Whistling sound
-
090810
Whistling sound
Good Morning!
I am using a Respirsonic Dreamstation Pro and a ResMed p10 nasal mask. Last night there was a whistling sound, I unhooked the tube from the machine since it had water in in. It stopped when I replaced the hose but only for a couple minutes then started again. This went on for a couple of hours, I tried everything I could think of to no avail. Finally I just turned the the machine off and finished the night without it. Anyone else experience this or and ideas as to what would cause this. Thank you.
I am using a Respirsonic Dreamstation Pro and a ResMed p10 nasal mask. Last night there was a whistling sound, I unhooked the tube from the machine since it had water in in. It stopped when I replaced the hose but only for a couple minutes then started again. This went on for a couple of hours, I tried everything I could think of to no avail. Finally I just turned the the machine off and finished the night without it. Anyone else experience this or and ideas as to what would cause this. Thank you.
Re: Whistling sound
Do you think the noise was from the machine itself, from the hose or from the mask? If you have a pinhole in the hose, duct tape will take care of it. If a connection between the hose and machine is off, try reseating it again. If it's the mask, see if you can trace the noise more exactly. If it happens when the machine's running but the hose disengaged - if it's internal - you might need to see the dealer.
Re: Whistling sound
I don't see to be having much luck with my Dreamstation either.
It is approx 9 mths old. After 3 months there was a clicking sound which seemed to be linked to the humidifier so the humidifier was replaced. Philips told me that 1 in 10-15 machines has the same problem and not to worry as they would just keeping replacing it under warranty! In Australia you can ring Philips direct but still need to return machine through the supplier.
New humidifier started making clicking sounds at 3mths too so Philips replaced the whole machine.
The new machine (approx 6 weeks) seems to be making groaning noises (as am I)during the night when the pressure is high. I am loathe to go back to the supplier AGAIN so will give it another week or so and then will take it back if it is still making noises.
I have had clicking and now groaning but no whistling
It is approx 9 mths old. After 3 months there was a clicking sound which seemed to be linked to the humidifier so the humidifier was replaced. Philips told me that 1 in 10-15 machines has the same problem and not to worry as they would just keeping replacing it under warranty! In Australia you can ring Philips direct but still need to return machine through the supplier.
New humidifier started making clicking sounds at 3mths too so Philips replaced the whole machine.
The new machine (approx 6 weeks) seems to be making groaning noises (as am I)during the night when the pressure is high. I am loathe to go back to the supplier AGAIN so will give it another week or so and then will take it back if it is still making noises.
I have had clicking and now groaning but no whistling
_________________
| Machine: DreamStation Auto CPAP Machine |
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
- upforsleep
- Posts: 51
- Joined: Mon Sep 28, 2009 6:13 am
- Location: Texas-South East
Re: Whistling sound
This happened on my Dreamstation once. Cleaning and changing both filters worked for me.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Simplus Full Face CPAP Mask with Headgear |
| Additional Comments: Cpap Pillow, HoseBuddy, soft cervical collar, mask liner |
-
Guest
Re: Whistling sound
Which mask?090810 wrote:Good Morning!
I am using a Respirsonic Dreamstation Pro and a ResMed p10 nasal mask. Last night there was a whistling sound, I unhooked the tube from the machine since it had water in in. It stopped when I replaced the hose but only for a couple minutes then started again. This went on for a couple of hours, I tried everything I could think of to no avail. Finally I just turned the the machine off and finished the night without it. Anyone else experience this or and ideas as to what would cause this. Thank you.
Whistling is caused by air moving (past something) - check your mask exhale port for restrictions of the small holes allowing air to escape. Depending on the size of the holes you can clear them with something like a toothpick or a cotton swab. The exhale holes can get clogged with gunk from time to time making the holes smaller and causing a whistle.
Re: Whistling sound
+1 on this! Check the filter.upforsleep wrote:This happened on my Dreamstation once. Cleaning and changing both filters worked for me.
-
Diamondminek
- Posts: 100
- Joined: Mon Sep 14, 2015 11:30 am
Re: Whistling sound
You can get whistling from condesation/ rain out - you mention water in the tube so I'd guess that's what it is. Things like a hose cover, heated hose, room temp, machine and hose position can help.
