I am going camping in a remote location. I have a Remstar Solo Lx. I will be out for 3-6 days with no electricity at all. I am considering purchasing a battery pack from cpap.com, but I need to recharge it after the battery has depleted. Has anyone used a solar power device to recharge battery? Does anyone know if this will work? Any recommendations?
Has anyone used a solar power device to recharge battery?
To get a solar power charger big enough to be effective would cost quite a bit. Solar power has been around for over 35 years, but hasn't really became cost effective. Using a small generator to recharge is the most effective, no power grid method to recharge, Two large Deep Cycle Batteries would get you a week without recharging. Jim
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
I've researched solar a bit
To use solar, you'll need a minimum amount of daytime sunlight, solar gear that will handle the power demands to recharge the battery in a day, and $ to buy the equipment. Everything will weigh a certain amount, mostly the battery(s). Don't use the heating circuit for the humidifier.
Recharging a small 20-amp-hour or so battery with adequate sunlight will require a larger panel and charging circuit that may run $400 or so. You can google around to find some options.
If weight is not an issue (you can haul the stuff in and out easily), I would purchase one or more deep cycle marine batteries of sufficient capacity to cover your 3-6 nights. Find out how many DC amps your CPAP draws (measure it so you know for sure). Then figure that amount (amps) times 8 hours= X amp-hours per night. Then multiply that answer by six to get the total ampere-hours needed. Then add 40-50% more if you want to be able to improve the life of your battery set-up. Then buy one or more batteries that combined will give you the total amp-hour number you calculated.
If $ were not an issue, buy the BreatheX and two spare batteries ~ $900. This gives you a light weight and very compact solution. My camping/travel requirement limits the size and weight, so I have the BreatheX and it works nicely.
Happy camping!
Recharging a small 20-amp-hour or so battery with adequate sunlight will require a larger panel and charging circuit that may run $400 or so. You can google around to find some options.
If weight is not an issue (you can haul the stuff in and out easily), I would purchase one or more deep cycle marine batteries of sufficient capacity to cover your 3-6 nights. Find out how many DC amps your CPAP draws (measure it so you know for sure). Then figure that amount (amps) times 8 hours= X amp-hours per night. Then multiply that answer by six to get the total ampere-hours needed. Then add 40-50% more if you want to be able to improve the life of your battery set-up. Then buy one or more batteries that combined will give you the total amp-hour number you calculated.
If $ were not an issue, buy the BreatheX and two spare batteries ~ $900. This gives you a light weight and very compact solution. My camping/travel requirement limits the size and weight, so I have the BreatheX and it works nicely.
Happy camping!