Charts after 1 month

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
bt_rox88
Posts: 13
Joined: Sat Jan 28, 2017 9:22 am

Charts after 1 month

Post by bt_rox88 » Sat Mar 04, 2017 9:37 am

I think I am right at about 1 month of using the CPAP machine. Overall I have had positive results including less afternoon irritability, less need for my ADD medication, and feeling more rested many days. The previous time that I posted a chart, it was suggested that I change my minimum pressure to 8, and that did seem to help.

Currently, there are definitely better nights than others, and I am looking for advice to clean up my charts a little more so that I can get more consistency. On a side note, it seems that I have best results when I go to bed more tired.

Here are my last weeks results, I would appreciate any feedback:
http://imgur.com/a/QdTn4
Machine: Phillips Respironics Dreamstation
Mask: Phillips Respironics Nuance Pro

User avatar
Pugsy
Posts: 65293
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Charts after 1 month

Post by Pugsy » Sat Mar 04, 2017 9:50 am

Maybe try 8.5 or 9 minimum?
You don't urgently need to clean things up. Some nights are a bit cluttered mainly with snores which could be related to sleep position or REM sleep. Snores are usually easily cleaned up with just a little more pressure.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
OkyDoky
Posts: 2869
Joined: Mon Aug 25, 2014 5:18 pm

Re: Charts after 1 month

Post by OkyDoky » Sat Mar 04, 2017 10:21 am

In all the screenshots it does look like a minimum pressure of 9 cm would better control the snores and events. If you want to go 8.5 first and see if that gets you where you want, it might. Up to you.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

bt_rox88
Posts: 13
Joined: Sat Jan 28, 2017 9:22 am

Re: Charts after 1 month

Post by bt_rox88 » Sat Mar 04, 2017 10:26 am

Thank you both, I really appreciate the advice.
Machine: Phillips Respironics Dreamstation
Mask: Phillips Respironics Nuance Pro

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Charts after 1 month

Post by palerider » Sat Mar 04, 2017 11:27 am

Pugsy wrote:Maybe try 8.5 or 9 minimum?
You don't urgently need to clean things up. Some nights are a bit cluttered mainly with snores which could be related to sleep position or REM sleep. Snores are usually easily cleaned up with just a little more pressure.
+1