S9 invalid data card issue

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
jaybee72
Posts: 159
Joined: Fri Aug 31, 2007 6:06 pm
Location: Atlanta area

S9 invalid data card issue

Post by jaybee72 » Thu Apr 13, 2017 7:43 pm

Hello - I've searched some on this, but while getting many results I don't see my particular issue coming up. Last fall my S9 auto suddenly started showing the "invalid data card" error, and I simply pulled the card out. Now that I'm looking at an upcoming physical, and need my doc to write a new prescription (last one is about 7 years old and DME won't take it) I would like to gather some data to carry in with me. When I got him to write that last one he gave me grief about it being time for a new sleep study, since the only one I had was in 2007. I convinced him instead to prescribe the auto, and that I'd watch the data myself.

Anyway, tried to clear the card and start over. After a format on my pc I could insert it in the machine, allow it to "erase" the card and then download summary data. If leave it in a couple of days I get detailed data, too. But, following all the safety procedures to pull it into sleepyhead like locking it, properly ejecting, etc the S9 gets angry when I put it back in. It gives the invalid message with the option of ejecting or erasing, but errors out when you tell it to erase. The only way to make it work again is to reformat on the PC, and then it will erase in the machine and start taking data.

I bought a new card and have exactly the same results.

Could Win10 still be writing a hidden file or something even though its write protected?

So, I can get data in a very cumbersome way, but why is it not just working right?

Thank you!

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Haven't used humidifier in at least 5 years! On the hose since Sept 2007 with less than 10 missed nights

jaybee72
Posts: 159
Joined: Fri Aug 31, 2007 6:06 pm
Location: Atlanta area

Re: S9 invalid data card issue

Post by jaybee72 » Tue Apr 18, 2017 8:30 am

I figured that others might have run into this. Any suggestions anyone?

Thank you

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Haven't used humidifier in at least 5 years! On the hose since Sept 2007 with less than 10 missed nights

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: S9 invalid data card issue

Post by prodigyplace » Tue Apr 18, 2017 8:40 am

Windows *should* not be writing data, but it might be possible.
Many years ago, I was working with a virus infected Mac computer. Even though I used a write-protected (locked) floppy disk to clean the computer, the virus wrote it's infection to the disk. The program I ran cleaned the virus off the original computer though.

I guess a test would be to put the locked card in your computer, open file eplorer, & try to copy a file to the card. If it copies, the computer is ignoring the fact the card is locked. A different card reader might behave differently.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

User avatar
Pugsy
Posts: 65417
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: S9 invalid data card issue

Post by Pugsy » Tue Apr 18, 2017 8:54 am

I would be wondering if the little lock thingy you have to slide on the SD card to write protect the card is working properly.
If it was me I would probably get a new SD card and see if the problem persists with a new card.

Alternately there is a way to remove the Windows file that the S9 is seeing and not liking.
I had to do it a couple of times myself because I would forget to write protect my card after I upgraded to Windows 10.


rmdir "System Volume Information" /s /q

at the contents of the SD card window hold down the shift key and right click mouse...and choose Open command window and paste the above in the black window.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: S9 invalid data card issue

Post by palerider » Tue Apr 18, 2017 9:55 am

jaybee72 wrote:Could Win10 still be writing a hidden file or something even though its write protected?
it could be, try sticking a SD card into your card reader, and try writing something on it when it's write protected.

the slider on a sd card is just a plastic tab, there should be a switch in the card reader that detects that tab, a few card readers don't have the write protect switch, and like all switches, it could go bad.

you can try telling windows not to write anything to removable media:
http://www.thewindowsclub.com/prevent-s ... folder-usb

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

jaybee72
Posts: 159
Joined: Fri Aug 31, 2007 6:06 pm
Location: Atlanta area

Re: S9 invalid data card issue

Post by jaybee72 » Tue Apr 18, 2017 8:30 pm

Ok, cool - lots of suggestions to try now.

Thank you!!!

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Haven't used humidifier in at least 5 years! On the hose since Sept 2007 with less than 10 missed nights