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
- CarpeNoctum
- Posts: 116
- Joined: Tue Nov 01, 2011 7:40 pm
- Location: Ilwaco WA
Re: Whistling sound
Hi folks,
I was getting a whistling sound from my old Legacy Respironics. Tracked it down to poor fit where the humidifier chamber connects to the machine. Solved it with some teflon joint tape. Inspect your connection points...could be a similar issue.
CN
I was getting a whistling sound from my old Legacy Respironics. Tracked it down to poor fit where the humidifier chamber connects to the machine. Solved it with some teflon joint tape. Inspect your connection points...could be a similar issue.
CN
- Wulfman...
- Posts: 6692
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: Whistling sound
You need to spend more time with it when you're not trying to sleep.......to figure out what the specific problem is.090810 wrote:Good Morning!
I am using a Respirsonic Dreamstation Pro and a ResMed p10 nasal mask. Last night there was a whistling sound, I unhooked the tube from the machine since it had water in in. It stopped when I replaced the hose but only for a couple minutes then started again. This went on for a couple of hours, I tried everything I could think of to no avail. Finally I just turned the the machine off and finished the night without it. Anyone else experience this or and ideas as to what would cause this. Thank you.
Could be any number of things, but YOU need to figure out what is causing it. We can't see your setup or what's happening.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
- DreamStalker
- Posts: 7503
- Joined: Mon Aug 07, 2006 9:58 am
- Location: Nowhere & Everywhere At Once
Re: Whistling sound
I used teflon tape with my Swift LT mask (about 10 years old) where the hose fits into the pillow frame due to whistling. Recently however I started thinking that maybe breathing in teflon dust may not be such a good idea? Because the teflon tape has worn over time to small shreds.CarpeNoctum wrote:Hi folks,
I was getting a whistling sound from my old Legacy Respironics. Tracked it down to poor fit where the humidifier chamber connects to the machine. Solved it with some teflon joint tape. Inspect your connection points...could be a similar issue.
CN
Soooo .... I may have to break down and by a new mask.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.
- CarpeNoctum
- Posts: 116
- Joined: Tue Nov 01, 2011 7:40 pm
- Location: Ilwaco WA
Re: Whistling sound
DreamStalker,DreamStalker wrote:I used teflon tape with my Swift LT mask (about 10 years old) where the hose fits into the pillow frame due to whistling. Recently however I started thinking that maybe breathing in teflon dust may not be such a good idea? Because the teflon tape has worn over time to small shreds.CarpeNoctum wrote:Hi folks,
I was getting a whistling sound from my old Legacy Respironics. Tracked it down to poor fit where the humidifier chamber connects to the machine. Solved it with some teflon joint tape. Inspect your connection points...could be a similar issue.
CN
Soooo .... I may have to break down and by a new mask.
First off, I love your avatar...reminds me of a Huichol yarn painting.
Teflon is very inert and designed for permanent use. I've never seen it deteriorate in a pipe fitting environment. And it should be installed so that it's never exposed in the interior of a pipe. Where I used it, there was no opportunity for it to fragment inside the cpap pressure system. And it's cheap enough to replace every now and then.
But I take your point. Might not be perfect for every situation.
CN
Re: Whistling sound
I once had a whistling/ sputtering noise, although I have a Resmed Airsense 10 Apap. I found that it was the moisture's fault in the hosing. I wrapped my house with a towel in the area where moisture was collecting, and I turned the temp of my hose up a several degrees above my room temp. Also turned my humidifier down a step. After these quick adjustments, The weird whistling was gone.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: Apap 8 - 11, Autoramp, Sleepyhead software |
Blake
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Whistling sound
Metallica (sp), cool username!
Now you have to pick a different password!
Now you have to pick a different password!
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: Whistling sound
No it's just my gamertag from xbox. I never have trouble making SN with it, so it is my go to haha.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: Apap 8 - 11, Autoramp, Sleepyhead software |
Blake
Re: Whistling sound
I discovered that when mine starts whistling I need to clean the gray foam filter and change the white filter.