Pressure Changes

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Lahree

Pressure Changes

Post by Lahree » Tue Sep 05, 2017 5:25 am

My CPAP machine is a DreamStation Auto CPAP. When I originally got it, my pressure range was set 4 - 20. About 2 weeks ago, I asked my doctor to allow a reset to a minimum pressure of 7, because I felt like at 4, I was really struggling for air. I was told that happened and when I turn on the machine, it now says "7" as the pressure. My questions about this are:

Do I need to/and if so, how do I --change settings on Dreammapper and Sleepyhead to show the reset range?

From my stats last night, I have a reading on both programs that said my minimum pressure was 5? How did I get below the lowest reset of 7?

Thanks for all your help!

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: Pressure Changes

Post by TedVPAP » Tue Sep 05, 2017 6:29 am

Lahree wrote:My CPAP machine is a DreamStation Auto CPAP. When I originally got it, my pressure range was set 4 - 20. About 2 weeks ago, I asked my doctor to allow a reset to a minimum pressure of 7, because I felt like at 4, I was really struggling for air. I was told that happened and when I turn on the machine, it now says "7" as the pressure. My questions about this are:

Do I need to/and if so, how do I --change settings on Dreammapper and Sleepyhead to show the reset range?

From my stats last night, I have a reading on both programs that said my minimum pressure was 5? How did I get below the lowest reset of 7?

Thanks for all your help!
I have the same machine. Sleepyhead will report what the machine settings are; you do not need to do anything.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

User avatar
Pugsy
Posts: 65313
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Pressure Changes

Post by Pugsy » Tue Sep 05, 2017 7:11 am

Do you have any sort of Flex exhale relief turned on?
If you do that will show up as the minimum because during exhale (EPAP) the pressure will drop from 7 to 5 (or whatever based on the setting).