Dreamstation GO - need help with set up

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Cardsfan
Posts: 1509
Joined: Wed Apr 29, 2015 3:02 pm
Location: Close to St. Louis, MO

Dreamstation GO - need help with set up

Post by Cardsfan » Mon Jul 24, 2017 2:13 pm

Just received my Dreamstation GO with battery. I have the clinician manual, and got it 99% set up. There is a setting that I don't understand. It is MASK TYPE System One Resistance Control, and it is a setting from 1 to 5. I believe this is similar to Resmed mask type, but the Resmed choice is clear as it states the mask style. All this machine states is pick a number 1 through 5, and it has to be a Phillips mask.
So, I called Phillips, no help, but to tell me to buy their mask.
I use a Resmed Swift FX Nano nasal mask. What setting should I use?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments:  CPAP 10 cmH20., User since 1/1/15.

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Dreamstation GO - need help with set up

Post by Pugsy » Mon Jul 24, 2017 2:25 pm

That's not the same as the ResMed mask selection.
Do you have a setting for 0 or off? That's what we normally use with the Respironics machines when using a non Respironics mask. Just turn it off.

This is NOT a critical setting and does nothing in terms of impacting leak calculations or anything like that.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
kbh209
Posts: 339
Joined: Wed Dec 30, 2015 9:12 pm

Re: Dreamstation GO - need help with set up

Post by kbh209 » Mon Jul 24, 2017 2:27 pm

I believe this is the information you might be looking for. http://incenter.medical.philips.com/doc ... ernum%3d-2
Phillips Dreamstation 2 Auto CPAP with Humidifier **

Cardsfan
Posts: 1509
Joined: Wed Apr 29, 2015 3:02 pm
Location: Close to St. Louis, MO

Re: Dreamstation GO - need help with set up

Post by Cardsfan » Mon Jul 24, 2017 2:29 pm

Pugsy, Thank you. Yes, there is an OFF setting option. I'll just turn it off.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments:  CPAP 10 cmH20., User since 1/1/15.
Last edited by Cardsfan on Mon Jul 24, 2017 2:32 pm, edited 1 time in total.

Cardsfan
Posts: 1509
Joined: Wed Apr 29, 2015 3:02 pm
Location: Close to St. Louis, MO

Re: Dreamstation GO - need help with set up

Post by Cardsfan » Mon Jul 24, 2017 2:31 pm

KBH- yes, that is the info, but it only lists Phillips brand masks. Mine is Resmed brand. Thanks.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments:  CPAP 10 cmH20., User since 1/1/15.

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Dreamstation GO - need help with set up

Post by Goofproof » Tue Jul 25, 2017 3:45 pm

xxyzx wrote:
Pugsy wrote:That's not the same as the ResMed mask selection.
Do you have a setting for 0 or off? That's what we normally use with the Respironics machines when using a non Respironics mask. Just turn it off.

This is NOT a critical setting and does nothing in terms of impacting leak calculations or anything like that.
========

how does the machine know how to deliver the right pressure without knowing the resistance/loss due to the mask type ?
If you bought a real XPAP, it wouldn't matter. The sensors would do their job and control the XPAP, no games required. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Dreamstation GO - need help with set up

Post by chunkyfrog » Tue Jul 25, 2017 3:55 pm

I have long suspected that the Respironics mask numbers are a crock.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

Cardsfan
Posts: 1509
Joined: Wed Apr 29, 2015 3:02 pm
Location: Close to St. Louis, MO

Re: Dreamstation GO - need help with set up

Post by Cardsfan » Tue Jul 25, 2017 6:32 pm

I set mask type at zero. I used it last night 7.5 hours, with an AHI of 1.0. Which is what my AHI is normally. So I think that mask type didn't make much difference in my case.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments:  CPAP 10 cmH20., User since 1/1/15.

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Dreamstation GO - need help with set up

Post by Pugsy » Tue Jul 25, 2017 6:51 pm

I tried all the various Resistance Settings, including OFF which is what Respironics advises for non Respironics masks, over a period of several nights and I couldn't spot any difference in anything. Not how it felt or my usual results.
If there was any difference in the pressure it was so tiny that I couldn't feel it and it sure didn't impact the data anywhere that I could spot.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.