Erroneous AHI / OSA event flag(s)

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
fog.apnea
Posts: 38
Joined: Wed Aug 23, 2017 10:18 am
Location: Maryland, USA

Erroneous AHI / OSA event flag(s)

Post by fog.apnea » Wed Sep 13, 2017 7:54 pm

I was wondering (perhaps questioning) the accuracy of the xpap device's reporting of the AHI events based on my following observations and would like to hear feedback

- I have averaged 0.95 AHI overall in the past 80 days and I occasionally review overnight video recordings ((trending to less frequent as I have a good grasp on my sleep pattern now) of my sleep. I noticed that the few AHI events (mostly obstructive) that are flagged are pretty much while turning and rolling in the bed. I am sure there are real ones but am pointing out ones that aren't (likely). There are couple instances that I could clearly see breathing (can clearly see chest movement) even though obstructive event was recorded.
- I usually don't take naps but one day I tried to though ended up taking the hose off after 30 mins without actually dozing off. So obviously I didn't sleep for that period but SH showed an obstructive event and I was clearly awake.

Any comments on this finding?

_________________
MachineMask
Additional Comments: Software - SleepyHead
APAP pressure: 9.2-10.6
EPR 2cm
Humidity Level: Auto
Diagnosis: OSA 35 AHI. Diagnosed May '17. Therapy started Jun '17.

User avatar
tooter
Posts: 124
Joined: Fri Apr 06, 2012 2:00 pm
Location: SW Florida

Re: Erroneous AHI / OSA event flag(s)

Post by tooter » Wed Sep 13, 2017 7:57 pm

You obviously stopped breathing for 10+ seconds during that period.
Fixed pressure at 11

User avatar
Pugsy
Posts: 65267
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Erroneous AHI / OSA event flag(s)

Post by Pugsy » Wed Sep 13, 2017 8:11 pm

These machines don't have anyway to evaluate sleep status. They don't know if you are asleep or awake.
They only measure air flow.
Our awake breathing is actually fairly irregular when compared to asleep breathing and sometimes the air flow irregularities while awake can cause the machine to think the irregularity is some sort of apnea event. This is quite common and we call it SWJ or Sleep/Wake/Junk.
While awake we pause our breathing all the time and don't even realize we are doing....like sometimes we hold our breath when changing position in bed.

The machine will sometimes mistakeningly flag these awake breathing irregularities...not much we can do about it but shrug our shoulders and move on. Sometimes it's really obvious when there are flags right during a known awake time. Sometimes not so obvious.
People who have insomnia issues will often see a lot of SWJ flags due to repeatedly moving, tossing and turn, sighing...whatever.

It's just one of the limitations of the machine and we deal with. When I have it happen to me I just mentally remove those SWJ events from my evaluation of the therapy. For me my AHI is usually low anyway and even if they were real it's not that big of a deal.

Also sometimes you will see an OA flag but maybe the flow rate isn't totally flat...to meet the OA definition criteria the reduction in flow rate has to be at least 80% and last 10 seconds or more. So a person doesn't necessarily have to have no air moving at all...they can have as much as 20% movement and still get an OA flag.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: Erroneous AHI / OSA event flag(s)

Post by TedVPAP » Wed Sep 13, 2017 8:14 pm

fog.apnea wrote:I was wondering (perhaps questioning) the accuracy of the xpap device's reporting of the AHI events based on my following observations and would like to hear feedback

- I have averaged 0.95 AHI overall in the past 80 days and I occasionally review overnight video recordings ((trending to less frequent as I have a good grasp on my sleep pattern now) of my sleep. I noticed that the few AHI events (mostly obstructive) that are flagged are pretty much while turning and rolling in the bed. I am sure there are real ones but am pointing out ones that aren't (likely). There are couple instances that I could clearly see breathing (can clearly see chest movement) even though obstructive event was recorded.
- I usually don't take naps but one day I tried to though ended up taking the hose off after 30 mins without actually dozing off. So obviously I didn't sleep for that period but SH showed an obstructive event and I was clearly awake.

