed, which is one of the reasons that resmed specs a longer change interval between recommended filter changes that respironics.chunkyfrog wrote:" . . . Properly maintained" would in my opinion, mean keeping fresh filters in the intake.
---to avoid over-working the motor by continually pulling air through a clogged filter.
Twice a month is the Medicare guideline. Once a year is OUTRAGEOUS.
I've been working on my roommate's oxygen concentrator, and it has a true HEPA filter on the compressor intake. It doesn't even specify a change time on that. Just "check during PM/between patients". (PMs are every three years/26,280 hours.)
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: How many thousand hours is considered an old machine?
I know he's never wrong and he won't admit it, but 5000 hours is less than 2 years use.xxyzx wrote:=========Suscamoer wrote:Looking at used machines for back up. Wouldn't get a car with over 100,000 miles on it. What is the equivalent in a cpap?
5000 is the design life roughly 100k miles on a car
they can legally sell cars as new (at least in this state) with up to 2999 miles on it
anything over 150 hours on it would be a used xpap if they had a similar law
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
Re: How many thousand hours is considered an old machine?
Lincare come out and checks mine monthly, test the O2 % and flow. Jimpalerider wrote:It all depends on the filter, basically, the bigger the filter, the less often it probably needs to be changed, which is one of the reasons that resmed specs a longer change interval between recommended filter changes that respironics.chunkyfrog wrote:" . . . Properly maintained" would in my opinion, mean keeping fresh filters in the intake.
---to avoid over-working the motor by continually pulling air through a clogged filter.
Twice a month is the Medicare guideline. Once a year is OUTRAGEOUS.
I've been working on my roommate's oxygen concentrator, and it has a true HEPA filter on the compressor intake. It doesn't even specify a change time on that. Just "check during PM/between patients". (PMs are every three years/26,280 hours.)
It Is a old model (Like my "TANK") They had brought our 3 newer units, all 3 failed. right out of his truck...
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
Re: How many thousand hours is considered an old machine?
https://www.resmed.com/us/en/consumer/p ... plies.htmlxxyzx wrote:==========LSAT wrote:I know he's never wrong and he won't admit it, but 5000 hours is less than 2 years use.xxyzx wrote:=========Suscamoer wrote:Looking at used machines for back up. Wouldn't get a car with over 100,000 miles on it. What is the equivalent in a cpap?
5000 is the design life roughly 100k miles on a car
they can legally sell cars as new (at least in this state) with up to 2999 miles on it
anything over 150 hours on it would be a used xpap if they had a similar law
i only admit to being wrong when i am wrong
i read their documents whre resmed says the design life is 5000 hours
most will run longer
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
Re: How many thousand hours is considered an old machine?
*Citation neededxxyzx wrote:==========LSAT wrote:I know he's never wrong and he won't admit it, but 5000 hours is less than 2 years use.xxyzx wrote:=========Suscamoer wrote:Looking at used machines for back up. Wouldn't get a car with over 100,000 miles on it. What is the equivalent in a cpap?
5000 is the design life roughly 100k miles on a car
they can legally sell cars as new (at least in this state) with up to 2999 miles on it
anything over 150 hours on it would be a used xpap if they had a similar law
i only admit to being wrong when i am wrong
i read their documents whre resmed says the design life is 5000 hours
most will run longer
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: How many thousand hours is considered an old machine?
I have reason to take Palerider at his word.
Resmed publications are usually intended only for DME's.;
--NOT for viewing/reprinting by inconsequential little patients--
He does not make things up.
Resmed publications are usually intended only for DME's.;
--NOT for viewing/reprinting by inconsequential little patients--
He does not make things up.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: How many thousand hours is considered an old machine?
So here you say that Resmed claims the design life is 5000 hours, but on another thread (viewtopic/t157071/Should-I-get-a-new-ma ... l#p1202825) you say "design life is 5 years". So which is it? 5 years or 5000 hours?xxyzx wrote:==========LSAT wrote:I know he's never wrong and he won't admit it, but 5000 hours is less than 2 years use.xxyzx wrote:=========Suscamoer wrote:Looking at used machines for back up. Wouldn't get a car with over 100,000 miles on it. What is the equivalent in a cpap?
5000 is the design life roughly 100k miles on a car
they can legally sell cars as new (at least in this state) with up to 2999 miles on it
anything over 150 hours on it would be a used xpap if they had a similar law
i only admit to being wrong when i am wrong
i read their documents whre resmed says the design life is 5000 hours
most will run longer
- ChicagoGranny
- Posts: 15419
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: How many thousand hours is considered an old machine?
My recommendation is to purchase a low hours used machine for backup. Get a machine that is the equivalent of your current machine. Rotate the two machines every three months so you are sure both work.Suscamoer wrote:Looking at used machines for back up. Wouldn't get a car with over 100,000 miles on it. What is the equivalent in a cpap?
You may pay more than for a used machine with low hours, but over the life of your two machines you should get your money's worth.
I did something like this buying a low-hours, gently used machine with one-year seller's warranty from http://secondwindcpap.com/.
Last edited by ChicagoGranny on Wed Oct 25, 2017 1:39 pm, edited 1 time in total.