Remstar Classic E-25 Error

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Snoredog
Posts: 6399
Joined: Sun Mar 19, 2006 5:09 pm

Remstar Classic E-25 Error

Post by Snoredog » Thu Dec 07, 2006 11:26 pm

Well after another 3-day trip to UCLA, I took along my old reliable RemstarPro Classic, it was one of the first C-Flex version machines to come out (reason I bought it).

Machine has 6500+ blower hours on it and never a single problem ...that is until last night. This is my travel machine I mainly use it when I camp or travel.

So on the 2nd night of my trip in LA I got the ole Pro out of its bag, connect the hose & mask to the machine and lay down to sleep. Hit the On/Off button, the machine revs up to full tilt blowing like 20cm!! and is really noisy, man must have a hole in the hose or something, just then the machine starts beeping shuts off with E-25 flashing on the LCD.

ah oh sounds fatal:
Image

Check the hose, mask no leaks man it is it noisy. Now this sucks, I have to drive like 5hrs the next day and without cpap that is going to be really difficult. In any case I made it home with lots of Starbucks.

So I got out a phillips screwdriver and removed the 3 screws holding the lid on the machine. Lifted up the main PCB, disconnected the motor wires, the round disc. Start looking for anything unusual, mmm.... this hose appears to be disconnected. Interesting the hose barb where the hose connects is inside the hose going to the pressure transducer. But where did it come from?

Cause of E-25 error:
The hose I used has very stiff cups on the end, the hose is difficult to put on the machine, so I give it a twist to get the hose on, but sometimes removal is also difficult. It appears that when connecting a hose directly to the machine you can actually "twist" the outlet port in the machine housing (this end of the machine only has 1 screw in the center, it can allow the top lid of the machine to "seperate" which can allow the port to twist, when it twists it can break off the Pressure transducer barb on the port

Outlet port where hose connects:
Image
Here is the part removed from the machine:
Image

Broken plastic barb was found in the hose going to the Pressure transducer on the main PCB. I used a brass brad nail to glue it back on, the nail made sure the hole aligned. I used some SuperGlue and some hot-melt glue to add extra strength.
Image

Image
Image
Image

I never bothered before but be careful connecting or disconnecting your cpap hose on your Remstar Classic (the Autopaps should be the same unless they changed the design).

Installing a hose with stiff ends can twist the port (top lid can seperate at the seam allowing the port to twist breaking off the internal hose barb.

I could have thrown this machine in the trash or sent it in for repairs, it is a old machine with 6500+ hours on it, been out of warranty for years and I could have sent it in if I wanted to pay for the repair, but simple repair job should hold up for what I use it for.


User avatar
rested gal
Posts: 12880
Joined: Thu Sep 09, 2004 10:14 pm
Location: Tennessee

Post by rested gal » Thu Dec 07, 2006 11:48 pm

Wow, Snoredog. Great detective work, great job, great pics!

Cool.
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435

-SWS
Posts: 5301
Joined: Tue Jan 11, 2005 7:06 pm

Post by -SWS » Thu Dec 07, 2006 11:55 pm

Great post and very nice repair work, Snoredog!!!

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Post by Goofproof » Fri Dec 08, 2006 8:50 am

Nice Save. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
GoofyUT
Posts: 1085
Joined: Sun Apr 09, 2006 9:45 am

BRAVO!!!!!!!!

Post by GoofyUT » Fri Dec 08, 2006 8:56 am

Once again, an AMAZINGLY helpful and useful post, very well done!!!

Thak you so much!!!

Chuck
People are dying every day in Darfur simply for who they are!!! PLEASE HELP THEM!
http://www.savedarfur.org

_______________________________

ignaciov
Posts: 2
Joined: Tue Jun 28, 2011 8:21 pm

Re: Remstar Classic E-25 Error

Post by ignaciov » Tue Jun 28, 2011 8:30 pm

I can't see the pictures of Snoredog explanation!
Please, Anybody can help me?

Thanks!

User avatar
Pugsy
Posts: 65207
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Remstar Classic E-25 Error

Post by Pugsy » Tue Jun 28, 2011 8:36 pm

ignaciov wrote:I can't see the pictures of Snoredog explanation!
Please, Anybody can help me?
Suggest you start a new thread with details of your problem.

Snoredog passed away and apparently image hosting site has closed the account.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Wulfman...

Re: Remstar Classic E-25 Error

Post by Wulfman... » Tue Jun 28, 2011 9:37 pm

I happen to have the instructions (including pictures) for the procedure Snoredog used to fix his machine......saved in both Word (.doc) and Adobe (.pdf) document files. Actually, the one I saved was from another thread he posted the instructions to.

The truth of the matter is that the outlet port kit to fix the broken part (hose nipple) can be purchased for about $3.00.......or, if you have the right connections, you can get the outlet port by itself for even less.