Any comments on this finding?
What are you looking at when you say that the event flag "are pretty much while turning and rolling", and "I could clearly see breathing"? Please share the data.
Also review the definition of event (help menu in sleepyhead or in your equipment manual) as apnea events are a reduction in breathing, not zero breathing.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Erroneous AHI / OSA event flag(s)

Post by palerider » Thu Sep 14, 2017 1:47 am

fog.apnea wrote:There are couple instances that I could clearly see breathing (can clearly see chest movement) even though obstructive event was recorded.
You see breathing EFFORT, but if the machine doesn't show air flow, then.... well, that's the *DEFINITION* of an obstructive apnea.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
fog.apnea
Posts: 38
Joined: Wed Aug 23, 2017 10:18 am
Location: Maryland, USA

Re: Erroneous AHI / OSA event flag(s)

Post by fog.apnea » Thu Sep 14, 2017 7:09 am

Pugsy wrote:These machines don't have anyway to evaluate sleep status. They don't know if you are asleep or awake.
They only measure air flow.
Our awake breathing is actually fairly irregular when compared to asleep breathing and sometimes the air flow irregularities while awake can cause the machine to think the irregularity is some sort of apnea event. This is quite common and we call it SWJ or Sleep/Wake/Junk.
While awake we pause our breathing all the time and don't even realize we are doing....like sometimes we hold our breath when changing position in bed.

The machine will sometimes mistakeningly flag these awake breathing irregularities...not much we can do about it but shrug our shoulders and move on. Sometimes it's really obvious when there are flags right during a known awake time. Sometimes not so obvious.
People who have insomnia issues will often see a lot of SWJ flags due to repeatedly moving, tossing and turn, sighing...whatever.

It's just one of the limitations of the machine and we deal with. When I have it happen to me I just mentally remove those SWJ events from my evaluation of the therapy. For me my AHI is usually low anyway and even if they were real it's not that big of a deal.

Also sometimes you will see an OA flag but maybe the flow rate isn't totally flat...to meet the OA definition criteria the reduction in flow rate has to be at least 80% and last 10 seconds or more. So a person doesn't necessarily have to have no air moving at all...they can have as much as 20% movement and still get an OA flag.
Thank you. Makes sense now and especially the irregularities as I looked at the data closely for that specific afternoon erroneous flag and the flow rate was down a bit (between 40-60% lasting 12 seconds) which explains why the machine flagged it. I'll keep this info in mind.

_________________
MachineMask
Additional Comments: Software - SleepyHead
APAP pressure: 9.2-10.6
EPR 2cm
Humidity Level: Auto
Diagnosis: OSA 35 AHI. Diagnosed May '17. Therapy started Jun '17.

User avatar
fog.apnea
Posts: 38
Joined: Wed Aug 23, 2017 10:18 am
Location: Maryland, USA

Re: Erroneous AHI / OSA event flag(s)

Post by fog.apnea » Thu Sep 14, 2017 7:13 am

palerider wrote:
fog.apnea wrote:There are couple instances that I could clearly see breathing (can clearly see chest movement) even though obstructive event was recorded.
You see breathing EFFORT, but if the machine doesn't show air flow, then.... well, that's the *DEFINITION* of an obstructive apnea.
Thanks for the clarification. What you mentioned is in-line with my response to pugsy's feedback above. I'll keep an eye on the finer level data if I have concern but for now my AHI is pretty low so not too concerned about impact of the actual event on my sleep quality but curious about what to expect from the device etc.. am only 2.5 months into the therapy .

_________________
MachineMask
Additional Comments: Software - SleepyHead
APAP pressure: 9.2-10.6
EPR 2cm
Humidity Level: Auto
Diagnosis: OSA 35 AHI. Diagnosed May '17. Therapy started Jun '17.

User avatar
fog.apnea
Posts: 38
Joined: Wed Aug 23, 2017 10:18 am
Location: Maryland, USA

Re: Erroneous AHI / OSA event flag(s)

Post by fog.apnea » Thu Sep 14, 2017 7:20 am

TedVPAP wrote:
What are you looking at when you say that the event flag "are pretty much while turning and rolling", and "I could clearly see breathing"? Please share the data.
SH 'Events Flag' graph in conjunction with IP camera video recording.
Just to note the xpap device, IP camera and computer system are all time synced down to the second.


Also review the definition of event (help menu in sleepyhead or in your equipment manual) as apnea events are a reduction in breathing, not zero breathing.
OK

_________________
MachineMask
Additional Comments: Software - SleepyHead
APAP pressure: 9.2-10.6
EPR 2cm
Humidity Level: Auto
Diagnosis: OSA 35 AHI. Diagnosed May '17. Therapy started Jun '17.

