Flecks in Humidifer Tank

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
MaxDarkside
Posts: 1199
Joined: Sun Dec 18, 2011 4:21 pm
Location: Minneapolis, MN

Flecks in Humidifer Tank

Post by MaxDarkside » Wed Jan 18, 2017 10:33 pm

Hi All,

Long time no post. I noticed light colored flecks with edges and pointy corners in my humidifier tank (Resmed S9 purchased Nov 2011) so I of course cleaned it, by hand, dismantling the tank as it is a pops-open dishwasher safe tank, disinfected it, buffed 'er to a shine inside, filled it with distilled water. The next morning... more flecks. Is this one of the failure modes of the Resmed S9, like some seal inside coming apart?

I replace the filter quite often, using CPAP.COM filters.

Thanks!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50E Oximeter, Zeo Bedside, Software: Comm'l grade AI analytics server & tools & SleepyHead
Do or Die... Sleep Apnea killed me, but I came back. Click for my story
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: Flecks in Humidifer Tank

Post by grayghost4 » Thu Jan 19, 2017 1:41 pm

There is nothing inside that would come apart and cause that .... silicone gasket ... black foam rubber .... all the rest is hard plastic.
Might it be something in the distilled water ?
If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Flecks in Humidifer Tank

Post by chunkyfrog » Thu Jan 19, 2017 1:48 pm

I wonder if flakes are coming off the plastic parts, due to age.
How old is the tank?

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Wulfman...
Posts: 6689
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Flecks in Humidifer Tank

Post by Wulfman... » Thu Jan 19, 2017 2:20 pm

MaxDarkside wrote:Hi All,

Long time no post. I noticed light colored flecks with edges and pointy corners in my humidifier tank (Resmed S9 purchased Nov 2011) so I of course cleaned it, by hand, dismantling the tank as it is a pops-open dishwasher safe tank, disinfected it, buffed 'er to a shine inside, filled it with distilled water. The next morning... more flecks. Is this one of the failure modes of the Resmed S9, like some seal inside coming apart?

I replace the filter quite often, using CPAP.COM filters.

Thanks!
Where ya been, Max?

Floating on the top or lying on the bottom of the tank?
In any case, they won't be carried out of the HH tank with the water vapor......too heavy.
Maybe try filtering them out on some clean, white material and inspect them with a magnifying glass.

Anyway, good to see you again.

Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
MaxDarkside
Posts: 1199
Joined: Sun Dec 18, 2011 4:21 pm
Location: Minneapolis, MN

Re: Flecks in Humidifer Tank

Post by MaxDarkside » Thu Jan 19, 2017 11:57 pm

chunkyfrog wrote:I wonder if flakes are coming off the plastic parts, due to age.
How old is the tank?
Hi ChunkyFrog!!

The tank is pretty new, less than a year. I got it here from cpap.com and it's an original new Resmed tank.

Thanks!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50E Oximeter, Zeo Bedside, Software: Comm'l grade AI analytics server & tools & SleepyHead
Do or Die... Sleep Apnea killed me, but I came back. Click for my story
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs

User avatar
MaxDarkside
Posts: 1199
Joined: Sun Dec 18, 2011 4:21 pm
Location: Minneapolis, MN

Re: Flecks in Humidifer Tank

Post by MaxDarkside » Fri Jan 20, 2017 12:05 am

Wulfman... wrote:
MaxDarkside wrote:Hi All,

Long time no post. I noticed light colored flecks with edges and pointy corners in my humidifier tank (Resmed S9 purchased Nov 2011) so I of course cleaned it, by hand, dismantling the tank as it is a pops-open dishwasher safe tank, disinfected it, buffed 'er to a shine inside, filled it with distilled water. The next morning... more flecks. Is this one of the failure modes of the Resmed S9, like some seal inside coming apart?

I replace the filter quite often, using CPAP.COM filters.

Thanks!
Where ya been, Max?

Floating on the top or lying on the bottom of the tank?
In any case, they won't be carried out of the HH tank with the water vapor......too heavy.
Maybe try filtering them out on some clean, white material and inspect them with a magnifying glass.

Anyway, good to see you again.

Den

.
Hey Wulfman!

They lurk on the bottom, heavier than water.

What have I been up too? Well, I got me 3 stents to the circumflex artery of the heart about a year and a half ago. Sleep Apnea related? The cardiologist thinks its possible. They put me on cardiac drugs then my Lisinopril decided to try to kill me with angio adema. 2 trips to the ER, one by ambulance with lights and sirens tracked that down. I went through cardiac rehab and then off to one of those cheap gyms in strip malls... started getting nasty aggressive staph infections monthly in my hands, once on my knee (think necretizing attack, hole in your leg forming almost before your eyes). Even used clorox wipes and all sorts of sanitation. After visits to internal medicine, immunologist, dermatologist and other docs, they said it was probably the gym. I stopped going and so did the infections, so far (knock on wood).