Den

ignaciov
Posts: 2
Joined: Tue Jun 28, 2011 8:21 pm

Re: Remstar Classic E-25 Error

Post by ignaciov » Wed Jun 29, 2011 12:01 pm

Wulfman... wrote:I happen to have the instructions (including pictures) for the procedure Snoredog used to fix his machine......saved in both Word (.doc) and Adobe (.pdf) document files. Actually, the one I saved was from another thread he posted the instructions to.

The truth of the matter is that the outlet port kit to fix the broken part (hose nipple) can be purchased for about $3.00.......or, if you have the right connections, you can get the outlet port by itself for even less.


Den
Wulfman, can you send me the files by email?
For another hand, where Can i buy the kit to fix the broken part?

thanks!!

Wulfman...

Re: Remstar Classic E-25 Error

Post by Wulfman... » Wed Jun 29, 2011 2:16 pm

ignaciov wrote:
Wulfman... wrote:I happen to have the instructions (including pictures) for the procedure Snoredog used to fix his machine......saved in both Word (.doc) and Adobe (.pdf) document files. Actually, the one I saved was from another thread he posted the instructions to.

The truth of the matter is that the outlet port kit to fix the broken part (hose nipple) can be purchased for about $3.00.......or, if you have the right connections, you can get the outlet port by itself for even less.


Den
Wulfman, can you send me the files by email?
For another hand, where Can i buy the kit to fix the broken part?

thanks!!
Sure. Please check your Private Messages (I need your e-mail address). I sent you one awhile ago. Just reply to it and include your e-mail address.


Den

nario
Posts: 1
Joined: Tue Aug 20, 2013 7:24 pm

Re: Remstar Classic E-25 Error

Post by nario » Tue Aug 20, 2013 7:34 pm

Wulfman... wrote:I happen to have the instructions (including pictures) for the procedure Snoredog used to fix his machine......saved in both Word (.doc) and Adobe (.pdf) document files. Actually, the one I saved was from another thread he posted the instructions to.

The truth of the matter is that the outlet port kit to fix the broken part (hose nipple) can be purchased for about $3.00.......or, if you have the right connections, you can get the outlet port by itself for even less.


Den

Den/Wulfman,

I realize your post is 2 years old -if you are still reading this, I'd like to get snoredog's instructions w pictures on them - and more importantly directions to buy the part(s)
Either doc or pdf would do.
Thank you kindly.

Wulfman...

Re: Remstar Classic E-25 Error

Post by Wulfman... » Tue Aug 20, 2013 8:23 pm

nario wrote:
Wulfman... wrote:I happen to have the instructions (including pictures) for the procedure Snoredog used to fix his machine......saved in both Word (.doc) and Adobe (.pdf) document files. Actually, the one I saved was from another thread he posted the instructions to.

The truth of the matter is that the outlet port kit to fix the broken part (hose nipple) can be purchased for about $3.00.......or, if you have the right connections, you can get the outlet port by itself for even less.


Den

Den/Wulfman,

I realize your post is 2 years old -if you are still reading this, I'd like to get snoredog's instructions w pictures on them - and more importantly directions to buy the part(s)
Either doc or pdf would do.
Thank you kindly.
Please check your PMs (Private Messages).


Den

.

User avatar
MikeyB
Posts: 61
Joined: Fri Nov 20, 2009 9:02 pm
Location: Lyles, Tennessee

Re: Remstar Classic E-25 Error

Post by MikeyB » Mon Apr 20, 2015 1:27 pm

Wulfman... wrote:I happen to have the instructions (including pictures) for the procedure Snoredog used to fix his machine......saved in both Word (.doc) and Adobe (.pdf) document files. Actually, the one I saved was from another thread he posted the instructions to.

The truth of the matter is that the outlet port kit to fix the broken part (hose nipple) can be purchased for about $3.00.......or, if you have the right connections, you can get the outlet port by itself for even less.


Den
Do you still have the file the machine I just got in the mail is having this error.
Started CPAP on 11-20-09
Pressure: 10, Ramp: No, C-Flex: 3 .
In Life Sometimes you only have two choices: NUT UP OR SHUTUP!!!!

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Remstar Classic E-25 Error

Post by Wulfman... » Mon Apr 20, 2015 1:34 pm

MikeyB wrote:
Wulfman... wrote:I happen to have the instructions (including pictures) for the procedure Snoredog used to fix his machine......saved in both Word (.doc) and Adobe (.pdf) document files. Actually, the one I saved was from another thread he posted the instructions to.

The truth of the matter is that the outlet port kit to fix the broken part (hose nipple) can be purchased for about $3.00.......or, if you have the right connections, you can get the outlet port by itself for even less.


Den
Do you still have the file the machine I just got in the mail is having this error.
Yes I do. If you need it, PM me with your email address.

I don't understand the last part about the "mail".


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Remstar Classic E-25 Error

Post by palerider » Mon Apr 20, 2015 1:53 pm

Wulfman... wrote: don't understand the last part about the "mail".
it's that disorganization that replaced the pony express for most of the country...

though it's not as fast, or as reliable at delivering things.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR