Rainout - Some Thoughts

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Rainout - Some Thoughts

Post by JDS74 » Sun Apr 02, 2017 5:32 am

Rainout is a symptom of the relative humidity of the air within the CPAP system being above the dew point of the surfaces of the hose and mask components. If the dew point is too low, condensation occurs and you can get rainout. The rainout can occur in the CPAP hose or in the mask components or both.

Fixes:
Lower the humidity in the system until the rainout stops. For some this works nicely, for others, the lower humidity causes issues with dry mouth or dry nasal passages.

Increase the dew point in the system by raising the temperature of the hose, etc. Using a heated hose is also another option. Here are some from various manufacturers.
Edit:

Hybernite Heayted Hose - works with and external humidifier.
Climateline for AirCurve10
for Respironics Dreamstation

Raising room temperature will raise the temperature of the hose surfaces and that may be enough to solve the problem. Insulating the CPAP hose with fleece fabric or a Hose Cozy may be enough.

Snuggle Skin Hose Wrap
Respironics Hose Wrap
Resmed Hose wrap

Sometimes, routing the hose under the bed covers is enough. Sometimes routing the hose with a hose hanger allows the condensation to drain back to the humidifier will do the trick.
Amazon hose hanger
DIY Hose Hanger



Rainout within the mask components is harder to fix depending on the mask. If the issue is the short hose from the mask to the CPAP hose, the fix is to insulate that short hose. If the issue is condensation in the mask itself, then some form of insulation will be required because the humidity level is controlled by the exhaled air.
Pad-A-Cheek Barrel Cozies

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Last edited by JDS74 on Sun Apr 02, 2017 8:39 am, edited 2 times in total.
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: Rainout - Some Thoughts

Post by prodigyplace » Sun Apr 02, 2017 7:42 am

Another solution I used to raise the temperature of the hose was tho switch to a heated hose. I cannot provide the link, because this solution is model-specific.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

User avatar
LSAT
Posts: 13408
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Rainout - Some Thoughts

Post by LSAT » Sun Apr 02, 2017 8:15 am

prodigyplace wrote:Another solution I used to raise the temperature of the hose was tho switch to a heated hose. I cannot provide the link, because this solution is model-specific.
That's what JDS74 suggested.......he wrote....." Using a heated hose is also another option."