Other than that, not much LOL

Keep Healthy!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50E Oximeter, Zeo Bedside, Software: Comm'l grade AI analytics server & tools & SleepyHead
Do or Die... Sleep Apnea killed me, but I came back. Click for my story
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs

User avatar
MaxDarkside
Posts: 1199
Joined: Sun Dec 18, 2011 4:21 pm
Location: Minneapolis, MN

Re: Flecks in Humidifer Tank

Post by MaxDarkside » Fri Jan 20, 2017 12:10 am

grayghost4 wrote:There is nothing inside that would come apart and cause that .... silicone gasket ... black foam rubber .... all the rest is hard plastic.
Might it be something in the distilled water ?
Hi GrayGhost!

I looked at tonight's and they are actually little globs, not hard after all. Soooo... since I sanitize the tank really well, with chemicals, it could be :

1. something in the air that gets past the filter
2. something inside the main unit, like bacteria, mold, mildew, ...
3. something in the water (thanks, I didn't think of that, I use store brand Target distilled water)

I recently went for a cruise in Hawaii and this started after that. Maybe a new buddy came back with me from the islands and is renting space inside my Resmed.

Thank you much!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50E Oximeter, Zeo Bedside, Software: Comm'l grade AI analytics server & tools & SleepyHead
Do or Die... Sleep Apnea killed me, but I came back. Click for my story
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: Flecks in Humidifer Tank

Post by grayghost4 » Fri Jan 20, 2017 12:46 am

If it is something for the Islands ... the unit will have to be taken apart and cleaned.
The foam pads could be a place that could harbour something.
If you are a little mechanical you can do it yourself ... there is a thread that will help you.
If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: Flecks in Humidifer Tank

Post by Jay Aitchsee » Fri Jan 20, 2017 5:44 am

Hey, Max!
I have no idea about your flecks, but it's good to hear from you.
j

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Flecks in Humidifer Tank

Post by chunkyfrog » Fri Jan 20, 2017 11:22 am

Indeed! I missed you!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
MaxDarkside
Posts: 1199
Joined: Sun Dec 18, 2011 4:21 pm
Location: Minneapolis, MN

Re: Flecks in Humidifer Tank

Post by MaxDarkside » Fri Jan 20, 2017 4:19 pm

grayghost4 wrote:If it is something for the Islands ... the unit will have to be taken apart and cleaned.
The foam pads could be a place that could harbour something.
If you are a little mechanical you can do it yourself ... there is a thread that will help you.
Thanks, GrayGhost. I may do that.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50E Oximeter, Zeo Bedside, Software: Comm'l grade AI analytics server & tools & SleepyHead
Do or Die... Sleep Apnea killed me, but I came back. Click for my story
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs

User avatar
MaxDarkside
Posts: 1199
Joined: Sun Dec 18, 2011 4:21 pm
Location: Minneapolis, MN

Re: Flecks in Humidifer Tank

Post by MaxDarkside » Fri Jan 20, 2017 4:21 pm

Jay Aitchsee wrote:Hey, Max!
I have no idea about your flecks, but it's good to hear from you.
j
Hi Jay!

Thanks. You still working with various devices and technology? I haven't much lately myself at least regarding sleep.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50E Oximeter, Zeo Bedside, Software: Comm'l grade AI analytics server & tools & SleepyHead
Do or Die... Sleep Apnea killed me, but I came back. Click for my story
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs

User avatar
MaxDarkside
Posts: 1199
Joined: Sun Dec 18, 2011 4:21 pm
Location: Minneapolis, MN

Re: Flecks in Humidifer Tank

Post by MaxDarkside » Fri Jan 20, 2017 4:22 pm

chunkyfrog wrote:Indeed! I missed you!
I've missed you too! I hope you are doing great!

Thanks!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50E Oximeter, Zeo Bedside, Software: Comm'l grade AI analytics server & tools & SleepyHead
Do or Die... Sleep Apnea killed me, but I came back. Click for my story
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: Flecks in Humidifer Tank

Post by Jay Aitchsee » Sat Jan 21, 2017 5:58 am

MaxDarkside wrote:Hi Jay!Thanks. You still working with various devices and technology? I haven't much lately myself at least regarding sleep.
No, not so much, Max. My sleep has been pretty good the last couple of years, so I've kind of lost interest. Once in a while, out of curiosity, I wear the Zeo, but mostly it serves as a nice clock for my nightstand. Most of my time, now, is taken up by maintenance, remodel and repair projects around the house that I can never seem to finish - which my wife likes to point out from time to time.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Flecks in Humidifer Tank

Post by palerider » Sat Jan 21, 2017 12:49 pm

MaxDarkside wrote:
grayghost4 wrote:If it is something for the Islands ... the unit will have to be taken apart and cleaned.
The foam pads could be a place that could harbour something.
If you are a little mechanical you can do it yourself ... there is a thread that will help you.
Thanks, GrayGhost. I may do that.
how to dissect a S9: http://imgur.com/a/S0ojG