User avatar
fog.apnea
Posts: 38
Joined: Wed Aug 23, 2017 10:18 am
Location: Maryland, USA

Re: Erroneous AHI / OSA event flag(s)

Post by fog.apnea » Mon Sep 25, 2017 3:26 pm

anyone know of any device/attachment (or other suggestion) to monitor/detect actual breathing (chest movement)? When I did my home sleep study, as part of the kit provided by the sleep doc office it had an attachment (guess some sort of electrode) that needed to be glued right below and near where the collar bone, chest and neck intersect.

_________________
MachineMask
Additional Comments: Software - SleepyHead
APAP pressure: 9.2-10.6
EPR 2cm
Humidity Level: Auto
Diagnosis: OSA 35 AHI. Diagnosed May '17. Therapy started Jun '17.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Erroneous AHI / OSA event flag(s)

Post by Julie » Mon Sep 25, 2017 3:52 pm

What are you trying to assess, why?

User avatar
fog.apnea
Posts: 38
Joined: Wed Aug 23, 2017 10:18 am
Location: Maryland, USA

Re: Erroneous AHI / OSA event flag(s)

Post by fog.apnea » Tue Sep 26, 2017 6:47 am

xxyzx wrote:
fog.apnea wrote:anyone know of any device/attachment (or other suggestion) to monitor/detect actual breathing (chest movement)? When I did my home sleep study, as part of the kit provided by the sleep doc office it had an attachment (guess some sort of electrode) that needed to be glued right below and near where the collar bone, chest and neck intersect.
the IHT has little tubes in the nose to do it to measure air flow
the xpap does the same thing to measure air flow

the IHT uses the belt to monitor breathing effort / chest movement
the xpap does not do that channel directly
it tells by the air flow
and if there is no air flow and it thinks the airway is open it assumes CA with no breathing effort
Thanks xxyxz. I am really looking into what options are available to try rather than how it works.

_________________
MachineMask
Additional Comments: Software - SleepyHead
APAP pressure: 9.2-10.6
EPR 2cm
Humidity Level: Auto
Diagnosis: OSA 35 AHI. Diagnosed May '17. Therapy started Jun '17.

User avatar
fog.apnea
Posts: 38
Joined: Wed Aug 23, 2017 10:18 am
Location: Maryland, USA

Re: Erroneous AHI / OSA event flag(s)

Post by fog.apnea » Tue Sep 26, 2017 6:48 am

Julie wrote:What are you trying to assess, why?
I monitored myself with overnight sleep recordings especially in the beginning of the therapy and noticed that all AHI events in the SH correlate to the rolling over in the bed. And based on the earlier responses in the thread they seem to be SWJ. There are times I zoomed in during those events and saw that I was continuing breathing but because of low lighting recording it is very difficult to be sure. Oximeter can record O2 levels but breathing cessation cannot be deducted from the data.

I have another thread viewtopic.php?f=1&t=156504&st=0&sk=t&sd=a regarding my relapsing symptoms. My AHI is pretty much <1 all the time but as outlined in the other thread I have some symptoms that are concerning. My doc indicated it could non-apnea related. So am attempting to use the info to fine tune my therapy, isolate issues that may be impacting the sleep quality in that regards OR be sure that problems that am experiencing are indeed non-apea related like the doc indicated.

Hope I answered your question.

_________________
MachineMask
Additional Comments: Software - SleepyHead
APAP pressure: 9.2-10.6
EPR 2cm
Humidity Level: Auto
Diagnosis: OSA 35 AHI. Diagnosed May '17. Therapy started Jun '17.

User avatar
TASmart
Posts: 1071
Joined: Wed Jan 04, 2017 7:23 pm
Location: Eugene, OR

Re: Erroneous AHI / OSA event flag(s)

Post by TASmart » Tue Sep 26, 2017 2:58 pm

xxyzx wrote:
fog.apnea wrote:anyone know of any device/attachment (or other suggestion) to monitor/detect actual breathing (chest movement)? When I did my home sleep study, as part of the kit provided by the sleep doc office it had an attachment (guess some sort of electrode) that needed to be glued right below and near where the collar bone, chest and neck intersect.
the IHT has little tubes in the nose to do it to measure air flow
the xpap does the same thing to measure air flow

the IHT uses the belt to monitor breathing effort / chest movement
the xpap does not do that channel directly
it tells by the air flow
and if there is no air flow and it thinks the airway is open it assumes CA with no breathing effort
I know of no xPAP that uses a "little tubes in the nose" to measure airflow. MIne like most, use an Internally mass flow device located at device inlet, (Bold added for emphasis)
digital mass flow type, -70 to +180 L/min.
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very