Airsense 10 problem

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
MeToo
Posts: 97
Joined: Thu Apr 28, 2005 12:18 pm
Location: Alabama

Airsense 10 problem

Post by MeToo » Wed Mar 01, 2017 12:21 pm

Last night, about 5 seconds after I put my mask on and and started the machine, it stopped. I couldn't get it to respond in any way. Checked the surge protector and it works fine. The A10 is dead, as far as I can tell. Does anyone know what might be wrong? I looked for a fuse that I could reset but couldn't find anything. I'd rather not buy another machine. Any ideas? Many thanks.

_________________
Machine
Additional Comments: Pressure Setting 8.4-10.6, flonase, buckwheat-Millet Pillow. side sleeper

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: Airsense 10 problem

Post by grayghost4 » Wed Mar 01, 2017 12:37 pm

check the plug into the machine ... is it in firmly ? Does the screen show any light at all ?
Most likley the power brick that failed.
Go to the DME and try it with a new power brick.
If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

User avatar
MeToo
Posts: 97
Joined: Thu Apr 28, 2005 12:18 pm
Location: Alabama

Re: Airsense 10 problem

Post by MeToo » Wed Mar 01, 2017 12:41 pm

Thank you! Yes, I tried unplugging and plugging in several times...no power indication at all.

Unfortunately, I bought it through CPAP.COM. I also have an S9, I wonder if the brick for that will work. I'll research it.

_________________
Machine
Additional Comments: Pressure Setting 8.4-10.6, flonase, buckwheat-Millet Pillow. side sleeper

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: Airsense 10 problem

Post by grayghost4 » Wed Mar 01, 2017 12:48 pm

no the S9 power brick has a different plug on the machine end
If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Airsense 10 problem

Post by Julie » Wed Mar 01, 2017 12:51 pm

Have you tried another appliance in that outlet?

User avatar
ChicagoGranny
Posts: 15473
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Airsense 10 problem

Post by ChicagoGranny » Wed Mar 01, 2017 12:53 pm

MeToo wrote:Unfortunately, I bought it through CPAP.COM.
2-year warranty expired?

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: Airsense 10 problem

Post by grayghost4 » Wed Mar 01, 2017 12:55 pm

do you have a DVOM to test the power brick ?

http://www.ebay.com/itm/112316893695?_t ... EBIDX%3AIT
If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

User avatar
MeToo
Posts: 97
Joined: Thu Apr 28, 2005 12:18 pm
Location: Alabama

Re: Airsense 10 problem

Post by MeToo » Wed Mar 01, 2017 1:14 pm

Thanks all.
Yes, my backup ran fine in that outlet, so that's not it.
Yes, two year warranty expired just 3 weeks ago!
I'm not sure what a DVOM is but I bet I can find one.
Thanks again!

_________________
Machine
Additional Comments: Pressure Setting 8.4-10.6, flonase, buckwheat-Millet Pillow. side sleeper

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: Airsense 10 problem

Post by grayghost4 » Wed Mar 01, 2017 1:16 pm

Digital volt ohm meter just a way to test the power brick
available at Radio shack also


http://www.ebay.com/itm/DT830B-Digital- ... SwUKxYiM3f
If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

User avatar
LSAT
Posts: 13397
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Airsense 10 problem

Post by LSAT » Wed Mar 01, 2017 1:25 pm

MeToo wrote:Thank you! Yes, I tried unplugging and plugging in several times...no power indication at all.

Unfortunately, I bought it through CPAP.COM. I also have an S9, I wonder if the brick for that will work. I'll research it.
Why is that unfortunate?....call them. It may be covered by warranty.
Also...try calling these people and see if they have a suggestion. ...http://acbio.com/cpap-repair/
Last edited by LSAT on Thu Nov 09, 2017 6:22 am, edited 2 times in total.

User avatar
OkyDoky
Posts: 2869
Joined: Mon Aug 25, 2014 5:18 pm

Re: Airsense 10 problem

Post by OkyDoky » Wed Mar 01, 2017 1:26 pm

One other thing you might try is to remove the SD card and see if it will power up.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
LoBattery
Posts: 158
Joined: Fri Nov 25, 2016 9:20 am

Re: Airsense 10 problem

Post by LoBattery » Wed Mar 01, 2017 2:23 pm

Is the power LED on on the power module? Seems a lot of those go bad. You can go to Harbor Freight and get a free DVM with a coupon. Measure voltage from outside metal to inside of the barrel wall. Should be 24V. You will need something like a paper clip to probe the inside. Try not to short to the center pin which should be 3.3V. I'd consider buying a dead machine if the quotes are too terrible.
Seeing and believing are often both wrong. FOW

User avatar
ChicagoGranny
Posts: 15473
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Airsense 10 problem

Post by ChicagoGranny » Wed Mar 01, 2017 4:35 pm

LoBattery wrote:You can go to Harbor Freight and get a free DVM with a coupon.
Well, I'll be darned - http://www.harborfreight.com/digitalsavings.html Free with purchase.

Friar Jim

Re: Airsense 10 problem

Post by Friar Jim » Thu Nov 09, 2017 6:11 am

Watch out for big problem. Machine made in Australia has wrong polarity for America!!!! Very nasty and thoughtless that they don't put out a big caution for this when selling in America.

My machine fried because of this problem. Who would check Polaity or think to check if machine is made in Australia. Very poor thinking by Resmed. Spread the word to save on frying machines.